Fwd: about sigma2 for exafs fitting
Shaofeng has given me permission to repost her question here on the ifeffit mailing list. It is quoted below my response.
Dear Shaofeng,
As Bruce and I said before, a sigma2 of 0.0007 A^2 is not impossible, although it indicates less disorder than is typically present. Your attached table does seem to show some improvement by using the model from the hydrogen-containing structure as compared to the arsenate. A more rigorous test for statistical improvement can be conducted using the Hamilton test (you mentioned you’ve consulted XAFS for Everyone; full details of the Hamilton test are given there).
It’s also encouraging that the uncertainties on your sigma2 determinations using the hydrogen-containing model are quite small; it appears that the fit is not getting confused by correlations even though it’s fitting both coordination number and sigma2, as that would generally also cause high uncertainties in the correlated parameters.
Is such a stiff sigma2 reasonable in this case? I have no idea. I just don’t know enough about this particular system; perhaps someone else on the list does. Oh, and one other question—was the data collected at room temperature? If it were collected at cryogenic temperatures, that would tend to reduce thermal disorder and thus lower sigma2’s.
Even if no one on the list has insight in to this particular system, anyone have good published examples of room-temperature systems with sigma2’s < 0.001? It might help Shaofeng with her referee…
Best,
Scott Calvin
Lehman College of the City University of New York
P. S. I certainly hope no reviewer is using the “typical values” I provide for parameters in XAFS for Everyone as rigid criteria for rejecting results! It is most certainly not the way I use them in the book. Some systems actually are atypical!
Begin forwarded message:
From: Shaofeng Wang
Dear Scott,
Thanks for your explaination. All the samples were measured at 77K using a homemade cryogenic box. I will try to do the Hamilton test according to your book.
Best,
Shaofeng
--------------------------------------
Shaofeng Wang, Ph.D of Geochemistry
Environmental Molecular Science Group
Institute of Applied Ecology, Chinese Academy of Sciences
Shenyang, 110016, China
wangshaofeng@iae.ac.cn
www.iae.cas.cn
From: Scott Calvin
Sent: Thursday, January 05, 2017 11:50 AM
To: XAFS Analysis using Ifeffit
Cc: Shaofeng Wang
Subject: Fwd: about sigma2 for exafs fitting
Shaofeng has given me permission to repost her question here on the ifeffit mailing list. It is quoted below my response.
Dear Shaofeng,
As Bruce and I said before, a sigma2 of 0.0007 A^2 is not impossible, although it indicates less disorder than is typically present. Your attached table does seem to show some improvement by using the model from the hydrogen-containing structure as compared to the arsenate. A more rigorous test for statistical improvement can be conducted using the Hamilton test (you mentioned you’ve consulted XAFS for Everyone; full details of the Hamilton test are given there).
It’s also encouraging that the uncertainties on your sigma2 determinations using the hydrogen-containing model are quite small; it appears that the fit is not getting confused by correlations even though it’s fitting both coordination number and sigma2, as that would generally also cause high uncertainties in the correlated parameters.
Is such a stiff sigma2 reasonable in this case? I have no idea. I just don’t know enough about this particular system; perhaps someone else on the list does. Oh, and one other question—was the data collected at room temperature? If it were collected at cryogenic temperatures, that would tend to reduce thermal disorder and thus lower sigma2’s.
Even if no one on the list has insight in to this particular system, anyone have good published examples of room-temperature systems with sigma2’s < 0.001? It might help Shaofeng with her referee…
Best,
Scott Calvin
Lehman College of the City University of New York
P. S. I certainly hope no reviewer is using the “typical values” I provide for parameters in XAFS for Everyone as rigid criteria for rejecting results! It is most certainly not the way I use them in the book. Some systems actually are atypical!
Begin forwarded message:
From: Shaofeng Wang
Dear Scott,
I have learned the Hamilton test. However, this method seems not suitable to distinguish our results because the Nvar are same for the two fittings (3 when S02 was fixed). So the b value should be zero and the calculator on the website http://www.danielsoper.com/statcalc/calculator.aspx?id=37 can not carry on.
In addition, I am not sure the value x. Is it the ratio of R-factors between two fits? I calculated the results using a=2.18 (Nidp and variables (Nvar) were 7.35 and 3), b=0.00001, and x=0.818 (the ratio of two R-factors) and got the Regularized lower incomplete beta function of 0.00000807. Does it mean something?
Cheers,
Shaofeng
--------------------------------------
Shaofeng Wang, Ph.D of Geochemistry
Environmental Molecular Science Group
Institute of Applied Ecology, Chinese Academy of Sciences
Shenyang, 110016, China
wangshaofeng@iae.ac.cn
www.iae.cas.cn
From: Scott Calvin
Sent: Thursday, January 05, 2017 11:50 AM
To: XAFS Analysis using Ifeffit
Cc: Shaofeng Wang
Subject: Fwd: about sigma2 for exafs fitting
Shaofeng has given me permission to repost her question here on the ifeffit mailing list. It is quoted below my response.
Dear Shaofeng,
As Bruce and I said before, a sigma2 of 0.0007 A^2 is not impossible, although it indicates less disorder than is typically present. Your attached table does seem to show some improvement by using the model from the hydrogen-containing structure as compared to the arsenate. A more rigorous test for statistical improvement can be conducted using the Hamilton test (you mentioned you’ve consulted XAFS for Everyone; full details of the Hamilton test are given there).
It’s also encouraging that the uncertainties on your sigma2 determinations using the hydrogen-containing model are quite small; it appears that the fit is not getting confused by correlations even though it’s fitting both coordination number and sigma2, as that would generally also cause high uncertainties in the correlated parameters.
Is such a stiff sigma2 reasonable in this case? I have no idea. I just don’t know enough about this particular system; perhaps someone else on the list does. Oh, and one other question—was the data collected at room temperature? If it were collected at cryogenic temperatures, that would tend to reduce thermal disorder and thus lower sigma2’s.
Even if no one on the list has insight in to this particular system, anyone have good published examples of room-temperature systems with sigma2’s < 0.001? It might help Shaofeng with her referee…
Best,
Scott Calvin
Lehman College of the City University of New York
P. S. I certainly hope no reviewer is using the “typical values” I provide for parameters in XAFS for Everyone as rigid criteria for rejecting results! It is most certainly not the way I use them in the book. Some systems actually are atypical!
Begin forwarded message:
From: Shaofeng Wang
Hi Shaofeng,
The parameter b does not have to do solely with the difference in N_var between the two fits! As I say in the book,
“Changes in the theoretical standards used by the model can often be accounted for as if they were changes in parameters. For example, if fit A assumed that the nearest neighbors were oxygen and included coordination number, bond length, and MSRD for those oxygen atoms as free parameters, while fit B assumed the nearest neighbors were sulfur and used the same three parameters, then a Hamilton test could be applied with the assumption that those three parameters were “changed”—instead of applying to oxygen, they now apply to sulfur. If E0 and S02 were also free parameters, they would not be considered to be changed, as they are primarily a property of the absorber, not the scatterer.”
In your case, it looks to me like two or three parameters are changed between models (I count four free parameters in your table, one of which is E0, so I’m not sure why you say there are only three…is there another constraint?) Suppose there are two free parameters changed. b = 2/2 = 1.
x is the ratio of R-factors between the fits.
Using the values you list below, then, gives a result of 0.64, which indicates such variation is quite likely to have taken place by chance.
You chose the worst case of your three, though; the pH 3 case. The pH 10 case shows the most improvement in the R-factor with x = 0.007/0.012 = 0.58. The probability of getting that much improvement by chance drops to 0.30—less likely, but still pretty close to a coin flip.
Even the fact that you got all three to show improvement doesn’t quite reach the 95% confidence level, although it’s becoming suggestive.
Why, then, are fits that seem to have much lower R-factors not attaining statistical significance? The main culprit is that you have a relatively small number of independent points for studying this kind of system.
So where’s that leave you?
I’ll stand by the results of the Hamilton test…there’s not quite enough data here, on its own, for me to decide in favor of your second model. If you have other evidence pointing in that direction, these results could be used as supporting evidence. But with the combination of the closeness of fit not improving to a statistically significant degree and the sigma2 taking on an unusual (but not impossible) value, there needs to be more to the argument, in my opinion.
Best,
Scott Calvin
Lehman College of the City University of New York
On Jan 5, 2017, at 12:19 AM, Shaofeng Wang
Dear Scott and demter users,
I also tried to use BaHAsO4 crystallographic data as the initial model for fitting of the EXAFS data. Please see the attached file. It was found the sigma2 was around 0.0012-0.0014. The Hamilton tests also show the improvement was not significant (> >0.05). But thease results could be considered to be better than using DFT optimized Configuration C2 as the initial model. This reuslt may imply the sigma2 is strongly denpendent on the initial model. Does this mean the sigma2 is not so important to determine the fitting quality?
Regards,
Shaofeng
--------------------------------------
Shaofeng Wang, Ph.D of Geochemistry
Environmental Molecular Science Group
Institute of Applied Ecology, Chinese Academy of Sciences
Shenyang, 110016, China
wangshaofeng@iae.ac.cn
www.iae.cas.cn
From: Scott Calvin
Sent: Friday, January 06, 2017 3:34 AM
To: XAFS Analysis using Ifeffit
Cc: Shaofeng Wang
Subject: Re: about sigma2 for exafs fitting
Hi Shaofeng,
The parameter b does not have to do solely with the difference in N_var between the two fits! As I say in the book,
“Changes in the theoretical standards used by the model can often be accounted for as if they were changes in parameters. For example, if fit A assumed that the nearest neighbors were oxygen and included coordination number, bond length, and MSRD for those oxygen atoms as free parameters, while fit B assumed the nearest neighbors were sulfur and used the same three parameters, then a Hamilton test could be applied with the assumption that those three parameters were “changed”—instead of applying to oxygen, they now apply to sulfur. If E0 and S02 were also free parameters, they would not be considered to be changed, as they are primarily a property of the absorber, not the scatterer.”
In your case, it looks to me like two or three parameters are changed between models (I count four free parameters in your table, one of which is E0, so I’m not sure why you say there are only three…is there another constraint?) Suppose there are two free parameters changed. b = 2/2 = 1.
x is the ratio of R-factors between the fits.
Using the values you list below, then, gives a result of 0.64, which indicates such variation is quite likely to have taken place by chance.
You chose the worst case of your three, though; the pH 3 case. The pH 10 case shows the most improvement in the R-factor with x = 0.007/0.012 = 0.58. The probability of getting that much improvement by chance drops to 0.30—less likely, but still pretty close to a coin flip.
Even the fact that you got all three to show improvement doesn’t quite reach the 95% confidence level, although it’s becoming suggestive.
Why, then, are fits that seem to have much lower R-factors not attaining statistical significance? The main culprit is that you have a relatively small number of independent points for studying this kind of system.
So where’s that leave you?
I’ll stand by the results of the Hamilton test…there’s not quite enough data here, on its own, for me to decide in favor of your second model. If you have other evidence pointing in that direction, these results could be used as supporting evidence. But with the combination of the closeness of fit not improving to a statistically significant degree and the sigma2 taking on an unusual (but not impossible) value, there needs to be more to the argument, in my opinion.
Best,
Scott Calvin
Lehman College of the City University of New York
On Jan 5, 2017, at 12:19 AM, Shaofeng Wang
Hi Shaofeng,
Thanks for providing a thorough description of what you did in the Word file—it makes it much easier to comment!
I wouldn’t say that sigma2 is not important, in this case. I would say, instead, that your fits have not yet successfully distinguished between the different proposed models, although some of those models yields significantly different values for sigma2.
Here’s my summary of what I’m seeing in your file and from your description:
The Hamilton test doesn’t indicate the differences between fits are statistically significant.
The Ba3(AsO4)2 model yields E0’s a bit smaller than the corresponding standard (although with overlapping error bars) and sigma2’s significantly larger. The larger sigma2 could just indicate it’s somewhat more disordered. I certainly can’t reject this model on the basis of the information you’ve provided, but it seems a little less likely than the others.
The BaHAsO4*H2O model yields parameters almost identical to the corresponding standard. Yay! The results are therefore consistent with the sample being BaHAsO4*H2O.
The C2 model again yields parameters that are a bit different from the fits to the standards. The sigma2’s are, as you mentioned, small. The E0’s are also large compared to the standard, although the error bars overlap.
This raises, for me, a question: did you run a fit with the BaHAsO4*H2O standard using the C2 model? I know that’s not the structure you expect for the standard, but it would be very nice to know if using that model with the standard also yields small sigma2’s, big E0’s, and smaller R-factors. If that happens, then you’d gain confidence that you had BaHAsO4*H2O, but you also wouldn’t have any evidence that the C2 model was a better description of your samples than just the regular BaHAsO4*H2O structure. If, on the other hand, the C2 model doesn’t yield a similar set of results for the standard as it did for the sample, then you’ve got some more investigating to do.
Does that make sense?
On Jan 8, 2017, at 6:53 AM, Shaofeng Wang
participants (2)
-
Scott Calvin -
Shaofeng Wang