HI Eugenio,
Have you seen this work? I came across it last night.
https://www.nature.com/articles/srep43791
"Local structures around the substituted elements in mixed layered oxides"
They experienced a similar issue:
"... In the O3 compound, we included the contribution from the first-
(O), second- (/M/), and third- (Na) nearest neighbor elements. In the P2
compounds, we included the contributions from the first- (O), third-
(/M/), and fifth- (O) nearest neighbor elements. We omitted the
contribution from the second- (Na2) and fourth- (Na1) nearest neighbor
elements because the occupancy (/g/ = 0.21–0.41) is small. In both
cases, we fixed /N/_j for the three elements at the crystallographic
value (=6)..."
"...We found that the R values of the third paths, /i.e/. the /M/–Na
distances in the O3 compounds and the longer /M/ - O distances of the P2
compounds, are seriously deviated from the crystallographic values. In
addition, the/σ/_j ^2 values for these paths are significantly large
(~0.01–0.03 Å^2 ) as compared with those of the first and second paths.
In this sense, the EXAFS analysis model including the third paths
remains ambiguous. We performed another EXAFS analysis with omitting the
third path to validate the discussion above. We observed the same trend
of the /M/ – O and /M/ – /M/ distances between the compounds even if we
omitted the third path...."
Disorder does seem to be an issue. Do you have a pure phase result to
which you can compare
i.e. unaltered or non-substituted?
cheers,
-R.
On 2024-01-16 11:40 p.m., Otal Eugenio wrote:
> Dear Carlo,
> Thanks for your answer.
> Yes, it is a layered material. I was measuring in situ the
> transformation from O3 to P3.
> I am not sure that Na atoms are ordered, it is a light element in a
> layered structure and I applied some cycles to form the
> electrolyte-cathode interphase, and disorder can be found there. I
> shook the structure a little, your observation is correct.
> I compared the spectra of a non-cycled one (blue) with a cycled one
> (red) and the differences are minimal.
> image.pngimage.png
> I was measuring all the edges, all showed the same problem, Na is
> impossible to include in the fitting.
> The goal was to monitor changes in the first coordination shell during
> the battery operation. I found all the necessary information to
> understand the system.
> To fit the second coordination shell was attempted to evaluate the Na
> concentration along the cycles and complement the electrochemical
> measurements.
> Best regards.
>
>
>
>
> (^ㅇᆽㅇ^)(=˃ᆺ˂=)(=🝦 ༝ 🝦=)
>
> Eugenio H. OTAL
> Assistant Professor
> Dept. of Materials Chemistry
> Shinshu University
> 4-17-1 Wakasato, Nagano 380-8553, JAPAN
> eugenio_otal at shinshu-u.ac.jp <mailto:eugenio_otal at shinshu-u.ac.jp>
> https://sites.google.com/view/zettsu-laboratory/news-updates
> <https://sites.google.com/view/zettsu-laboratory/news-updates>
>
> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>
> Who is John Galt?
>
> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>
>
>
> On Wed, Jan 17, 2024 at 2:56 PM Carlo Segre <segre at iit.edu> wrote:
>
> Hi Eugenio:
>
> This looks like a layered structure, most likely a P2 or P3
> phase. Is this correct? How sure are you that you have a single
> phase and that the Na ions are well ordered? Is it possible that
> there is so much disorder in the M-Na distances that they wash out
> any contribution? In that case, the second M-O distance would
> probably be more important. You also do not mention if this is
> Ni, Fe or Mn edge data. Is it possible that these differ?
>
> Carlo
>
>
>
> On Tue, Jan 16, 2024 at 11:30 PM Otal Eugenio
> <eugenio_otal at shinshu-u.ac.jp> wrote:
>
> Dear Robert,
> Thanks for your detailed answer.
> I am cutting at R=3.35. I also tried to go up to 4.0 to
> provide more freedom and to include the SS TM-O at 3.6.
> It seems that including MS was not fruitful at the moment. I
> found the contribution very tiny, Importance < 10.
> I included the path at 3.6, and it fits well, but when I
> include the Na, the CN_Na goes to high values (>10) and errors
> (~+/- 50)
> I also tried to restrict the CN of TM to force the Na to take
> place and get good results. Also useless.
> I was checking the data in R_real and the TM and Na paths. I
> see the TM path is in phase while the Na path is not. I wonder
> if this is affecting the fit. How should I constrain the
> fitting to solve this situation?
> Best regards.
>
> image.png
>
>
>
> (^ㅇᆽㅇ^)(=˃ᆺ˂=)(=🝦 ༝ 🝦=)
>
> Eugenio H. OTAL
> Assistant Professor
> Dept. of Materials Chemistry
> Shinshu University
> 4-17-1 Wakasato, Nagano 380-8553, JAPAN
> eugenio_otal at shinshu-u.ac.jp <mailto:eugenio_otal at shinshu-u.ac.jp>
> https://sites.google.com/view/zettsu-laboratory/news-updates
> <https://sites.google.com/view/zettsu-laboratory/news-updates>
>
> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>
> Who is John Galt?
>
> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>
>
>
> On Wed, Jan 17, 2024 at 1:24 AM Robert Gordon
> <ragordon at alumni.sfu.ca> wrote:
>
> Hi Otal,
>
> I ran a quick FEFF simulation just to see what is present
> under that second peak in the FT.
> There are many more shells than just TM and Na...e.g.
>
>
> There is significant multiple scattering and another
> backscattering peak (Fe-O) at
> 3.6Angstroms. All of these are going to spread underneath
> that peak in the FT. You might
> also be getting some interference from the 4.07A MS stuff.
> It doesn't seem a surprise that
> just including TM and Na is giving problems.
>
> What is your high-R cutoff? Have you tried including
> multiple-scattering or that 4th backscattering
> contribution?
>
> cheers,
> Robert
>
> On 2024-01-09 12:06 a.m., Otal Eugenio wrote:
>> Dear all,
>> Happy New Year!
>> I am dealing with the fitting of NaMn0.3Fe0.4Ni0.3O2
>> second shell.
>> I have Mn, Ni, and Fe at 2.9772Å and Na at 3.1859Å,
>> showing only one shell. If I use any of the transition
>> metals (dZ <5) for fitting the second coordination shell.
>> fitting is fine, but when I introduce the Na path, the
>> fitting has no sense, CN_Na grows and also its error.
>> Even including Na, I have Nvary = 10 and Ninp = 13.98, so
>> I am getting out of ideas and nothing helpful in the
>> ifeffit Archive until now.
>> Is there any strategy for dealing with mixed second
>> coordination shells? I am attaching the file exported
>> from Larch and the cif file.
>> Best regards.
>>
>>
>> (^ㅇᆽㅇ^)(=˃ᆺ˂=)(=🝦 ༝ 🝦=)
>>
>> Eugenio H. OTAL
>> Assistant Professor
>> Dept. of Materials Chemistry
>> Shinshu University
>> 4-17-1 Wakasato, Nagano 380-8553, JAPAN
>> eugenio_otal at shinshu-u.ac.jp
>> <mailto:eugenio_otal at shinshu-u.ac.jp>
>> https://sites.google.com/view/zettsu-laboratory/news-updates
>> <https://sites.google.com/view/zettsu-laboratory/news-updates>
>>
>> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>>
>> Who is John Galt?
>>
>> ¯\_(ツ)_/¯¯\_(ツ)_/¯
>>
>>
>> _______________________________________________
>> Ifeffit mailing list
>> Ifeffit at millenia.cars.aps.anl.gov
>> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>> Unsubscribe:http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe:
> http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe:
> http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>
>
>
> --
> Carlo U. Segre (he/him) -- Duchossois Leadership Professor of Physics
> Professor of Materials Science & Engineering
> Director, Center for Synchrotron Radiation Research and
> Instrumentation
> Illinois Institute of Technology
> Phone: 312.567.3498
> segre at iit.edu http://phys.iit.edu/~segre segre at debian.org
>
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe:http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20240117/58c326f1/attachment-0001.htm>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 142712 bytes
Desc: not available
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20240117/58c326f1/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 113062 bytes
Desc: not available
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20240117/58c326f1/attachment-0005.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 97278 bytes
Desc: not available
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20240117/58c326f1/attachment-0006.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 6m1qwMkGPObtzPU2.png
Type: image/png
Size: 49298 bytes
Desc: not available
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20240117/58c326f1/attachment-0007.png>
More information about the Ifeffit
mailing list