Hi Valerie,
On Tue, Mar 2, 2021 at 10:48 AM Schoepfer, Valerie <
valerie.schoepfer at usask.ca> wrote:
> Hi,
>
>
>
> I am using Larch XAS Viewer for the first time to analyze some Mo XANES
> data. It is fairly straightforward, however I am running into problems with
> energy calibration, which leads to problems with Linear Combination
> Fitting.
>
>
>
> Right now, which is likely wrong, I’m calibrating the energy of a standard
> to the theoretical edge, then auto-aligning the samples to the standard.
> But how do reference foils fit in here? Reference foils don’t seem to be
> tied to the sample like they are in Athena. Should I be manually aligning?
>
>
>
> Is there a general guidance or work flow?
>
It's possible that I do not fully understand the question or that this
answer will veer a bit off the topic of your question.
For sure, energies need to be aligned properly for any multi-spectra
comparison or linear method to work well. But it should be that you will
have groups of spectra that all share a consistent energy calibration, say
from the same beamline/beamtime.
If you do have a reference channel for every measurement, you can compare
those reference channels. Ideally, these will not vary for every
measurement - that would indicate a serious problem. So, I think you
should be able to group spectra together as uniformly calibrated (hopefully
all data from a day or more of beamtime at a particular beamline) and then
make sure that the different groups of spectra. Does that seem reasonable?
I have to admit that at my beamline I don't often have the luxury or need
to run a reference foil for every scan, so we calibrate consistently ahead
of time. I'm sure that leads to a bias in the software. I guess I forgot
that Athena had the ability to read and tie a second spectrum as a
"reference" and use that to auto-apply calibration.
Is it generally necessary to calibrate many spectra individually, or do
people find that doing them in a few large groups is sufficient?
--Matt