[Ifeffit] Unwanted rescaling of energy axis when importing ascii file into Athena (Demeter 0.9.26)
Robert Gordon
ragordon at alumni.sfu.ca
Wed Jan 9 11:26:39 CST 2019
Hi,
I switched your data from keV to eV before importing, and that does not
have the same issue.
-R.
On 2019-01-09 6:35 a.m., Lukas Grote wrote:
>
> Hello,
>
> I am reporting an apparent bug in Athena (Demeter 0.9.26, Windows 10).
> When I import high energy resolution XAS data from an ascii file, the
> energy axis is automatically rescaled to 1.2eV step size, thus
> drastically reducing the resolution of the data. When the imported
> data has energy steps larger than 1.2eV, no rescaling occurs.
>
> This issue may be similar to the one reported in March 2018 under the
> topic "Athena (Demeter 0.9.26) does not import all datapoints from
> ascci file". However, the issue I am seeing always results in 1.2eV
> steps in the imported data. If my original file hast 0.3eV steps, the
> number of points is reduced by a factor of 4. If it has 0.6eV steps,
> the number of points is reduced by a factor of 2.
>
> I couldn't find any option to switch this recaling on / off, or to
> change the target energy step size.
>
> The problem does not occur with Demeter 0.9.25. Attached please find
> two screenshots of the data import dialog for both versions of
> Demeter. The plots show the difference in energy scale. I also
> attached the original data file "no_rebin.dat" as well as spectra
> exported from both Demeter versions, one of them showing the 1.2eV steps.
>
> Thank you and all the best,
> Lukas
>
> --
> Lukas Grote
>
> University of Hamburg
> Center for Hybrid Nanostructures
> Institute for Nanostructures and Solid State Physics
> Luruper Chaussee 149, Bld. 600, Room 2.62
> 22761 Hamburg
> Germany
>
> +49 40 42838 1614
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20190109/c4c23b31/attachment-0001.html>
-------------- next part --------------
# created: Wed Jan 09 09:11:08 2019
# from file: no_rebin.dat
# No: 300 [E0: 0,0,0,-1]
#---------------------------------------------------------------------
# X Y
7.6950000000E+003 1.6270000000E-003
7.6953000000E+003 0.0000000000E+000
7.6956000000E+003 0.0000000000E+000
7.6959000000E+003 4.0000000000E-006
7.6962000000E+003 0.0000000000E+000
7.6965000000E+003 0.0000000000E+000
7.6968000000E+003 0.0000000000E+000
7.6971000000E+003 0.0000000000E+000
7.6974000000E+003 1.9000000000E-004
7.6977000000E+003 2.2300000000E-004
7.6980000000E+003 0.0000000000E+000
7.6983000000E+003 9.5000000000E-005
7.6986000000E+003 0.0000000000E+000
7.6989000000E+003 0.0000000000E+000
7.6992000000E+003 5.1300000000E-004
7.6995000000E+003 0.0000000000E+000
7.6998000000E+003 0.0000000000E+000
7.7001000000E+003 0.0000000000E+000
7.7004000000E+003 0.0000000000E+000
7.7007000000E+003 6.8500000000E-004
7.7010000000E+003 0.0000000000E+000
7.7013000000E+003 0.0000000000E+000
7.7016000000E+003 1.2120000000E-003
7.7019000000E+003 1.2520000000E-003
7.7022000000E+003 0.0000000000E+000
7.7025000000E+003 0.0000000000E+000
7.7028000000E+003 0.0000000000E+000
7.7031000000E+003 5.8500000000E-004
7.7034000000E+003 0.0000000000E+000
7.7037000000E+003 2.2300000000E-004
7.7040000000E+003 0.0000000000E+000
7.7043000000E+003 0.0000000000E+000
7.7046000000E+003 0.0000000000E+000
7.7049000000E+003 0.0000000000E+000
7.7052000000E+003 0.0000000000E+000
7.7055000000E+003 1.0510000000E-003
7.7058000000E+003 1.5050000000E-003
7.7061000000E+003 2.8670000000E-003
7.7064000000E+003 3.2340000000E-003
7.7067000000E+003 5.9870000000E-003
7.7070000000E+003 8.7590000000E-003
7.7073000000E+003 1.1040000000E-002
7.7076000000E+003 2.1544000000E-002
7.7079000000E+003 2.4197000000E-002
7.7082000000E+003 2.7295000000E-002
7.7085000000E+003 2.5455000000E-002
7.7088000000E+003 2.6229000000E-002
7.7091000000E+003 2.8788000000E-002
7.7094000000E+003 2.1862000000E-002
7.7097000000E+003 2.1495000000E-002
7.7100000000E+003 1.5811000000E-002
7.7103000000E+003 1.2947000000E-002
7.7106000000E+003 1.2381000000E-002
7.7109000000E+003 1.4895000000E-002
7.7112000000E+003 1.5022000000E-002
7.7115000000E+003 1.7291000000E-002
7.7118000000E+003 2.0151000000E-002
7.7121000000E+003 1.7761000000E-002
7.7124000000E+003 2.0716000000E-002
7.7127000000E+003 2.2911000000E-002
7.7130000000E+003 3.0756000000E-002
7.7133000000E+003 3.4165000000E-002
7.7136000000E+003 3.8639000000E-002
7.7139000000E+003 5.0613000000E-002
7.7142000000E+003 6.4706000000E-002
7.7145000000E+003 8.6778000000E-002
7.7148000000E+003 9.9867000000E-002
7.7151000000E+003 1.2476600000E-001
7.7154000000E+003 1.4072200000E-001
7.7157000000E+003 1.7525500000E-001
7.7160000000E+003 2.0230000000E-001
7.7163000000E+003 2.3248200000E-001
7.7166000000E+003 2.8919900000E-001
7.7169000000E+003 3.3403200000E-001
7.7172000000E+003 3.6579700000E-001
7.7175000000E+003 4.2088200000E-001
7.7178000000E+003 4.6328900000E-001
7.7181000000E+003 5.0946800000E-001
7.7184000000E+003 5.5738400000E-001
7.7187000000E+003 6.1336900000E-001
7.7190000000E+003 6.7899100000E-001
7.7193000000E+003 7.4188800000E-001
7.7196000000E+003 8.3672200000E-001
7.7199000000E+003 9.2037700000E-001
7.7202000000E+003 1.0117220000E+000
7.7205000000E+003 1.0654910000E+000
7.7208000000E+003 1.1327360000E+000
7.7211000000E+003 1.2186160000E+000
7.7214000000E+003 1.2526620000E+000
7.7217000000E+003 1.3064720000E+000
7.7220000000E+003 1.3403760000E+000
7.7223000000E+003 1.3708070000E+000
7.7226000000E+003 1.3870930000E+000
7.7229000000E+003 1.3998280000E+000
7.7232000000E+003 1.3977760000E+000
7.7235000000E+003 1.4361000000E+000
7.7238000000E+003 1.4106380000E+000
7.7241000000E+003 1.3939930000E+000
7.7244000000E+003 1.3845650000E+000
7.7247000000E+003 1.3617420000E+000
7.7250000000E+003 1.3478580000E+000
7.7253000000E+003 1.3001120000E+000
7.7256000000E+003 1.3252590000E+000
7.7259000000E+003 1.2916750000E+000
7.7262000000E+003 1.2817510000E+000
7.7265000000E+003 1.2542000000E+000
7.7268000000E+003 1.2537460000E+000
7.7271000000E+003 1.2388830000E+000
7.7274000000E+003 1.1935380000E+000
7.7277000000E+003 1.1836590000E+000
7.7280000000E+003 1.1413890000E+000
7.7283000000E+003 1.1278830000E+000
7.7286000000E+003 1.1217750000E+000
7.7289000000E+003 1.1111540000E+000
7.7292000000E+003 1.0972530000E+000
7.7295000000E+003 1.0851250000E+000
7.7298000000E+003 1.0795580000E+000
7.7301000000E+003 1.0682330000E+000
7.7304000000E+003 1.0606470000E+000
7.7307000000E+003 1.0611010000E+000
7.7310000000E+003 1.0624460000E+000
7.7313000000E+003 1.0492600000E+000
7.7316000000E+003 1.0427060000E+000
7.7319000000E+003 1.0497640000E+000
7.7322000000E+003 1.0331290000E+000
7.7325000000E+003 1.0500160000E+000
7.7328000000E+003 1.0582000000E+000
7.7331000000E+003 1.0471490000E+000
7.7334000000E+003 1.0478420000E+000
7.7337000000E+003 1.0336490000E+000
7.7340000000E+003 1.0434330000E+000
7.7343000000E+003 1.0501010000E+000
7.7346000000E+003 1.0385220000E+000
7.7349000000E+003 1.0648180000E+000
7.7352000000E+003 1.0539400000E+000
7.7355000000E+003 1.0445600000E+000
7.7358000000E+003 1.0482300000E+000
7.7361000000E+003 1.0446040000E+000
7.7364000000E+003 1.0474000000E+000
7.7367000000E+003 1.0404650000E+000
7.7370000000E+003 1.0294330000E+000
7.7373000000E+003 1.0244230000E+000
7.7376000000E+003 1.0020190000E+000
7.7379000000E+003 1.0146750000E+000
7.7382000000E+003 1.0036220000E+000
7.7385000000E+003 9.8587800000E-001
7.7388000000E+003 9.8001300000E-001
7.7391000000E+003 9.7647700000E-001
7.7394000000E+003 9.6251000000E-001
7.7397000000E+003 9.7169100000E-001
7.7400000000E+003 9.4698100000E-001
7.7403000000E+003 9.4486900000E-001
7.7406000000E+003 9.3564700000E-001
7.7409000000E+003 9.1952000000E-001
7.7412000000E+003 9.0984100000E-001
7.7415000000E+003 9.1730600000E-001
7.7418000000E+003 9.0328000000E-001
7.7421000000E+003 8.9026900000E-001
7.7424000000E+003 8.9206200000E-001
7.7427000000E+003 8.7939800000E-001
7.7430000000E+003 8.6704100000E-001
7.7433000000E+003 8.5196400000E-001
7.7436000000E+003 8.5953900000E-001
7.7439000000E+003 8.5234100000E-001
7.7442000000E+003 8.5346900000E-001
7.7445000000E+003 8.5185700000E-001
7.7448000000E+003 8.4505800000E-001
7.7451000000E+003 8.1112200000E-001
7.7454000000E+003 8.1307800000E-001
7.7457000000E+003 8.1330100000E-001
7.7460000000E+003 8.1150600000E-001
7.7463000000E+003 8.0718600000E-001
7.7466000000E+003 8.2164600000E-001
7.7469000000E+003 8.0993000000E-001
7.7472000000E+003 8.0849000000E-001
7.7475000000E+003 8.0133800000E-001
7.7478000000E+003 8.0793400000E-001
7.7481000000E+003 8.0836300000E-001
7.7484000000E+003 7.9739300000E-001
7.7487000000E+003 7.8875800000E-001
7.7490000000E+003 7.9589500000E-001
7.7493000000E+003 7.9255900000E-001
7.7496000000E+003 8.0122400000E-001
7.7499000000E+003 8.0643600000E-001
7.7502000000E+003 7.9144200000E-001
7.7505000000E+003 8.1059200000E-001
7.7508000000E+003 8.1554700000E-001
7.7511000000E+003 8.0557700000E-001
7.7514000000E+003 8.2321500000E-001
7.7517000000E+003 8.1028500000E-001
7.7520000000E+003 8.1815700000E-001
7.7523000000E+003 8.2647800000E-001
7.7526000000E+003 8.1768400000E-001
7.7529000000E+003 8.4463100000E-001
7.7532000000E+003 8.2934100000E-001
7.7535000000E+003 8.5531500000E-001
7.7538000000E+003 8.5262100000E-001
7.7541000000E+003 8.4997500000E-001
7.7544000000E+003 8.4999200000E-001
7.7547000000E+003 8.5613500000E-001
7.7550000000E+003 8.7571200000E-001
7.7553000000E+003 8.8303700000E-001
7.7556000000E+003 8.8327400000E-001
7.7559000000E+003 8.9278900000E-001
7.7562000000E+003 8.9119400000E-001
7.7565000000E+003 8.9025200000E-001
7.7568000000E+003 8.9610100000E-001
7.7571000000E+003 9.2876900000E-001
7.7574000000E+003 8.9869300000E-001
7.7577000000E+003 9.3840100000E-001
7.7580000000E+003 9.3693900000E-001
7.7583000000E+003 9.3925300000E-001
7.7586000000E+003 9.4987700000E-001
7.7589000000E+003 9.7060000000E-001
7.7592000000E+003 9.7269900000E-001
7.7595000000E+003 9.5926600000E-001
7.7598000000E+003 9.7475800000E-001
7.7601000000E+003 9.9450700000E-001
7.7604000000E+003 9.8679200000E-001
7.7607000000E+003 9.9637600000E-001
7.7610000000E+003 9.9448500000E-001
7.7613000000E+003 9.9349600000E-001
7.7616000000E+003 1.0290070000E+000
7.7619000000E+003 1.0082360000E+000
7.7622000000E+003 1.0290480000E+000
7.7625000000E+003 1.0224410000E+000
7.7628000000E+003 1.0460270000E+000
7.7631000000E+003 1.0316740000E+000
7.7634000000E+003 1.0536490000E+000
7.7637000000E+003 1.0357630000E+000
7.7640000000E+003 1.0374740000E+000
7.7643000000E+003 1.0479700000E+000
7.7646000000E+003 1.0228440000E+000
7.7649000000E+003 1.0428670000E+000
7.7652000000E+003 1.0607740000E+000
7.7655000000E+003 1.0639820000E+000
7.7658000000E+003 1.0711730000E+000
7.7661000000E+003 1.0445540000E+000
7.7664000000E+003 1.0516270000E+000
7.7667000000E+003 1.0749230000E+000
7.7670000000E+003 1.0758750000E+000
7.7673000000E+003 1.0531360000E+000
7.7676000000E+003 1.0559800000E+000
7.7679000000E+003 1.0566810000E+000
7.7682000000E+003 1.0618070000E+000
7.7685000000E+003 1.0578580000E+000
7.7688000000E+003 1.0816530000E+000
7.7691000000E+003 1.0538710000E+000
7.7694000000E+003 1.0784860000E+000
7.7697000000E+003 1.0690980000E+000
7.7700000000E+003 1.0656370000E+000
7.7703000000E+003 1.0669170000E+000
7.7706000000E+003 1.0644230000E+000
7.7709000000E+003 1.0466280000E+000
7.7712000000E+003 1.0472260000E+000
7.7715000000E+003 1.0710470000E+000
7.7718000000E+003 1.0633170000E+000
7.7721000000E+003 1.0470670000E+000
7.7724000000E+003 1.0643990000E+000
7.7727000000E+003 1.0693280000E+000
7.7730000000E+003 1.0657710000E+000
7.7733000000E+003 1.0855940000E+000
7.7736000000E+003 1.0584100000E+000
7.7739000000E+003 1.0342810000E+000
7.7742000000E+003 1.0797130000E+000
7.7745000000E+003 1.0510800000E+000
7.7748000000E+003 1.0669390000E+000
7.7751000000E+003 1.0437220000E+000
7.7754000000E+003 1.0438700000E+000
7.7757000000E+003 1.0551690000E+000
7.7760000000E+003 1.0418180000E+000
7.7763000000E+003 1.0627320000E+000
7.7766000000E+003 1.0586930000E+000
7.7769000000E+003 1.0461400000E+000
7.7772000000E+003 1.0604780000E+000
7.7775000000E+003 1.0422920000E+000
7.7778000000E+003 1.0681820000E+000
7.7781000000E+003 1.0490640000E+000
7.7784000000E+003 1.0471910000E+000
7.7787000000E+003 1.0462610000E+000
7.7790000000E+003 1.0537860000E+000
7.7793000000E+003 1.0401120000E+000
7.7796000000E+003 1.0618420000E+000
7.7799000000E+003 1.0595490000E+000
7.7802000000E+003 1.0554840000E+000
7.7805000000E+003 1.0299450000E+000
7.7808000000E+003 1.0479990000E+000
7.7811000000E+003 1.0402910000E+000
7.7814000000E+003 1.0602710000E+000
7.7817000000E+003 1.0199390000E+000
7.7820000000E+003 1.0576500000E+000
7.7823000000E+003 1.0504850000E+000
7.7826000000E+003 1.0542060000E+000
7.7829000000E+003 1.0488840000E+000
7.7832000000E+003 1.0401740000E+000
7.7835000000E+003 1.0534550000E+000
7.7838000000E+003 1.0454490000E+000
7.7841000000E+003 1.0437770000E+000
7.7844000000E+003 1.0643510000E+000
7.7847000000E+003 1.0362300000E+000
-------------- next part --------------
# XDI/1.0 Athena/0.9.26
# Element.edge: K
# Element.symbol: Co
# Column.1: energy eV
# Column.2: xmu
# Column.3: bkg
# Column.4: pre_edge
# Column.5: post_edge
# Column.6: der
# Column.7: sec
# Column.8: i0
# Column.9: chi(e)
# Athena.e0: 7719.6
# Athena.eshift: 0
# Athena.rbkg: 1.0
# Athena.importance: 1
# Athena.standard: None
# Athena.bkg_kweight: 2
# Athena.edge_step: 0.8101498
# Athena.fixed_step: no
# Athena.pre_edge_range: -24.600 -14.600
# Athena.pre_edge_line: 0.051531 - 6.66718e-006 * E
# Athena.normalization_range: 15.000 65.100
# Athena.post_edge_polynomial: -30.206407 + 0.00401791 * E + 0 * E^2 + 0 * E^3
# Athena.spline_range_energy: 0.000 65.112
# Athena.clamps: 0 24
# Athena.spline_range_k: 0.000 4.134
# Athena.kweight: 2
# Athena.window: hanning
# Athena.phase_correction: no
# Athena.k_range: 2.134 3.000
# Athena.dk: 1
# Athena.r_range: 1 3
# Athena.dr: 0.0
# Athena.window: hanning
# Athena.plot_multiplier: 1
# Athena.y_offset: 0
# ///
#------------------------
# e xmu bkg pre_edge post_edge der sec i0
7695.0000 0.16270000E-02 0.0000000 0.22738487E-03 0.71137270 -0.54233333E-02 0.90388889E-02 1.0000000
7695.3000 0.0000000 0.0000000 0.22538471E-03 0.71257807 -0.27116667E-02 0.90500000E-02 1.0000000
7695.6000 0.0000000 0.0000000 0.22338456E-03 0.71378344 0.66666667E-05 0.45194444E-02 1.0000000
7695.9000 0.40000000E-05 0.0000000 0.22138441E-03 0.71498881 0.0000000 -0.22222222E-04 1.0000000
7696.2000 0.0000000 0.0000000 0.21938426E-03 0.71619418 -0.66666667E-05 0.0000000 1.0000000
7696.5000 0.0000000 0.0000000 0.21738410E-03 0.71739956 0.0000000 0.11111111E-04 1.0000000
7696.8000 0.0000000 0.0000000 0.21538395E-03 0.71860493 0.0000000 0.52777778E-03 1.0000000
7697.1000 0.0000000 0.0000000 0.21338380E-03 0.71981030 0.31666667E-03 0.61944444E-03 1.0000000
7697.4000 0.19000000E-03 0.0000000 0.21138365E-03 0.72101567 0.37166667E-03 -0.10555556E-02 1.0000000
7697.7000 0.22300000E-03 0.0000000 0.20938349E-03 0.72222104 -0.31666667E-03 -0.97500000E-03 1.0000000
7698.0000 0.0000000 0.0000000 0.20738334E-03 0.72342641 -0.21333333E-03 0.52777778E-03 1.0000000
7698.3000 0.95000000E-04 0.0000000 0.20538319E-03 0.72463179 0.0000000 0.91666667E-04 1.0000000
7698.6000 0.0000000 0.0000000 0.20338303E-03 0.72583716 -0.15833333E-03 0.14250000E-02 1.0000000
7698.9000 0.0000000 0.0000000 0.20138288E-03 0.72704253 0.85500000E-03 0.26388889E-03 1.0000000
7699.2000 0.51300000E-03 0.0000000 0.19938273E-03 0.72824790 0.0000000 -0.28500000E-02 1.0000000
7699.5000 0.0000000 0.0000000 0.19738258E-03 0.72945327 -0.85500000E-03 0.0000000 1.0000000
7699.8000 0.0000000 0.0000000 0.19538242E-03 0.73065864 0.0000000 0.14250000E-02 1.0000000
7700.1000 0.0000000 0.0000000 0.19338227E-03 0.73186401 0.0000000 0.19027778E-02 1.0000000
7700.4000 0.0000000 0.0000000 0.19138212E-03 0.73306939 0.11416667E-02 0.0000000 1.0000000
7700.7000 0.68500000E-03 0.0000000 0.18938196E-03 0.73427476 0.0000000 -0.38055556E-02 1.0000000
7701.0000 0.0000000 0.0000000 0.18738181E-03 0.73548013 -0.11416667E-02 0.33666667E-02 1.0000000
7701.3000 0.0000000 0.0000000 0.18538166E-03 0.73668550 0.20200000E-02 0.53805556E-02 1.0000000
7701.6000 0.12120000E-02 0.0000000 0.18338151E-03 0.73789087 0.20866667E-02 -0.67333333E-02 1.0000000
7701.9000 0.12520000E-02 0.0000000 0.18138135E-03 0.73909624 -0.20200000E-02 -0.69555556E-02 1.0000000
7702.2000 0.0000000 0.0000000 0.17938120E-03 0.74030162 -0.20866667E-02 0.33666667E-02 1.0000000
7702.5000 0.0000000 0.0000000 0.17738105E-03 0.74150699 0.0000000 0.51027778E-02 1.0000000
7702.8000 0.0000000 0.0000000 0.17538089E-03 0.74271236 0.97500000E-03 0.0000000 1.0000000
7703.1000 0.58500000E-03 0.0000000 0.17338074E-03 0.74391773 0.0000000 -0.26305556E-02 1.0000000
7703.4000 0.0000000 0.0000000 0.17138059E-03 0.74512310 -0.60333333E-03 0.0000000 1.0000000
7703.7000 0.22300000E-03 0.0000000 0.16938044E-03 0.74632847 0.0000000 0.38611111E-03 1.0000000
7704.0000 0.0000000 0.0000000 0.16738028E-03 0.74753384 -0.37166667E-03 0.0000000 1.0000000
7704.3000 0.0000000 0.0000000 0.16538013E-03 0.74873922 0.0000000 0.61944444E-03 1.0000000
7704.6000 0.0000000 0.0000000 0.16337998E-03 0.74994459 0.0000000 0.0000000 1.0000000
7704.9000 0.0000000 0.0000000 0.16137983E-03 0.75114996 0.0000000 0.29194444E-02 1.0000000
7705.2000 0.0000000 0.0000000 0.15937967E-03 0.75235533 0.17516667E-02 0.41805556E-02 1.0000000
7705.5000 0.10510000E-02 0.0000000 0.15737952E-03 0.75356070 0.25083333E-02 0.21250000E-02 1.0000000
7705.8000 0.15050000E-02 0.0000000 0.15537937E-03 0.75476607 0.30266667E-02 0.62222222E-03 1.0000000
7706.1000 0.28670000E-02 0.0000000 0.15337921E-03 0.75597145 0.28816667E-02 0.36222222E-02 1.0000000
7706.4000 0.32340000E-02 0.0000000 0.15137906E-03 0.75717682 0.52000000E-02 0.10544444E-01 1.0000000
7706.7000 0.59870000E-02 0.0000000 0.14937891E-03 0.75838219 0.92083333E-02 0.53694444E-02 1.0000000
7707.0000 0.87590000E-02 0.0000000 0.14737876E-03 0.75958756 0.84216667E-02 0.20166667E-01 1.0000000
7707.3000 0.11040000E-01 0.0000000 0.14537860E-03 0.76079293 0.21308333E-01 0.22511111E-01 1.0000000
7707.6000 0.21544000E-01 0.0000000 0.14337845E-03 0.76199830 0.21928333E-01 -0.19538889E-01 1.0000000
7707.9000 0.24197000E-01 0.0000000 0.14137830E-03 0.76320367 0.95850000E-02 -0.33052778E-01 1.0000000
7708.2000 0.27295000E-01 0.0000000 0.13937814E-03 0.76440905 0.20966667E-02 -0.18936111E-01 1.0000000
7708.5000 0.25455000E-01 0.0000000 0.13737799E-03 0.76561442 -0.17766667E-02 0.57638889E-02 1.0000000
7708.8000 0.26229000E-01 0.0000000 0.13537784E-03 0.76681979 0.55550000E-02 -0.91694444E-02 1.0000000
7709.1000 0.28788000E-01 0.0000000 0.13337769E-03 0.76802516 -0.72783333E-02 -0.29516667E-01 1.0000000
7709.4000 0.21862000E-01 0.0000000 0.13137753E-03 0.76923053 -0.12155000E-01 -0.46777778E-02 1.0000000
7709.7000 0.21495000E-01 0.0000000 0.12937738E-03 0.77043590 -0.10085000E-01 -0.34861111E-02 1.0000000
7710.0000 0.15811000E-01 0.0000000 0.12737723E-03 0.77164127 -0.14246667E-01 0.72805556E-02 1.0000000
7710.3000 0.12947000E-01 0.0000000 0.12537708E-03 0.77284665 -0.57166667E-02 0.29155556E-01 1.0000000
7710.6000 0.12381000E-01 0.0000000 0.12337692E-03 0.77405202 0.32466667E-02 0.16863889E-01 1.0000000
7710.9000 0.14895000E-01 0.0000000 0.12137677E-03 0.77525739 0.44016667E-02 0.12444444E-02 1.0000000
7711.2000 0.15022000E-01 0.0000000 0.11937662E-03 0.77646276 0.39933333E-02 0.69111111E-02 1.0000000
7711.5000 0.17291000E-01 0.0000000 0.11737646E-03 0.77766813 0.85483333E-02 -0.53500000E-02 1.0000000
7711.8000 0.20151000E-01 0.0000000 0.11537631E-03 0.77887350 0.78333333E-03 -0.12677778E-01 1.0000000
7712.1000 0.17761000E-01 0.0000000 0.11337616E-03 0.78007888 0.94166667E-03 0.13000000E-01 1.0000000
7712.4000 0.20716000E-01 0.0000000 0.11137601E-03 0.78128425 0.85833333E-02 0.26319444E-01 1.0000000
7712.7000 0.22911000E-01 0.0000000 0.10937585E-03 0.78248962 0.16733333E-01 0.16955556E-01 1.0000000
7713.0000 0.30756000E-01 0.0000000 0.10737570E-03 0.78369499 0.18756667E-01 -0.59916667E-02 1.0000000
7713.3000 0.34165000E-01 0.0000000 0.10537555E-03 0.78490036 0.13138333E-01 0.14427778E-01 1.0000000
7713.6000 0.38639000E-01 0.0000000 0.10337539E-03 0.78610573 0.27413333E-01 0.50511111E-01 1.0000000
7713.9000 0.50613000E-01 0.0000000 0.10137524E-03 0.78731110 0.43445000E-01 0.54769444E-01 1.0000000
7714.2000 0.64706000E-01 0.0000000 0.99375089E-04 0.78851648 0.60275000E-01 0.25261111E-01 1.0000000
7714.5000 0.86778000E-01 0.0000000 0.97374936E-04 0.78972185 0.58601667E-01 0.50638889E-02 1.0000000
7714.8000 0.99867000E-01 0.0000000 0.95374783E-04 0.79092722 0.63313333E-01 0.15816667E-01 1.0000000
7715.1000 0.12476600 0.0000000 0.93374631E-04 0.79213259 0.68091667E-01 0.34725000E-01 1.0000000
7715.4000 0.14072200 0.0000000 0.91374478E-04 0.79333796 0.84148333E-01 0.57563889E-01 1.0000000
7715.7000 0.17525500 0.0000000 0.89374325E-04 0.79454333 0.10263000 0.18716667E-01 1.0000000
7716.0000 0.20230000 0.0000000 0.87374172E-04 0.79574871 0.95378333E-01 0.70336111E-01 1.0000000
7716.3000 0.23248200 0.0000000 0.85374020E-04 0.79695408 0.14483167 0.12311944 1.0000000
7716.6000 0.28919900 0.0000000 0.83373867E-04 0.79815945 0.16925000 -0.28613889E-01 1.0000000
7716.9000 0.33403200 0.0000000 0.81373714E-04 0.79936482 0.12766333 -0.40833333E-01 1.0000000
7717.2000 0.36579700 0.0000000 0.79373561E-04 0.80057019 0.14475000 0.58038889E-01 1.0000000
7717.5000 0.42088200 0.0000000 0.77373408E-04 0.80177556 0.16248667 0.48222222E-02 1.0000000
7717.8000 0.46328900 0.0000000 0.75373256E-04 0.80298093 0.14764333 -0.94361111E-02 1.0000000
7718.1000 0.50946800 0.0000000 0.73373103E-04 0.80418631 0.15682500 0.42541667E-01 1.0000000
7718.4000 0.55738400 0.0000000 0.71372950E-04 0.80539168 0.17316833 0.76422222E-01 1.0000000
7718.7000 0.61336900 0.0000000 0.69372797E-04 0.80659705 0.20267833 0.68383333E-01 1.0000000
7719.0000 0.67899100 0.0000000 0.67372644E-04 0.80780242 0.21419833 0.10034444 1.0000000
7719.3000 0.74188800 0.0000000 0.65372492E-04 0.80900779 0.26288500 0.13880556 1.0000000
7719.6000 0.83672200 0.0000000 0.63372339E-04 0.81021316 0.29748167 0.47969444E-01 1.0000000
7719.9000 0.92037700 0.0000000 0.61372186E-04 0.81141854 0.29166667 -0.92708333E-01 1.0000000
7720.2000 1.0117220 0.0000000 0.59372033E-04 0.81262391 0.24185667 -0.14996111 1.0000000
7720.5000 1.0654910 0.0000000 0.57371880E-04 0.81382928 0.20169000 0.22252778E-01 1.0000000
7720.8000 1.1327360 0.0000000 0.55371728E-04 0.81503465 0.25520833 -0.30222222E-02 1.0000000
7721.1000 1.2186160 0.0000000 0.53371575E-04 0.81624002 0.19987667 -0.18130278 1.0000000
7721.4000 1.2526620 0.0000000 0.51371422E-04 0.81744539 0.14642667 -0.89477778E-01 1.0000000
7721.7000 1.3064720 0.0000000 0.49371269E-04 0.81865076 0.14619000 -0.65336111E-01 1.0000000
7722.0000 1.3403760 0.0000000 0.47371117E-04 0.81985614 0.10722500 -0.11388056 1.0000000
7722.3000 1.3708070 0.0000000 0.45370964E-04 0.82106151 0.77861667E-01 -0.98094444E-01 1.0000000
7722.6000 1.3870930 0.0000000 0.43370811E-04 0.82226688 0.48368333E-01 -0.10009444 1.0000000
7722.9000 1.3998280 0.0000000 0.41370658E-04 0.82347225 0.17805000E-01 0.20141667E-01 1.0000000
7723.2000 1.3977760 0.0000000 0.39370505E-04 0.82467762 0.60453333E-01 0.60527778E-02 1.0000000
7723.5000 1.4361000 0.0000000 0.37370353E-04 0.82588299 0.21436667E-01 -0.21771944 1.0000000
7723.8000 1.4106380 0.0000000 0.35370200E-04 0.82708836 -0.70178333E-01 -0.10815278 1.0000000
7724.1000 1.3939930 0.0000000 0.33370047E-04 0.82829374 -0.43455000E-01 0.27377778E-01 1.0000000
7724.4000 1.3845650 0.0000000 0.31369894E-04 0.82949911 -0.53751667E-01 -0.29538889E-01 1.0000000
7724.7000 1.3617420 0.0000000 0.29369741E-04 0.83070448 -0.61178333E-01 -0.81608333E-01 1.0000000
7725.0000 1.3478580 0.0000000 0.27369589E-04 0.83190985 -0.10271667 0.39188889E-01 1.0000000
7725.3000 1.3001120 0.0000000 0.25369436E-04 0.83311522 -0.37665000E-01 0.14775833 1.0000000
7725.6000 1.3252590 0.0000000 0.23369283E-04 0.83432059 -0.14061667E-01 -0.58080556E-01 1.0000000
7725.9000 1.2916750 0.0000000 0.21369130E-04 0.83552597 -0.72513333E-01 -0.80661111E-01 1.0000000
7726.2000 1.2817510 0.0000000 0.19368978E-04 0.83673134 -0.62458333E-01 0.43063889E-01 1.0000000
7726.5000 1.2542000 0.0000000 0.17368825E-04 0.83793671 -0.46675000E-01 0.61550000E-01 1.0000000
7726.8000 1.2537460 0.0000000 0.15368672E-04 0.83914208 -0.25528333E-01 -0.89452778E-01 1.0000000
7727.1000 1.2388830 0.0000000 0.13368519E-04 0.84034745 -0.10034667 -0.11085278 1.0000000
7727.4000 1.1935380 0.0000000 0.11368366E-04 0.84155282 -0.92040000E-01 0.22386111E-01 1.0000000
7727.7000 1.1836590 0.0000000 0.93682136E-05 0.84275819 -0.86915000E-01 -0.15333333E-02 1.0000000
7728.0000 1.1413890 0.0000000 0.73680608E-05 0.84396357 -0.92960000E-01 0.90375000E-01 1.0000000
7728.3000 1.1278830 0.0000000 0.53679080E-05 0.84516894 -0.32690000E-01 0.10846389 1.0000000
7728.6000 1.1217750 0.0000000 0.33677552E-05 0.84637431 -0.27881667E-01 -0.13633333E-01 1.0000000
7728.9000 1.1111540 0.0000000 0.13676024E-05 0.84757968 -0.40870000E-01 -0.25833333E-01 1.0000000
7729.2000 1.0972530 0.0000000 -0.63255036E-06 0.84878505 -0.43381667E-01 0.18963889E-01 1.0000000
7729.5000 1.0851250 0.0000000 -0.26327032E-05 0.84999042 -0.29491667E-01 0.25380556E-01 1.0000000
7729.8000 1.0795580 0.0000000 -0.46328559E-05 0.85119580 -0.28153333E-01 -0.33777778E-02 1.0000000
7730.1000 1.0682330 0.0000000 -0.66330087E-05 0.85240117 -0.31518333E-01 0.27111111E-01 1.0000000
7730.4000 1.0606470 0.0000000 -0.86331615E-05 0.85360654 -0.11886667E-01 0.57527778E-01 1.0000000
7730.7000 1.0611010 0.0000000 -0.10633314E-04 0.85481191 0.29983333E-02 -0.13080556E-01 1.0000000
7731.0000 1.0624460 0.0000000 -0.12633467E-04 0.85601728 -0.19735000E-01 -0.59830556E-01 1.0000000
7731.3000 1.0492600 0.0000000 -0.14633620E-04 0.85722265 -0.32900000E-01 0.34291667E-01 1.0000000
7731.6000 1.0427060 0.0000000 -0.16633773E-04 0.85842802 0.84000000E-03 0.28230556E-01 1.0000000
7731.9000 1.0497640 0.0000000 -0.18633925E-04 0.85963340 -0.15961667E-01 -0.70000000E-03 1.0000000
7732.2000 1.0331290 0.0000000 -0.20634078E-04 0.86083877 0.42000000E-03 0.96244444E-01 1.0000000
7732.5000 1.0500160 0.0000000 -0.22634231E-04 0.86204414 0.41785000E-01 -0.86638889E-02 1.0000000
7732.8000 1.0582000 0.0000000 -0.24634384E-04 0.86324951 -0.47783333E-02 -0.98413889E-01 1.0000000
7733.1000 1.0471490 0.0000000 -0.26634537E-04 0.86445488 -0.17263333E-01 -0.29536111E-01 1.0000000
7733.4000 1.0478420 0.0000000 -0.28634689E-04 0.86566025 -0.22500000E-01 0.16525000E-01 1.0000000
7733.7000 1.0336490 0.0000000 -0.30634842E-04 0.86686563 -0.73483333E-02 0.83200000E-01 1.0000000
7734.0000 1.0434330 0.0000000 -0.32634995E-04 0.86807100 0.27420000E-01 -0.13944444E-02 1.0000000
7734.3000 1.0501010 0.0000000 -0.34635148E-04 0.86927637 -0.81850000E-02 -0.48194444E-02 1.0000000
7734.6000 1.0385220 0.0000000 -0.36635301E-04 0.87048174 0.24528333E-01 0.56469444E-01 1.0000000
7734.9000 1.0648180 0.0000000 -0.38635453E-04 0.87168711 0.25696667E-01 -0.97152778E-01 1.0000000
7735.2000 1.0539400 0.0000000 -0.40635606E-04 0.87289248 -0.33763333E-01 -0.58688889E-01 1.0000000
7735.5000 1.0445600 0.0000000 -0.42635759E-04 0.87409785 -0.95166667E-02 0.56394444E-01 1.0000000
7735.8000 1.0482300 0.0000000 -0.44635912E-04 0.87530323 0.73333333E-04 0.13555556E-01 1.0000000
7736.1000 1.0446040 0.0000000 -0.46636064E-04 0.87650860 -0.13833333E-02 -0.11619444E-01 1.0000000
7736.4000 1.0474000 0.0000000 -0.48636217E-04 0.87771397 -0.68983333E-02 -0.47602778E-01 1.0000000
7736.7000 1.0404650 0.0000000 -0.50636370E-04 0.87891934 -0.29945000E-01 -0.33063889E-01 1.0000000
7737.0000 1.0294330 0.0000000 -0.52636523E-04 0.88012471 -0.26736667E-01 -0.26241667E-01 1.0000000
7737.3000 1.0244230 0.0000000 -0.54636676E-04 0.88133008 -0.45690000E-01 0.17483333E-01 1.0000000
7737.6000 1.0020190 0.0000000 -0.56636828E-04 0.88253545 -0.16246667E-01 0.80602778E-01 1.0000000
7737.9000 1.0146750 0.0000000 -0.58636981E-04 0.88374083 0.26716667E-02 -0.52913889E-01 1.0000000
7738.2000 1.0036220 0.0000000 -0.60637134E-04 0.88494620 -0.47995000E-01 -0.70033333E-01 1.0000000
7738.5000 0.98587800 0.0000000 -0.62637287E-04 0.88615157 -0.39348333E-01 0.53877778E-01 1.0000000
7738.8000 0.98001300 0.0000000 -0.64637440E-04 0.88735694 -0.15668333E-01 0.16961111E-01 1.0000000
7739.1000 0.97647700 0.0000000 -0.66637592E-04 0.88856231 -0.29171667E-01 0.12819444E-01 1.0000000
7739.4000 0.96251000 0.0000000 -0.68637745E-04 0.88976768 -0.79766667E-02 0.54833333E-02 1.0000000
7739.7000 0.97169100 0.0000000 -0.70637898E-04 0.89097306 -0.25881667E-01 -0.61211111E-01 1.0000000
7740.0000 0.94698100 0.0000000 -0.72638051E-04 0.89217843 -0.44703333E-01 0.11652778E-01 1.0000000
7740.3000 0.94486900 0.0000000 -0.74638204E-04 0.89338380 -0.18890000E-01 0.40916667E-02 1.0000000
7740.6000 0.93564700 0.0000000 -0.76638356E-04 0.89458917 -0.42248333E-01 -0.40200000E-01 1.0000000
7740.9000 0.91952000 0.0000000 -0.78638509E-04 0.89579454 -0.43010000E-01 0.64263889E-01 1.0000000
7741.2000 0.90984100 0.0000000 -0.80638662E-04 0.89699991 -0.36900000E-02 0.53458333E-01 1.0000000
7741.5000 0.91730600 0.0000000 -0.82638815E-04 0.89820528 -0.10935000E-01 -0.68952778E-01 1.0000000
7741.8000 0.90328000 0.0000000 -0.84638967E-04 0.89941066 -0.45061667E-01 -0.12936111E-01 1.0000000
7742.1000 0.89026900 0.0000000 -0.86639120E-04 0.90061603 -0.18696667E-01 0.44905556E-01 1.0000000
7742.4000 0.89206200 0.0000000 -0.88639273E-04 0.90182140 -0.18118333E-01 -0.38341667E-01 1.0000000
7742.7000 0.87939800 0.0000000 -0.90639426E-04 0.90302677 -0.41701667E-01 -0.46008333E-01 1.0000000
7743.0000 0.86704100 0.0000000 -0.92639579E-04 0.90423214 -0.45723333E-01 0.48663889E-01 1.0000000
7743.3000 0.85196400 0.0000000 -0.94639731E-04 0.90543751 -0.12503333E-01 0.77252778E-01 1.0000000
7743.6000 0.85953900 0.0000000 -0.96639884E-04 0.90664289 0.62833333E-03 0.39777778E-02 1.0000000
7743.9000 0.85234100 0.0000000 -0.98640037E-04 0.90784826 -0.10116667E-01 -0.23916667E-02 1.0000000
7744.2000 0.85346900 0.0000000 -0.10064019E-03 0.90905363 -0.80666667E-03 -0.65027778E-02 1.0000000
7744.5000 0.85185700 0.0000000 -0.10264034E-03 0.91025900 -0.14018333E-01 -0.11180833 1.0000000
7744.8000 0.84505800 0.0000000 -0.10464050E-03 0.91146437 -0.67891667E-01 -0.65469444E-01 1.0000000
7745.1000 0.81112200 0.0000000 -0.10664065E-03 0.91266974 -0.53300000E-01 0.11920556 1.0000000
7745.4000 0.81307800 0.0000000 -0.10864080E-03 0.91387511 0.36316667E-02 0.84466667E-01 1.0000000
7745.7000 0.81330100 0.0000000 -0.11064095E-03 0.91508049 -0.26200000E-02 -0.23038889E-01 1.0000000
7746.0000 0.81150600 0.0000000 -0.11264111E-03 0.91628586 -0.10191667E-01 0.32533333E-01 1.0000000
7746.3000 0.80718600 0.0000000 -0.11464126E-03 0.91749123 0.16900000E-01 0.24608333E-01 1.0000000
7746.6000 0.82164600 0.0000000 -0.11664141E-03 0.91869660 0.45733333E-02 -0.64711111E-01 1.0000000
7746.9000 0.80993000 0.0000000 -0.11864156E-03 0.91990197 -0.21926667E-01 -0.31488889E-01 1.0000000
7747.2000 0.80849000 0.0000000 -0.12064172E-03 0.92110734 -0.14320000E-01 0.35000000E-01 1.0000000
7747.5000 0.80133800 0.0000000 -0.12264187E-03 0.92231272 -0.92666667E-03 0.43380556E-01 1.0000000
7747.8000 0.80793400 0.0000000 -0.12464202E-03 0.92351809 0.11708333E-01 -0.27736111E-01 1.0000000
7748.1000 0.80836300 0.0000000 -0.12664218E-03 0.92472346 -0.17568333E-01 -0.73972222E-01 1.0000000
7748.4000 0.79739300 0.0000000 -0.12864233E-03 0.92592883 -0.32675000E-01 0.25119444E-01 1.0000000
7748.7000 0.78875800 0.0000000 -0.13064248E-03 0.92713420 -0.24966667E-02 0.65016667E-01 1.0000000
7749.0000 0.79589500 0.0000000 -0.13264263E-03 0.92833957 0.63350000E-02 0.18963889E-01 1.0000000
7749.3000 0.79255900 0.0000000 -0.13464279E-03 0.92954494 0.88816667E-02 0.27988889E-01 1.0000000
7749.6000 0.80122400 0.0000000 -0.13664294E-03 0.93075032 0.23128333E-01 -0.41975000E-01 1.0000000
7749.9000 0.80643600 0.0000000 -0.13864309E-03 0.93195569 -0.16303333E-01 -0.27002778E-01 1.0000000
7750.2000 0.79144200 0.0000000 -0.14064325E-03 0.93316106 0.69266667E-02 0.94130556E-01 1.0000000
7750.5000 0.81059200 0.0000000 -0.14264340E-03 0.93436643 0.40175000E-01 -0.25475000E-01 1.0000000
7750.8000 0.81554700 0.0000000 -0.14464355E-03 0.93557180 -0.83583333E-02 -0.45658333E-01 1.0000000
7751.1000 0.80557700 0.0000000 -0.14664370E-03 0.93677717 0.12780000E-01 0.27008333E-01 1.0000000
7751.4000 0.82321500 0.0000000 -0.14864386E-03 0.93798255 0.78466667E-02 -0.35350000E-01 1.0000000
7751.7000 0.81028500 0.0000000 -0.15064401E-03 0.93918792 -0.84300000E-02 0.31902778E-01 1.0000000
7752.0000 0.81815700 0.0000000 -0.15264416E-03 0.94039329 0.26988333E-01 0.12736111E-01 1.0000000
7752.3000 0.82647800 0.0000000 -0.15464432E-03 0.94159866 -0.78833333E-03 0.54444444E-02 1.0000000
7752.6000 0.81768400 0.0000000 -0.15664447E-03 0.94280403 0.30255000E-01 0.33694444E-01 1.0000000
7752.9000 0.84463100 0.0000000 -0.15864462E-03 0.94400940 0.19428333E-01 -0.20747222E-01 1.0000000
7753.2000 0.82934100 0.0000000 -0.16064477E-03 0.94521477 0.17806667E-01 0.32286111E-01 1.0000000
7753.5000 0.85531500 0.0000000 -0.16264493E-03 0.94642015 0.38800000E-01 -0.44511111E-01 1.0000000
7753.8000 0.85262100 0.0000000 -0.16464508E-03 0.94762552 -0.89000000E-02 -0.71969444E-01 1.0000000
7754.1000 0.84997500 0.0000000 -0.16664523E-03 0.94883089 -0.43816667E-02 0.31944444E-01 1.0000000
7754.4000 0.84999200 0.0000000 -0.16864538E-03 0.95003626 0.10266667E-01 0.78747222E-01 1.0000000
7754.7000 0.85613500 0.0000000 -0.17064554E-03 0.95124163 0.42866667E-01 0.57616667E-01 1.0000000
7755.0000 0.87571200 0.0000000 -0.17264569E-03 0.95244700 0.44836667E-01 -0.50438889E-01 1.0000000
7755.3000 0.88303700 0.0000000 -0.17464584E-03 0.95365237 0.12603333E-01 -0.47638889E-01 1.0000000
7755.6000 0.88327400 0.0000000 -0.17664600E-03 0.95485775 0.16253333E-01 0.99444444E-03 1.0000000
7755.9000 0.89278900 0.0000000 -0.17864615E-03 0.95606312 0.13200000E-01 -0.34136111E-01 1.0000000
7756.2000 0.89119400 0.0000000 -0.18064630E-03 0.95726849 -0.42283333E-02 -0.83694444E-02 1.0000000
7756.5000 0.89025200 0.0000000 -0.18264645E-03 0.95847386 0.81783333E-02 0.11403889 1.0000000
7756.8000 0.89610100 0.0000000 -0.18464661E-03 0.95967923 0.64195000E-01 -0.64305556E-02 1.0000000
7757.1000 0.92876900 0.0000000 -0.18664676E-03 0.96088460 0.43200000E-02 -0.80236111E-01 1.0000000
7757.4000 0.89869300 0.0000000 -0.18864691E-03 0.96208998 0.16053333E-01 0.99038889E-01 1.0000000
7757.7000 0.93840100 0.0000000 -0.19064707E-03 0.96329535 0.63743333E-01 -0.24388889E-01 1.0000000
7758.0000 0.93693900 0.0000000 -0.19264722E-03 0.96450072 0.14200000E-02 -0.70300000E-01 1.0000000
7758.3000 0.93925300 0.0000000 -0.19464737E-03 0.96570609 0.21563333E-01 0.84708333E-01 1.0000000
7758.6000 0.94987700 0.0000000 -0.19664752E-03 0.96691146 0.52245000E-01 0.27455556E-01 1.0000000
7758.9000 0.97060000 0.0000000 -0.19864768E-03 0.96811683 0.38036667E-01 -0.11855833 1.0000000
7759.2000 0.97269900 0.0000000 -0.20064783E-03 0.96932220 -0.18890000E-01 -0.57675000E-01 1.0000000
7759.5000 0.95926600 0.0000000 -0.20264798E-03 0.97052758 0.34316667E-02 0.12937500 1.0000000
7759.8000 0.97475800 0.0000000 -0.20464813E-03 0.97173295 0.58735000E-01 0.27708333E-01 1.0000000
7760.1000 0.99450700 0.0000000 -0.20664829E-03 0.97293832 0.20056667E-01 -0.92700000E-01 1.0000000
7760.4000 0.98679200 0.0000000 -0.20864844E-03 0.97414369 0.31150000E-02 -0.12058333E-01 1.0000000
7760.7000 0.99637600 0.0000000 -0.21064859E-03 0.97534906 0.12821667E-01 -0.13191667E-01 1.0000000
7761.0000 0.99448500 0.0000000 -0.21264875E-03 0.97655443 -0.48000000E-02 0.74525000E-01 1.0000000
7761.3000 0.99349600 0.0000000 -0.21464890E-03 0.97775981 0.57536667E-01 0.48944444E-01 1.0000000
7761.6000 1.0290070 0.0000000 -0.21664905E-03 0.97896518 0.24566667E-01 -0.95780556E-01 1.0000000
7761.9000 1.0082360 0.0000000 -0.21864920E-03 0.98017055 0.68333333E-04 -0.14861111E-02 1.0000000
7762.2000 1.0290480 0.0000000 -0.22064936E-03 0.98137592 0.23675000E-01 0.47050000E-01 1.0000000
7762.5000 1.0224410 0.0000000 -0.22264951E-03 0.98258129 0.28298333E-01 -0.13811111E-01 1.0000000
7762.8000 1.0460270 0.0000000 -0.22464966E-03 0.98378666 0.15388333E-01 -0.25991667E-01 1.0000000
7763.1000 1.0316740 0.0000000 -0.22664982E-03 0.98499203 0.12703333E-01 -0.14288889E-01 1.0000000
7763.4000 1.0536490 0.0000000 -0.22864997E-03 0.98619741 0.68150000E-02 -0.66102778E-01 1.0000000
7763.7000 1.0357630 0.0000000 -0.23065012E-03 0.98740278 -0.26958333E-01 0.22550000E-01 1.0000000
7764.0000 1.0374740 0.0000000 -0.23265027E-03 0.98860815 0.20345000E-01 0.42916667E-02 1.0000000
7764.3000 1.0479700 0.0000000 -0.23465043E-03 0.98981352 -0.24383333E-01 -0.48083333E-01 1.0000000
7764.6000 1.0228440 0.0000000 -0.23665058E-03 0.99101889 -0.85050000E-02 0.14600000 1.0000000
7764.9000 1.0428670 0.0000000 -0.23865073E-03 0.99222426 0.63216667E-01 0.72827778E-01 1.0000000
7765.2000 1.0607740 0.0000000 -0.24065088E-03 0.99342964 0.35191667E-01 -0.76475000E-01 1.0000000
7765.5000 1.0639820 0.0000000 -0.24265104E-03 0.99463501 0.17331667E-01 -0.11261944 1.0000000
7765.8000 1.0711730 0.0000000 -0.24465119E-03 0.99584038 -0.32380000E-01 -0.83180556E-01 1.0000000
7766.1000 1.0445540 0.0000000 -0.24665134E-03 0.99704575 -0.32576667E-01 0.13832500 1.0000000
7766.4000 1.0516270 0.0000000 -0.24865150E-03 0.99825112 0.50615000E-01 0.12165000 1.0000000
7766.7000 1.0749230 0.0000000 -0.25065165E-03 0.99945649 0.40413333E-01 -0.14487778 1.0000000
7767.0000 1.0758750 0.0000000 -0.25265180E-03 1.0006619 -0.36311667E-01 -0.12261944 1.0000000
7767.3000 1.0531360 0.0000000 -0.25465195E-03 1.0018672 -0.33158333E-01 0.70366667E-01 1.0000000
7767.6000 1.0559800 0.0000000 -0.25665211E-03 1.0030726 0.59083333E-02 0.71450000E-01 1.0000000
7767.9000 1.0566810 0.0000000 -0.25865226E-03 1.0042780 0.97116667E-02 -0.65777778E-02 1.0000000
7768.2000 1.0618070 0.0000000 -0.26065241E-03 1.0054834 0.19616667E-02 0.38941667E-01 1.0000000
7768.5000 1.0578580 0.0000000 -0.26265257E-03 1.0066887 0.33076667E-01 -0.14344444E-01 1.0000000
7768.8000 1.0816530 0.0000000 -0.26465272E-03 1.0078941 -0.66450000E-02 -0.63925000E-01 1.0000000
7769.1000 1.0538710 0.0000000 -0.26665287E-03 1.0090995 -0.52783333E-02 0.53372222E-01 1.0000000
7769.4000 1.0784860 0.0000000 -0.26865302E-03 1.0103048 0.25378333E-01 -0.26894444E-01 1.0000000
7769.7000 1.0690980 0.0000000 -0.27065318E-03 1.0115102 -0.21415000E-01 -0.48355556E-01 1.0000000
7770.0000 1.0656370 0.0000000 -0.27265333E-03 1.0127156 -0.36350000E-02 0.32319444E-01 1.0000000
7770.3000 1.0669170 0.0000000 -0.27465348E-03 1.0139210 -0.20233333E-02 -0.50300000E-01 1.0000000
7770.6000 1.0644230 0.0000000 -0.27665364E-03 1.0151263 -0.33815000E-01 -0.44397222E-01 1.0000000
7770.9000 1.0466280 0.0000000 -0.27865379E-03 1.0163317 -0.28661667E-01 0.12418889 1.0000000
7771.2000 1.0472260 0.0000000 -0.28065394E-03 1.0175371 0.40698333E-01 0.92466667E-01 1.0000000
7771.5000 1.0710470 0.0000000 -0.28265409E-03 1.0187424 0.26818333E-01 -0.13444167 1.0000000
7771.8000 1.0633170 0.0000000 -0.28465425E-03 1.0199478 -0.39966667E-01 -0.41691667E-01 1.0000000
7772.1000 1.0470670 0.0000000 -0.28665440E-03 1.0211532 0.18033333E-02 0.12844722 1.0000000
7772.4000 1.0643990 0.0000000 -0.28865455E-03 1.0223586 0.37101667E-01 0.80555556E-03 1.0000000
7772.7000 1.0693280 0.0000000 -0.29065470E-03 1.0235639 0.22866667E-02 -0.16652778E-01 1.0000000
7773.0000 1.0657710 0.0000000 -0.29265486E-03 1.0247693 0.27110000E-01 -0.24258333E-01 1.0000000
7773.3000 1.0855940 0.0000000 -0.29465501E-03 1.0259747 -0.12268333E-01 -0.18771944 1.0000000
7773.6000 1.0584100 0.0000000 -0.29665516E-03 1.0271800 -0.85521667E-01 0.79622222E-01 1.0000000
7773.9000 1.0342810 0.0000000 -0.29865532E-03 1.0283854 0.35505000E-01 0.18920000 1.0000000
7774.2000 1.0797130 0.0000000 -0.30065547E-03 1.0295908 0.27998333E-01 -0.94658333E-01 1.0000000
7774.5000 1.0510800 0.0000000 -0.30265562E-03 1.0307962 -0.21290000E-01 -0.67102778E-01 1.0000000
7774.8000 1.0669390 0.0000000 -0.30465577E-03 1.0320015 -0.12263333E-01 -0.28597222E-01 1.0000000
7775.1000 1.0437220 0.0000000 -0.30665593E-03 1.0332069 -0.38448333E-01 0.52236111E-01 1.0000000
7775.4000 1.0438700 0.0000000 -0.30865608E-03 1.0344123 0.19078333E-01 0.58380556E-01 1.0000000
7775.7000 1.0551690 0.0000000 -0.31065623E-03 1.0356176 -0.34200000E-02 -0.10788889E-01 1.0000000
7776.0000 1.0418180 0.0000000 -0.31265639E-03 1.0368230 0.12605000E-01 0.52575000E-01 1.0000000
7776.3000 1.0627320 0.0000000 -0.31465654E-03 1.0380284 0.28125000E-01 -0.67097222E-01 1.0000000
7776.6000 1.0586930 0.0000000 -0.31665669E-03 1.0392338 -0.27653333E-01 -0.41916667E-01 1.0000000
7776.9000 1.0461400 0.0000000 -0.31865684E-03 1.0404391 0.29750000E-02 0.35400000E-01 1.0000000
7777.2000 1.0604780 0.0000000 -0.32065700E-03 1.0416445 -0.64133333E-02 0.16441667E-01 1.0000000
7777.5000 1.0422920 0.0000000 -0.32265715E-03 1.0428499 0.12840000E-01 0.29500000E-01 1.0000000
7777.8000 1.0681820 0.0000000 -0.32465730E-03 1.0440552 0.11286667E-01 -0.79708333E-01 1.0000000
7778.1000 1.0490640 0.0000000 -0.32665745E-03 1.0452606 -0.34985000E-01 -0.26597222E-01 1.0000000
7778.4000 1.0471910 0.0000000 -0.32865761E-03 1.0464660 -0.46716667E-02 0.76627778E-01 1.0000000
7778.7000 1.0462610 0.0000000 -0.33065776E-03 1.0476714 0.10991667E-01 -0.92944444E-02 1.0000000
7779.0000 1.0537860 0.0000000 -0.33265791E-03 1.0488767 -0.10248333E-01 0.40583333E-02 1.0000000
7779.3000 1.0401120 0.0000000 -0.33465807E-03 1.0500821 0.13426667E-01 0.71072222E-01 1.0000000
7779.6000 1.0618420 0.0000000 -0.33665822E-03 1.0512875 0.32395000E-01 -0.40038889E-01 1.0000000
7779.9000 1.0595490 0.0000000 -0.33865837E-03 1.0524928 -0.10596667E-01 -0.13622500 1.0000000
7780.2000 1.0554840 0.0000000 -0.34065852E-03 1.0536982 -0.49340000E-01 -0.31305556E-02 1.0000000
7780.5000 1.0299450 0.0000000 -0.34265868E-03 1.0549036 -0.12475000E-01 0.11097222 1.0000000
7780.8000 1.0479990 0.0000000 -0.34465883E-03 1.0561090 0.17243333E-01 0.54880556E-01 1.0000000
7781.1000 1.0402910 0.0000000 -0.34665898E-03 1.0573143 0.20453333E-01 -0.85272222E-01 1.0000000
7781.4000 1.0602710 0.0000000 -0.34865914E-03 1.0585197 -0.33920000E-01 -0.41369444E-01 1.0000000
7781.7000 1.0199390 0.0000000 -0.35065929E-03 1.0597251 -0.43683333E-02 0.14138333 1.0000000
7782.0000 1.0576500 0.0000000 -0.35265944E-03 1.0609304 0.50910000E-01 -0.22861111E-02 1.0000000
7782.3000 1.0504850 0.0000000 -0.35465959E-03 1.0621358 -0.57400000E-02 -0.89297222E-01 1.0000000
7782.6000 1.0542060 0.0000000 -0.35665975E-03 1.0633412 -0.26683333E-02 -0.29411111E-01 1.0000000
7782.9000 1.0488840 0.0000000 -0.35865990E-03 1.0645466 -0.23386667E-01 0.17144444E-01 1.0000000
7783.2000 1.0401740 0.0000000 -0.36066005E-03 1.0657519 0.76183333E-02 0.53630556E-01 1.0000000
7783.5000 1.0534550 0.0000000 -0.36266020E-03 1.0669573 0.87916667E-02 -0.39580556E-01 1.0000000
7783.8000 1.0454490 0.0000000 -0.36466036E-03 1.0681627 -0.16130000E-01 0.37852778E-01 1.0000000
7784.1000 1.0437770 0.0000000 -0.36666051E-03 1.0693680 0.31503333E-01 0.59194444E-02 1.0000000
7784.4000 1.0643510 0.0000000 -0.36866066E-03 1.0705734 -0.12578333E-01 -0.20873333 1.0000000
7784.7000 1.0362300 0.0000000 -0.37066082E-03 1.0717788 -0.93736667E-01 -0.27052778 1.0000000
More information about the Ifeffit