The sample was a 1 cm square 2400Å thick
YBa2Cu3O6.6
(001) film grown epitaxially on a MgO (001) substrate by pulsed laser
deposition. The crystal was fully twinned in the ab plane and the
c-axis mosaic spread was about
FWHM. The x-ray polarization
was perpendicular to both the scattering plane and the c-axis of the
crystal. Simultaneous fluorescence measurements were used for the
absorption correction and were also used as an
reference to
facilitate accurate comparisons between different Bragg peaks. The DAFS
signals were measured across the Cu K-edge for the eight different (00L)
Bragg peaks listed in Table i. As illustrated in
Fig. 8, the measured contrast is very different for
different Bragg peaks. When the peak is weak, the oscillating DAFS terms,
and
, are more
important. The contrast is then larger and the DAFS signal can be measured
more easily
for these weak, high contrast reflections than it can for the strong, low
contrast reflections, or for the XAFS. This contrast enhancement is not
possible in conventional XAFS because the XAFS contrast is fixed by the
relative numbers of inequivalent absorbing atoms (with the same atomic
type). The DAFS intensities were corrected for absorption using the
measured fluorescence XAFS
signal. The measured
values were converted to absorption coefficient values by fitting the
measured XAFS to the smooth absorption coefficient calculated using the
values of McMaster [28]. The properly normalized fluorescence
XAFS data was then used to construct the absorption correction function,
. The measured DAFS data was also corrected for instrument
contributions.