Hi all, I've been pondering the McMaster correction recently. My understanding is that it is a correction because while chi(k) is defined relative to the embedded-atom background mu_o(E), we almost always extract it from our data by normalizing by the edge step. Since mu_o(E) drops gradually above the edge, the normalization procedure results in oscillations that are too small well above edge, which the McMaster correction then compensates for. It's also my understanding that this correction is the same whether the data is measured in absorption or fluorescence, because in this context mu_o(E) refers only to absorption due to the edge of interest, which is a characteristic of the atom in its local environment and is thus independent of measurement mode. So here's my question: why is existing software structured so that we have to put this factor in by hand? Feff, for instance, could simply define chi(k) consistently with the usual procedure, so that it was normalized by the edge step rather than mu_o(E). A card could be set to turn that off if a user desired. Alternatively, a correction could be done to the experimental data by Athena, or automatically within the fitting procedure by Ifeffit. Of course, having more than one of those options could cause trouble, just as the ability to put sigma2 into a feff calculation and in to Ifeffit sometimes does now. But wouldn't it make sense to have it available (perhaps even the default) at one of those stages? --Scott Calvin Sarah Lawrence College
Scott, Is this a discussion topic or a feature request? :) B On Thursday, June 16, 2011 08:28:18 pm Scott Calvin wrote:
Hi all,
I've been pondering the McMaster correction recently.
My understanding is that it is a correction because while chi(k) is defined relative to the embedded-atom background mu_o(E), we almost always extract it from our data by normalizing by the edge step. Since mu_o(E) drops gradually above the edge, the normalization procedure results in oscillations that are too small well above edge, which the McMaster correction then compensates for. It's also my understanding that this correction is the same whether the data is measured in absorption or fluorescence, because in this context mu_o(E) refers only to absorption due to the edge of interest, which is a characteristic of the atom in its local environment and is thus independent of measurement mode.
So here's my question: why is existing software structured so that we have to put this factor in by hand? Feff, for instance, could simply define chi(k) consistently with the usual procedure, so that it was normalized by the edge step rather than mu_o(E). A card could be set to turn that off if a user desired. Alternatively, a correction could be done to the experimental data by Athena, or automatically within the fitting procedure by Ifeffit.
Of course, having more than one of those options could cause trouble, just as the ability to put sigma2 into a feff calculation and in to Ifeffit sometimes does now. But wouldn't it make sense to have it available (perhaps even the default) at one of those stages?
--Scott Calvin Sarah Lawrence College
-- Bruce Ravel ------------------------------------ bravel@bnl.gov National Institute of Standards and Technology Synchrotron Methods Group at NSLS --- Beamlines U7A, X24A, X23A2 Building 535A Upton NY, 11973 My homepage: http://xafs.org/BruceRavel EXAFS software: http://cars9.uchicago.edu/~ravel/software/exafs/
Hi Scott, I maybe misunderstood your question but let me comment on it the way I see the problem. First, chi(k) is defined only above k=0 (E=E0), of course, and its defined as mu=mu_o*(1+chi(k)). chi(k) results from the scattering electron wave function being non-monotonic and is stripped of any information about mu_o. In that sense, chi(k) is completely independent on mu_o, and all parts of chi(k) including f(k), delta(k) and lambda(k) that FEFF calculates do not have any idea about mu_o either. Thus, if FEFF defined chi(k) using mu-mu_o/edge step instead of mu-mu_o/mu_o which is a correct definition of chi(k), then FEFF would have to know the McMaster correction to recover a more accurate chi(k), or would have to know a mu_o, which would be even better, but either option introduces additional theoretical approximations to the more accurate chi(k) that FEFF can produce without tampering with McMaster correction. Maybe another way to handle that would be to normalize mu not by the edge step but by mu_o, and that would make experimental chi(k) in a better agreement with the way it is calculated by FEFFand theoretical EXAFS equation, and not require any McMaster correction? That would make chi(k), of course, not realistic near the edge since mu_o near the edge is not reliable, but we are not using chi(k) in the k-range between 0 and k_min anyway, so nobody who uses chi(k) for Fourier transform purposes (in fits or whatever) will not notice the difference. Thus, such procedure would get away with McMaster correction alltogether. Anatoly ________________________________ From: ifeffit-bounces@millenia.cars.aps.anl.gov on behalf of Scott Calvin Sent: Thu 6/16/2011 8:28 PM To: XAFS Analysis using Ifeffit Subject: [Ifeffit] McMaster correction Hi all, I've been pondering the McMaster correction recently. My understanding is that it is a correction because while chi(k) is defined relative to the embedded-atom background mu_o(E), we almost always extract it from our data by normalizing by the edge step. Since mu_o(E) drops gradually above the edge, the normalization procedure results in oscillations that are too small well above edge, which the McMaster correction then compensates for. It's also my understanding that this correction is the same whether the data is measured in absorption or fluorescence, because in this context mu_o(E) refers only to absorption due to the edge of interest, which is a characteristic of the atom in its local environment and is thus independent of measurement mode. So here's my question: why is existing software structured so that we have to put this factor in by hand? Feff, for instance, could simply define chi(k) consistently with the usual procedure, so that it was normalized by the edge step rather than mu_o(E). A card could be set to turn that off if a user desired. Alternatively, a correction could be done to the experimental data by Athena, or automatically within the fitting procedure by Ifeffit. Of course, having more than one of those options could cause trouble, just as the ability to put sigma2 into a feff calculation and in to Ifeffit sometimes does now. But wouldn't it make sense to have it available (perhaps even the default) at one of those stages? --Scott Calvin Sarah Lawrence College
Hi Scott,
I would say that it does not belong in Feff, as it is a correction to
the approximation of using constant value for normalizing mu(E) to get
to chi(k). That is, in
chi(E) = [mu(E) - mu_0(E)] / Edge_Step
The approximation is in using a constant value for Edge_Step, while we
ought to normalize by a value that decays as mu(E) should. I don't
think Feff should try to include a correction that makes assumptions
on how chi(k) is extracted from mu(E).
A McMaster-like correction could be included as a default when
normalizing data, but for a few points:
1. Hardly anyone measures mu(E) with enough accuracy to be useful
by itself -- tabulated values for the decay of mu(E) would have to be
used (hence the name McMaster). Of course, the McMaster correction
should be applied to fluorescence as well as transmission data, and
they typically have very different instrumentation drifts.
2. Which tables? How accurate are these?
3. You have to know the absorbing element and excited edge in order
to make the correction. I think this is not so easy to automate.
4. Ignoring the McMaster correction adds a small static component
to sigma^2, at least for most hard x-ray edges (it's more serious for
low energy edges). So the error is in the absolute value of sigma^2,
but not the relative value of two spectra on the same element/edge.
Hope that helps. I have to admit I'm a little uneasy with the
frequency of "I've been pondering..." discussion topics alternating
with requests to review book chapters, and find myself being more
cautious in my response than I would if someone was actually asking a
question. On the other hand, I don't think anyone would object if
you added a button to Athena that normalized the data in a way that
included a correction for the expected decay of mu(E).
Cheers,
--Matt
On Thu, Jun 16, 2011 at 7:28 PM, Scott Calvin
Hi all, I've been pondering the McMaster correction recently. My understanding is that it is a correction because while chi(k) is defined relative to the embedded-atom background mu_o(E), we almost always extract it from our data by normalizing by the edge step. Since mu_o(E) drops gradually above the edge, the normalization procedure results in oscillations that are too small well above edge, which the McMaster correction then compensates for. It's also my understanding that this correction is the same whether the data is measured in absorption or fluorescence, because in this context mu_o(E) refers only to absorption due to the edge of interest, which is a characteristic of the atom in its local environment and is thus independent of measurement mode. So here's my question: why is existing software structured so that we have to put this factor in by hand? Feff, for instance, could simply define chi(k) consistently with the usual procedure, so that it was normalized by the edge step rather than mu_o(E). A card could be set to turn that off if a user desired. Alternatively, a correction could be done to the experimental data by Athena, or automatically within the fitting procedure by Ifeffit. Of course, having more than one of those options could cause trouble, just as the ability to put sigma2 into a feff calculation and in to Ifeffit sometimes does now. But wouldn't it make sense to have it available (perhaps even the default) at one of those stages? --Scott Calvin Sarah Lawrence College _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Hi Scott et al., Thanks for bringing up this issue. Whether or not McMaster corrections are useful does seem to depend on details of the measurement. But my question is: for the cases where they are useful, can one do better? As the data & theory get better and better, perhaps we should try to extract more accurate cross sections mu(E). For example, is it at all of interest to have embedded atom cross-sections to replace the atomic based Cromer-Liberman cross sections or empirical tables? John On Thu, 16 Jun 2011, Scott Calvin wrote:
Hi all, I've been pondering the McMaster correction recently.
My understanding is that it is a correction because while chi(k) is defined relative to the embedded-atom background mu_o(E), we almost always extract it from our data by normalizing by the edge step. Since mu_o(E) drops gradually above the edge, the normalization procedure results in oscillations that are too small well above edge, which the McMaster correction then compensates for. It's also my understanding that this correction is the same whether the data is measured in absorption or fluorescence, because in this context mu_o(E) refers only to absorption due to the edge of interest, which is a characteristic of the atom in its local environment and is thus independent of measurement mode.
So here's my question: why is existing software structured so that we have to put this factor in by hand? Feff, for instance, could simply define chi(k) consistently with the usual procedure, so that it was normalized by the edge step rather than mu_o(E). A card could be set to turn that off if a user desired. Alternatively, a correction could be done to the experimental data by Athena, or automatically within the fitting procedure by Ifeffit.
Of course, having more than one of those options could cause trouble, just as the ability to put sigma2 into a feff calculation and in to Ifeffit sometimes does now. But wouldn't it make sense to have it available (perhaps even the default) at one of those stages?
--Scott Calvin Sarah Lawrence College