Hi all, I read your papers Anatoly. You fixed the value of S02 found in the Pt metal, and the expression in Artemis would be like this: N:1 x S02: amp·N, right?? I have a low value of S02 in the case of platinum (0.53), how can I solve the problem? Bond distances = delR + Reff (with phase correction)? Thank you very much, Best regards, JA
If you mean that your S02 value for platinum foil is 0.53 instead of the expected 0.85 or so, you have to first find out why that happened. There are too many (for a quick list) setup-related and sample-related reasons that can explain why it happened. You should be careful with your NP data analysis, since your foil EXAFS acts up. Anatoly -----Original Message----- From: ifeffit-bounces@millenia.cars.aps.anl.gov [mailto:ifeffit-bounces@millenia.cars.aps.anl.gov] On Behalf Of Juan Antonio Maciá Agulló Sent: Monday, December 04, 2006 2:03 PM To: XAFS Analysis using Ifeffit Subject: [Ifeffit] CN and bond distances in Artemis Hi all, I read your papers Anatoly. You fixed the value of S02 found in the Pt metal, and the expression in Artemis would be like this: N:1 x S02: amp·N, right?? I have a low value of S02 in the case of platinum (0.53), how can I solve the problem? Bond distances = delR + Reff (with phase correction)? Thank you very much, Best regards, JA
Hi Juan, You're saying you have that low a value of S02 for a bulk platinum standard? Anatoly's right...you'd better figure out why before you continue. But one piece of advice I have: for a simple crystalline standard like platinum, by all means fit more than one coordination shell (with coordination numbers constrained)! That will be very helpful in diagnosing what is going on, and, for certain cases, might even solve the problem. Which brings up the first important question to ask...what is the uncertainty on the S02 of 0.53? --Scott Calvin Sarah Lawrence College At 02:03 PM 12/4/2006, you wrote:
Hi all,
I read your papers Anatoly. You fixed the value of S02 found in the Pt metal, and the expression in Artemis would be like this: N:1 x S02: amp·N, right?? I have a low value of S02 in the case of platinum (0.53), how can I solve the problem? Bond distances = delR + Reff (with phase correction)?
Did you check your data against someone else's? Maybe you have hole effect
or overabsorption.
mam
----- Original Message -----
From: "Scott Calvin"
Hi Juan,
You're saying you have that low a value of S02 for a bulk platinum standard? Anatoly's right...you'd better figure out why before you continue. But one piece of advice I have: for a simple crystalline standard like platinum, by all means fit more than one coordination shell (with coordination numbers constrained)! That will be very helpful in diagnosing what is going on, and, for certain cases, might even solve the problem.
Which brings up the first important question to ask...what is the uncertainty on the S02 of 0.53?
--Scott Calvin Sarah Lawrence College
At 02:03 PM 12/4/2006, you wrote:
Hi all,
I read your papers Anatoly. You fixed the value of S02 found in the Pt metal, and the expression in Artemis would be like this: N:1 x S02: amp·N, right?? I have a low value of S02 in the case of platinum (0.53), how can I solve the problem? Bond distances = delR + Reff (with phase correction)?
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Hi all, I did a multiple shell fit for platinum foil and the results were similar: S02 = 0.51 and the uncertainty varies from 0.02 to 0.03 depending on the fit. Please find enclosed a file with values of energy, I0 and I of the platinum foil. I include this file if someone has curiosity on raw data and has time, otherwise forget it. Maybe I can suppose a value of 0.85 for S02 and follow with the analysis of the data, then...CN equation proposed in the first e-mail is ok? Thank you very much Best regards, JA 11297.93828 2878 372358 11306.20956 2906 372292 11314.43751 2914 372058 11322.95624 2928 371952 11331.26478 2944 371682 11339.80927 2960 371614 11348.08699 2992 371372 11356.65728 3006 371244 11365.01615 3044 371104 11373.66867 3082 370976 11381.94038 3120 370776 11390.5627 3152 370662 11399.02884 3190 370420 11407.73406 3218 370214 11415.99953 3270 370056 11424.61761 3328 369524 11433.13526 3366 368060 11441.89347 3444 367558 11450.2663 3570 368788 11458.87993 3644 363456 11467.62108 10656 1026572 11468.65034 10738 1030714 11469.27942 11064 1052010 11469.96578 11292 1071180 11470.76663 11450 1079322 11471.45316 11466 1080080 11472.08256 11502 1078352 11472.76926 11576 1081712 11473.51327 11638 1082000 11474.31463 11626 1079298 11475.00159 11692 1078474 11475.80316 11694 1073826 11476.54757 11730 1076298 11477.17754 11728 1072708 11477.92213 11640 1061240 11478.49496 11654 1054028 11479.23973 11748 1055294 11480.0419 11720 1047422 11480.67225 11572 1030644 11481.35999 11586 1028928 11482.10514 11862 1048432 11482.85038 12238 1074564 11483.48105 12450 1089808 11484.05444 12570 1096612 11484.8573 12674 1099990 11485.71762 12760 1101194 11486.40597 12826 1101086 11487.15178 12860 1099050 11487.95507 12864 1091706 11488.6437 12902 1088520 11489.27501 12960 1091260 11489.9064 13108 1097922 11490.71008 13234 1102234 11491.45646 13308 1103140 11492.14551 13402 1103496 11492.83465 13476 1103560 11493.58131 13564 1103510 11494.27062 13622 1103130 11494.90256 13704 1101982 11495.59203 13762 1100862 11496.39652 13848 1101250 11497.14365 13950 1102366 11497.89088 14076 1102820 11498.6382 14154 1102564 11499.44312 14252 1102140 11500.13315 14346 1101342 11500.82326 14426 1099622 11501.45593 14466 1094870 11502.1462 14442 1084690 11503.00916 14442 1076538 11503.64207 14650 1080264 11504.33261 14896 1090874 11505.13834 15128 1099578 11505.82906 15316 1102778 11506.40473 15462 1103656 11507.03802 15548 1103634 11507.90172 15696 1103636 11508.65037 15846 1103346 11509.39912 15976 1103184 11510.14796 16108 1103142 11510.95452 16272 1102862 11511.58833 16428 1102636 11512.27983 16586 1102466 11512.97142 16752 1101876 11513.6631 16900 1101276 11514.47016 17038 1100160 11515.10436 17186 1098416 11515.85396 17302 1092602 11516.60365 17316 1079634 11517.29577 17206 1062014 11517.98797 17124 1044168 11518.56487 17188 1038548 11519.37263 17570 1050028 11520.1805 18102 1067308 11520.93077 18648 1085278 11521.68114 19042 1093338 11522.43161 19346 1096746 11523.18218 19662 1098266 11523.81735 19934 1099186 11524.4526 20208 1099696 11525.26119 20538 1100070 11526.0699 20846 1100292 11526.7054 21174 1100530 11527.39875 21476 1100532 11528.20777 21898 1100872 11528.9013 22222 1100802 11529.53712 22596 1100874 11530.23081 23024 1100924 11530.98241 23440 1101092 11531.79194 23860 1101046 11532.48591 24308 1101016 11533.29565 24830 1100974 11534.04766 25300 1101202 11534.7419 25834 1101138 11535.37837 26340 1101038 11536.07278 26990 1100918 11536.82516 27586 1101058 11537.63551 28202 1100970 11538.3302 28932 1100944 11539.02497 29656 1100886 11539.71983 30380 1100882 11540.41477 31206 1100734 11541.10979 32164 1100714 11541.74698 33074 1100598 11542.5001 34036 1100446 11543.36921 35154 1100304 11544.0646 36308 1100154 11544.87599 37486 1099964 11545.62953 38816 1099796 11546.3252 40330 1099748 11547.02095 41866 1099738 11547.6008 43480 1099556 11548.4127 45452 1099332 11549.22471 47438 1099152 11549.8628 49624 1099186 11550.617 52256 1098970 11551.3713 54904 1098916 11552.06767 57896 1098896 11552.64804 61190 1098838 11553.34456 65264 1098862 11554.15728 69526 1098712 11554.91205 74346 1098854 11555.72499 80362 1098758 11556.47997 86768 1098642 11557.23505 94306 1098736 11557.93213 103146 1098598 11558.6293 114598 1098670 11559.26845 127236 1098714 11560.02389 142650 1098682 11560.89569 163064 1098582 11561.53509 186190 1098508 11562.17456 214738 1098638 11562.9304 249650 1098506 11563.68633 295344 1098518 11564.32605 342848 1098482 11564.96583 391768 1098424 11565.83838 439014 1098346 11566.65289 469606 1098328 11567.29293 485060 1098214 11568.10764 487490 1098262 11568.92247 480756 1098044 11569.56277 470524 1098138 11570.20315 459672 1098000 11570.90182 448844 1097946 11571.65881 440328 1097878 11572.47415 433624 1097858 11573.11485 428168 1097892 11573.81388 422572 1097878 11574.57126 418230 1097864 11575.27047 414744 1097808 11575.96976 412272 1097668 11576.61086 411238 1097700 11577.36861 411580 1097562 11578.18476 412934 1097552 11578.94272 415342 1097446 11579.70078 417852 1097482 11580.45894 420508 1097482 11581.2172 422460 1097512 11581.85889 423052 1097368 11582.50065 422400 1097442 11583.25918 420474 1097268 11584.07617 417308 1097152 11584.77655 414140 1096948 11585.41863 411696 1096896 11586.23594 409922 1096674 11586.93658 409178 1096546 11587.57891 409106 1096484 11588.22131 409384 1096376 11589.03901 410036 1096314 11589.85684 411344 1096128 11590.61635 413220 1096236 11591.4344 416198 1096118 11592.19412 419516 1095998 11592.83704 423462 1096044 11593.53849 427360 1095864 11594.24003 431064 1095882 11594.94165 433598 1095716 11595.76032 434834 1095568 11596.46213 434648 1095618 11597.16402 433270 1095280 11597.92451 431106 1095010 11598.62658 428464 1095158 11599.32874 425664 1094714 11599.97247 423308 1094518 11600.73332 421102 1094448 11601.61136 419332 1094362 11602.31389 418196 1094292 11603.07506 417268 1094092 11603.83633 416808 1094052 11604.59771 416208 1094096 11605.24202 415492 1093946 11605.88641 414580 1093862 11606.64806 413432 1093804 11607.46841 412342 1093760 11608.17165 411344 1093674 11608.81638 410764 1093650 11609.63703 410552 1093530 11610.34055 410764 1093428 11610.98551 411534 1093392 11611.63055 412692 1093346 11612.45161 414156 1093426 11613.27279 415772 1093404 11613.97676 417432 1093340 11614.79816 419356 1093448 11615.56099 421052 1093428 11616.20654 422846 1093376 11616.96956 424750 1093322 11617.61528 426576 1093392 11618.31977 428298 1093368 11619.1418 429302 1093386 11619.84648 430398 1093272 11620.55125 431388 1093328 11621.31485 431922 1093302 11622.07856 432106 1093330 11622.72485 431842 1093410 11623.37121 431430 1093236 11624.13519 430892 1093216 11625.01682 430116 1093212 11625.72223 429482 1093170 11626.48652 428634 1093172 11627.25091 285567 728871 11630.48601 283523 728715 11633.84068 282067 728587 11637.19731 280855 728462 11640.49696 278516 728231 11643.85749 275981 728187 11647.16099 274539 728036 11650.82064 274314 727745 11654.12814 275478 727701 11657.55578 278283 727540 11660.92632 281070 727439 11664.35801 282680 727248 11667.73254 282990 727115 11671.0498 282742 726852 11674.66542 282377 726735 11678.046 281902 726512 11681.36921 281938 726314 11684.81313 282837 726171 11688.19968 283756 725923 11691.64769 284093 725664 11695.03826 283470 725496 11698.60944 281149 723161 11702.00411 279180 723368 11705.28157 277894 725094 11708.79946 277044 724984 11712.20013 277322 724750 11715.66253 278242 724546 11718.94775 279890 724387 11722.59355 281420 724214 11726.00236 282594 724029 11729.35333 283389 723259 11732.88592 284385 723606 11736.18095 284760 723233 11739.71771 284627 722876 11743.07664 284716 722834 11746.67765 284961 722710 11750.1007 285510 722036 11753.46565 286484 722275 11757.01299 287014 722193 11760.38197 286538 721966 11763.93355 285147 721770 11767.3668 283246 721581 11770.92266 281918 721343 11774.42038 281231 721229 11777.67877 280908 721008 11781.30135 280749 720773 11784.68447 280921 720623 11788.25096 281069 720345 11791.69863 281356 720189 11795.20888 281906 720057 11798.72126 282694 719819 11801.99331 283817 719685 11805.69173 285333 719516 11809.02839 286677 719373 11812.60985 287925 719153 11816.01123 288901 719035 11819.59698 289127 718738 11823.12411 288819 718657 11826.47077 288124 718457 11830.18478 287084 718245 11833.47456 285693 718187 11837.07107 284518 717893 11840.54778 283450 717775 11844.14865 282880 717554 11847.56849 282867 717374 11850.99033 283207 717276 11854.65881 283789 717166 11858.08481 284387 716961 11861.69651 284400 716660 11865.24918 283522 715044 11868.804 283811 716118 11872.2383 283024 713604 11875.55185 284504 715639 11879.35861 284465 713386 11882.79911 286034 715537 11886.36461 287058 715437 11889.93228 287976 715146 11893.44056 288692 715035 11896.95093 289188 714778 11900.34013 289391 714642 11904.16303 289124 714600 11907.61812 288282 714361 11911.19874 287245 714152 11914.65797 286002 714005 11918.24289 285530 713889 11921.76813 285091 713768 11925.17169 284936 713329 11929.07273 284897 712470 11932.5425 285563 713155 11936.01431 285736 713063 11939.61228 286126 712962 11943.15036 286410 712769 11946.69056 286719 712528 11950.17073 286609 712154 11953.90177 286371 711944 11957.44844 286192 711832 11960.99724 286203 711481 11964.6105 286790 711411 11968.22598 287436 711152 11971.78128 288063 710972 11975.21386 288418 710726 11979.02324 287369 706829 11982.58505 289227 710887 11986.08646 289468 710639 11989.71511 289643 710355 11993.28336 289790 710114 11996.85378 289632 709518 12000.42634 287565 704647 12004.25201 288347 707416 12007.82905 286762 704668 12011.28262 288081 709732 12014.92676 287257 709565 12018.44737 286566 709310 12022.09591 286260 709164 12025.68374 286074 708751 12029.52576 285521 705795 12033.18113 286666 708594 12036.58643 287044 708449 12040.30927 287544 708269 12043.84498 287889 708025 12047.50918 288249 707886 12051.11241 288571 707567 12054.84437 288995 707419 12058.45204 289308 707370 12061.99856 289621 707014 12065.80075 289764 706772 12069.35165 290072 706741 12073.09508 290149 706443 12076.58683 290290 706282 12080.39835 290494 706185 12084.02155 290568 705940 12087.51972 290628 705790 12091.33824 290417 705691 12094.8407 290226 705525 12098.60017 289947 705353 12102.23445 289291 703844 12106.0624 289361 704597 12109.70123 288029 701506 12113.21448 288117 701802 12116.98554 288859 703994 12120.50308 288107 702413 12124.34276 288515 703604 12127.99274 288323 703433 12131.77314 288192 703198 12135.42765 288116 702964 12138.89189 288141 702753 12142.8076 288209 702558 12146.4046 288364 702264 12150.13235 288144 700822 12153.86242 288583 701084 12157.53044 288324 699626 12161.1363 289296 700976 12164.74432 289592 701034 12168.67696 290027 700955 12172.28952 290336 700758 12176.09796 290573 700489 12179.71499 290648 700166 12183.4635 290950 699904 12187.14965 291119 699793 12190.70861 291206 699598 12194.72306 291140 699307 12198.2865 291164 699151 12202.0466 291065 698993 12205.80906 290947 698654 12209.57386 290686 698600 12213.27605 290378 698271 12216.85049 290196 698094 12220.81736 290038 697923 12224.39627 289836 697704 12228.17271 289858 697605 12231.88634 289725 697415 12235.66747 289642 697287 12239.38571 289650 696988 12242.97566 289739 696863 12247.02508 289596 696683 12250.68495 289664 696589 12254.41245 289476 696403 12258.14225 289351 696201 12261.87435 289369 696088 12265.60875 289356 695842 12269.21431 288734 694214 12273.28134 289520 695382 12277.02279 289685 695169 12280.76655 290036 695208 12284.51263 290379 695014 12288.26102 290732 694816 12292.01174 291112 694666 12295.69891 291440 694494 12299.78375 291690 694298 12303.47564 292046 694149 12307.16979 292201 693914 12310.99823 292317 693771 12314.76303 292317 693547 12318.59626 292274 693319 12322.29961 292215 693127 12326.26999 291965 692836 12330.04426 291808 692632 12333.75459 291595 692403 12337.59981 291335 692148 12341.38111 291001 691937 12345.16476 290955 691729 12350.08037 290809 691520 12354.86693 290661 691119 12359.9235 290664 690932 12364.91764 290616 690647 12369.71587 290634 690271 12374.51787 290730 689977 12379.65752 290652 689702 12384.46735 290604 689496 12389.41473 290625 689123 12394.43305 290674 688738 12399.3885 290680 688542 12404.34797 290744 688287 12409.17725 291016 687989 12414.14465 291002 687449 12419.31772 291269 687590 12424.15882 291420 687108 12429.00375 291650 686897 12434.18936 291963 686902 12439.04222 292196 686707 12444.10136 292224 686296 12448.96206 292441 686270 12454.02938 292424 685932 12459.16853 292350 685574 12464.10886 292458 685504 12468.91765 292231 685159 12474.20488 292220 684814 12479.02157 292022 684475 12484.11371 292049 684219 12489.07411 291743 683842 12494.17457 291640 683558 12499.21115 291434 683196 12504.25186 291299 682884 12509.16028 291177 682425 12514.48214 291088 682098 12519.33039 291109 681695 12524.31918 290492 679816 12529.51726 291162 681178 12534.51428 290789 679750 12539.6524 291363 680637
Hi Juan Antonio: It looks like this is fluorescence data. It is likely that you have significant self-absorption if this is a Pt foil. Try to apply an "infinite" sample correction to this data before fitting. That should correct the SO2 somewhat. A much better thing to do is to take this data in transmission. i suspect that if it is a standard foil, it is optimized for transmission and should be measured in that mode. Carlo On Tue, 5 Dec 2006, Juan Antonio Maciá Agulló wrote:
Hi all,
I did a multiple shell fit for platinum foil and the results were similar: S02 = 0.51 and the uncertainty varies from 0.02 to 0.03 depending on the fit. Please find enclosed a file with values of energy, I0 and I of the platinum foil. I include this file if someone has curiosity on raw data and has time, otherwise forget it. Maybe I can suppose a value of 0.85 for S02 and follow with the analysis of the data, then...CN equation proposed in the first e-mail is ok?
Thank you very much
Best regards, JA
-- Carlo U. Segre -- Professor of Physics Associate Dean for Special Projects, Graduate College Illinois Institute of Technology Voice: 312.567.3498 Fax: 312.567.3494 segre@iit.edu http://www.iit.edu/~segre segre@debian.org
On Tuesday 05 December 2006 09:03, Carlo Segre wrote:
It looks like this is fluorescence data. It is likely that you have significant self-absorption if this is a Pt foil. Try to apply an "infinite" sample correction to this data before fitting. That should correct the SO2 somewhat.
I agree with Carlo. I tried correcting your data as he suggested and it more or less doubled the amplitude of chi(k). This does suggest one of the cardinal rules of EXAFS data analysis, which is that good data analysis starts with careful sample preparation. B -- Bruce Ravel ---------------------------------------------- bravel@anl.gov Molecular Environmental Science Group, Building 203, Room E-165 MRCAT, Sector 10, Advanced Photon Source, Building 433, Room B007 Argonne National Laboratory phone and voice mail: (1) 630 252 5033 Argonne IL 60439, USA fax: (1) 630 252 9793 My homepage: http://cars9.uchicago.edu/~ravel EXAFS software: http://cars9.uchicago.edu/~ravel/software/exafs/
Ok, how can I apply an "infinite" sample correction? The other samples are measured also in fluorescence mode, then is it better to measure Ptfoil in the same conditions (fluorescence)? Thanks a lot Best regards, JA
Juan, There are 2 ways to apply the self-absorption correction in Athena. The first way is to pick thicknesses starting with a couple microns and going thicker and thicker until the correction has no effect. I like this method because it gives an idea of the effects that thickness and concentration can have on the fluorescence. The other way is to simply use a thickness that will be in the infinite thickness domain to begin with. Either way, you get the information needed. To do the self-absorption correction in Athena go to Data then self-absorption, pick your algorithm (I suggest reading the Athena document section to see which you should use), plug in the formula for the sample and then either plot the information depth or plot data and correction. Rich Juan Antonio Maciá Agulló wrote:
Ok, how can I apply an "infinite" sample correction? The other samples are measured also in fluorescence mode, then is it better to measure Ptfoil in the same conditions (fluorescence)?
Thanks a lot
Best regards, JA
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
On Tue, 5 Dec 2006, Juan Antonio Maciá Agulló wrote:
Ok, how can I apply an "infinite" sample correction? The other samples are measured also in fluorescence mode, then is it better to measure Ptfoil in the same conditions (fluorescence)?
The question about applying the correction has been answered. As to your question about the standard being measured the same way as the samples. yes, it is important to have standards be the same but not if you are compromising the data quality. If the sample is mad efor transmission it is not necessarily the right sample to measure in fluorescence. Perhaps a thin film of Pt would be a better fluorescence sample. Something with signficantly less than an absorption length of thickness. In my experience, mixing modes (transmission and fluorescence) is OK if you are careful about drawing conclusions. For example, a Pt foil is a geed standard because it permits you to see if you can get a consistent amplitude reduction factor which you can then use for your nanoparticles. For this purpose transmission is fine. If you want a Pt standard that is like your samples, then Pt black would be better but you know that it will have a reduced SO2 as well because it is in nanoparticle form. Cheers, Carlo
Thanks a lot
Best regards, JA
-- Carlo U. Segre -- Professor of Physics Associate Dean for Special Projects, Graduate College Illinois Institute of Technology Voice: 312.567.3498 Fax: 312.567.3494 segre@iit.edu http://www.iit.edu/~segre segre@debian.org
Hi all, I did fits with self absorption corrected data (Fluo and Booth algorithms) and the value for S02 is 1.20 (0.30). It seems that I am not very lucky, if I take an average value (1.20 and 0.50), it should be ok (0.85) jeje, I am joking. Then...what can I do now? Thank you very much, Best regards, JA P.S. If I do not answer until Monday do not hate me, I will be on holidays.
I would remeasure Pt foil in transmission, by all means. Anatoly -----Original Message----- From: ifeffit-bounces@millenia.cars.aps.anl.gov [mailto:ifeffit-bounces@millenia.cars.aps.anl.gov]On Behalf Of Juan Antonio Macia Agullo Sent: Tuesday, December 05, 2006 1:32 PM To: XAFS Analysis using Ifeffit Subject: Re: [Ifeffit] CN and bond distances in Artemis Hi all, I did fits with self absorption corrected data (Fluo and Booth algorithms) and the value for S02 is 1.20 (0.30). It seems that I am not very lucky, if I take an average value (1.20 and 0.50), it should be ok (0.85) jeje, I am joking. Then...what can I do now? Thank you very much, Best regards, JA P.S. If I do not answer until Monday do not hate me, I will be on holidays.
I imagine somebody has done so and would be willing to provide that data. Unfortunately, the Newville database doesn't have it,
and I don't think I do. Anatoly, surely you must have done Pt foil? Yes, I know it's best to measure everything on the same
beamline, but it's
not essential.
mam
----- Original Message -----
From: "Anatoly Frenkel"
I would remeasure Pt foil in transmission, by all means.
Anatoly
-----Original Message----- From: ifeffit-bounces@millenia.cars.aps.anl.gov [mailto:ifeffit-bounces@millenia.cars.aps.anl.gov]On Behalf Of Juan Antonio Macia Agullo Sent: Tuesday, December 05, 2006 1:32 PM To: XAFS Analysis using Ifeffit Subject: Re: [Ifeffit] CN and bond distances in Artemis
Hi all,
I did fits with self absorption corrected data (Fluo and Booth algorithms) and the value for S02 is 1.20 (0.30). It seems that I am not very lucky, if I take an average value (1.20 and 0.50), it should be ok (0.85) jeje, I am joking.
Then...what can I do now?
Thank you very much,
Best regards, JA
P.S. If I do not answer until Monday do not hate me, I will be on holidays.
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Hi Matthew,
It should be very easy to send to him Pt foil data, especially because,
based on my and others' tests, Pt data collected at different beamlines are
"identical" and thus transferable, but what if the beamline settings
(detuning, slit size, gases, etc.) were not optimized for his experiment? To
make sure everything is tuned up correctly, he should take his own Pt foil
data and compare with theory other other people standards (good idea!).
Alternatively, he will have to operate under assumption that everything was
tuned up correctly and his NP data are perfect, but why to take chances?
Anatoly
-----Original Message-----
From: Matthew Marcus [mailto:mamarcus@lbl.gov]
Sent: Tuesday, December 05, 2006 1:54 PM
To: anatoly.frenkel@yu.edu; XAFS Analysis using Ifeffit
Subject: Re: [Ifeffit] CN and bond distances in Artemis
I imagine somebody has done so and would be willing to provide that data.
Unfortunately, the Newville database doesn't have it,
and I don't think I do. Anatoly, surely you must have done Pt foil? Yes, I
know it's best to measure everything on the same
beamline, but it's
not essential.
mam
----- Original Message -----
From: "Anatoly Frenkel"
I would remeasure Pt foil in transmission, by all means.
Anatoly
-----Original Message----- From: ifeffit-bounces@millenia.cars.aps.anl.gov [mailto:ifeffit-bounces@millenia.cars.aps.anl.gov]On Behalf Of Juan Antonio Macia Agullo Sent: Tuesday, December 05, 2006 1:32 PM To: XAFS Analysis using Ifeffit Subject: Re: [Ifeffit] CN and bond distances in Artemis
Hi all,
I did fits with self absorption corrected data (Fluo and Booth algorithms) and the value for S02 is 1.20 (0.30). It seems that I am not very lucky, if I take an average value (1.20 and 0.50), it should be ok (0.85) jeje, I am joking.
Then...what can I do now?
Thank you very much,
Best regards, JA
P.S. If I do not answer until Monday do not hate me, I will be on holidays.
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Hi JA, I agree with Anatoly that remeasuring in transmission is a good idea, if you have the beam time. But I would also note that the value you have for S02 with the correction is within the realm of possibility, considering the uncertainty: 1.20 - 0.30 = 0.90. And part of the reason for starting with a standard in the first place is to work out some of the other difficulties that may arise; e.g. questions of normalization. If it is not practical for you to remeasure right now, you can try some of the things that can reduce the uncertainty in S02: fitting with multiple k-weights, changing the k-range being fit, fitting multiple coordination shells, etc.. I also just checked a past paper of mine, and see I ended up with an S02 for a set of platinum samples (transmission) of 1.16 +/- 0.05, suggesting that I wasn't doing something quite right (probably not sample preparation or measurement issues, since most of those tend to lower S02). But since that set of samples included a standard, and I was really interested in the differences between samples (including the standard), the fact that I have some systematic error that's causing the S02 to come out a tad large was acceptable, if a bit troubling. The paper is S. Calvin et al., Appl. Phys. Lett. 87, 233102 (2005). --Scott Calvin Sarah Lawrence College P.S. Just as I finished writing this, Matthew Marcus' post came through. I do have Pt foil data in transmission that I can dig up for you, if you'd like. Maybe I'll send it to Matt to put in the database; it's pretty good quality, as I recall (good data out to 20 inverse angstroms or so). At 01:31 PM 12/5/2006, Juan Antonio Maciá Agulló wrote:
Hi all,
I did fits with self absorption corrected data (Fluo and Booth algorithms) and the value for S02 is 1.20 (0.30). It seems that I am not very lucky, if I take an average value (1.20 and 0.50), it should be ok (0.85) jeje, I am joking.
Then...what can I do now?
On Tuesday 05 December 2006 13:02, Scott Calvin wrote:
P.S. Just as I finished writing this, Matthew Marcus' post came through. I do have Pt foil data in transmission that I can dig up for you, if you'd like. Maybe I'll send it to Matt to put in the database; it's pretty good quality, as I recall (good data out to 20 inverse angstroms or so).
A few years ago, when I was commissioning X11b at NSLS, I prepared a some reports on how things were progressing. This one: http://cars9.uchicago.edu/~ravel/misc/X11b/update_sep02.html includes a link to an athena project file containing my measurements there on several foils and filters that we had lying around X11, including a Pt foil. It comes with all the caveats that the others have already discussed. I agree with what others have said -- Juan would be best served by remeasuring the Pt foil in transmission. That's not, however, the most immediately useful advice. Juan has a problem that he needs to solve *now*. It is quite possible that he will not be able to go to a synchrotron for quite some time. Suggesting that he is stuck until he returns to the synchrotron just isn't especially helpful. What I would recommend, Juan, to get you over this road block is to compare your measured data to the data from my project file or to someone else's data (or both!). Work on the self-absorption correction until your fluorescence data most closely resembles the data you are comparing to. Then proceed with your data analysis using the many excellent suggestions you have received so far. The *next* time you go to measure data, think really hard about how best to measure each of your samples and standards. Try to figure out how to measure the best possible data from each sample. Different samples require different experimental protocols. The goal is, of course, to obtain consistently excellent data, so you will need to be resourceful and do what each sample requires. As helpful as it is to have a slew of "experts" (y'know, disreputable sorts like me) telling you what you need to do, you must remember that your education is a process. If you learn from the mistakes you made last time and improve upon them next time, then you are proceeding in the correct direction. B -- Bruce Ravel ---------------------------------------------- bravel@anl.gov Molecular Environmental Science Group, Building 203, Room E-165 MRCAT, Sector 10, Advanced Photon Source, Building 433, Room B007 Argonne National Laboratory phone and voice mail: (1) 630 252 5033 Argonne IL 60439, USA fax: (1) 630 252 9793 My homepage: http://cars9.uchicago.edu/~ravel EXAFS software: http://cars9.uchicago.edu/~ravel/software/exafs/
Konichiwa from Japan, Alas Bruce, not everyone has a unix without training wheels. The darwin OS is case insensitive. Sort of a good thing in many ways. For instance take the example of "atoms" and "Atoms" in the horae-063 directory. This frozzed tar in the unpacking stage and I didn't realize the conflict until today (I was in Italy until last Friday so didn't get a chance to compile horae until today). Can you remove the conflict with uppercase and lowercase "atoms/ Atoms" in the horae directory so this doesn' happen every time. It would also be sort of neat if there were a some sort of "honest" way to compile the ifeffit_wrapper.c source without having to hack the ucky MakeMaker output (a switch perhaps). All this is of course, when you have time. Thanks for the new release. Paul