Dear all
This is an excellent thread. Just to give some background literature, you will be aware of the calibration of X-ray energies by diffraction (either like Kraft or better accuracy from the NIST group and Dick Deslattes, etc). This remains the best method for beamline calibration, though the mono is usually affected by heatload and detuning to shift the energies. [So you want either a 4-bounce or an analyser 'mono'].
You will also be aware of the International Tables of Crystallography Volume C Section 4.2.2 [Not our XAFS one, Volume I, but the main Physical Tables Volume]. The old issue summarised a lot of details about current known and not-known accuracy of X-ray energies and is embedded in the NIST website, though the original publication[s] Deslattes et al. are in many ways superior. However, Volume C is coming out with a new edition [soon*] and the summary of the known and unknown accuracies and what to do about them is in the new section 4.2.2. It should be available this year, I understand.
Secondly, by far the best edge energy determinations are those of Kraft et al., and these were based around the subset of transition metals and K edges. Most other edges have not been published with an absolute energy measurement, so they really do flop around in the literature. This is a major and important [urgent] task that maybe we can all help with. The issue of stability of lines, crystals etc. is a closely related and important issue.
The issue is compounded at lower energies and for L edges. One of the best current examples is that of the recent Mendenhall et al. publication on copper...
I hope this thread has emphasised the importance of this work and that we would all like more progress in this area
Very best wishes
Chris
--------------------------------------------------------------------------------------
Christopher Chantler, Professor, FAIP, Fellow American Physical Society
Editor-in-Chief, Radiation Physics and Chemistry
Chair, International IUCr Commission on XAFS; CIT, CCN
IPP, International Radiation Physics Society
School of Physics, University of Melbourne
Parkville Victoria 3010 Australia
+61-3-83445437 FAX +61-3-93474783
chantler@unimelb.edu.auhttps://owa.unimelb.edu.au/owa/redir.aspx?C=c7BoS0kVVkC1_S95-9x9l5cBu6YTjdAITgSrfUpfDAUV5oUH1LFYBcz08w8xvHMJoosZRdagfQM.&URL=mailto%3achantler%40unimelb.edu.au chantler@me.comhttps://owa.unimelb.edu.au/owa/redir.aspx?C=c7BoS0kVVkC1_S95-9x9l5cBu6YTjdAITgSrfUpfDAUV5oUH1LFYBcz08w8xvHMJoosZRdagfQM.&URL=mailto%3achantler%40me.com
http://optics.ph.unimelb.edu.au/~chantler/xrayopt/xrayopt.htmlhttps://owa.unimelb.edu.au/owa/redir.aspx?C=c7BoS0kVVkC1_S95-9x9l5cBu6YTjdAITgSrfUpfDAUV5oUH1LFYBcz08w8xvHMJoosZRdagfQM.&URL=http%3a%2f%2foptics.ph.unimelb.edu.au%2f%7echantler%2fxrayopt%2fxrayopt.html
http://optics.ph.unimelb.edu.au/~chantler/home.htmlhttps://owa.unimelb.edu.au/owa/redir.aspx?C=c7BoS0kVVkC1_S95-9x9l5cBu6YTjdAITgSrfUpfDAUV5oUH1LFYBcz08w8xvHMJoosZRdagfQM.&URL=http%3a%2f%2foptics.ph.unimelb.edu.au%2f%7echantler%2fhome.html
________________________________
From: Ifeffit
Hi Matt,
Indeed, in my experience (which is limited to one beamline at one synchrotron facility for P XAS), once it is calibrated, the energy selection tends to be quite stable, so I think you're on-target there.
The trouble I still run into, though, is comparability of data between studies. The difficulty is magnified by the fact that people tend to identify certain near-edge features by the energy range at which they occur. I do the same, of course, but I also try to carefully document the material and energy I used to calibrate the monochromator.
For the P K-edge, it doesn't really seem like people have settled on a convention for calibrating the monochromator, unlike in the case of iron, for example (where one just uses a foil and sets some feature of that spectrum to their preferred value). If everyone was using the same thing all would be happy, but most people use different materials and different values. So datasets for P at the K-edge really aren't too comparable just yet.
Sorry to hijack the conversation, it's just an issue I've been mulling over for a few years. The discussion of energy calibration values made me think of it again.
Best,
Mike
On May 8, 2020, at 8:51 AM, Matt Newville
wrote: ? Hi Mike,
On Tue, May 5, 2020 at 10:56 PM Mike Massey
mailto:mmassey@gmail.com> wrote: On a tangentially related topic, I find that phosphorus K-edge XAS energy calibration conventions are still in a bit of a "Wild West" state, with a wide variety of materials and values in use for energy calibration. As an extreme example, one or two frequently cited papers in my field from the 2000s don't even report the material or value used for energy calibration, and only show portions of the spectra on an energy axis with values relative to an unknown E0.
I have never measured a P K edge, or indeed any edge lower in energy than the S K edge (ignoring some X-ray raman work).? But if one is using a Si(111) double-crystal monochromator where P or S is approximately the low-energy (high-angle) limit, then it really should be that the calibration does not drift much and cannot be too wrong at low energies.
That is, a mono calibration is controlled by a d-spacing and angular offset. Normally (or perhaps, in my experience), "re-calibrating" is done by changing the angular offset, leaving the d-spacing alone. That is, the d-spacing is presumably known, at least to within some thermal drift. If that is the case that the d-spacing really is not changing and what needs to be refined is the angular?offset, then setting the offset at relatively high energy edges will be much more sensitive, and changing the angular offset to that a high-energy edge is correct should move lower energy edges by a smaller amount.? ?The corollary is that you have to move the offset a lot to move the P? K edge around, and that would have a larger (and ever-increasing) impact on higher energy edges such as Ca, Fe, Cu or Mo.
The counter-argument is also true:? d-spacing has a bigger effect on the high-angle / low-energy edges.
So, if you believe the mono d-spacing (or you believe the beamline scientist who believes it ;)) then calibrate at the highest energy you can.? ?The Kraft values don't go very low in energy.
All that said, if using a different mono crystal such as InSb or more exotic crystals, I have no idea how stable those are.
I too have picked my own material and value, and will be the first to acknowledge that I did so out of necessity and ease of comparison to other available data, rather than because I thought it was correct.
The issue of calibration conventions and values definitely seems to be one that merits continued discussion. It has been interesting to watch things evolve over time in the case of iron, for example (it's good to know that 7110.75 is a candidate calibration value...) I appreciate Matt's detailed thoughts, and the data that he's been working with. Thanks Matt!
Cheers,
Mike
On May 6, 2020, at 3:32 PM, Matt Newville
mailto:newville@cars.uchicago.edu> wrote: ? Hi Simon,
This is definitely a timely discussion for me, as I've been spending part of the quartine?working on collating data and expanding datasets for an XAFS spectral database.? I'm hoping to have something ready for public comment and to start asking for contributions of data in a few weeks, but I'll be happy to have more discussion about that sooner too.
I generally believe that the monochromator I use at GSECARS is both well-calibrated and reasonably accurate.? That is, with 2 angular encoders with a resolution of >130,000 lines per degree and an air-bearing, I believe the angular accuracy and repeatability are very good.? I believe there are equally good moons in existence.? ?As Matthew Marcus pointed to the Kraft paper (which used an older source but 4-bounce mono to improve resolution), we find that Fe foil is definitely?better defined as 7110.75 and Cu foil is between 8980.0 and 8980.5 eV.? That is, we've measured multiple foils, found their first derivatives, and refined the d-spacing and angular offset.? We do this about once per run, and the offsets tend to be very consistent.? ?For sure, there is some question about whether the Kraft numbers are perfect.? ?For sure, putting Fe foil at 7110.75?+/- 0.25 eV appears to be "most right" to us.
I also believe that we should probably re-measure these metal foils (and other compounds) with a single calibration set for both Si(111) and Si(311).? We will probably have time to do that this summer in the time between "beamline staff can get back to the beamline" and "open for outside users".
What I can tell you now is:? I have some data on W metal, WO2, and WO3 measured all at the same time on our bending magnet line, with Si(111).? An Athena project for this is attached (W.prj). ?I cannot vouch for the absolute calibration.
I also attach a set of foils (V, Fe, Cu, Mo) measured with the same calibration (and Si(111) on our ID line), after adjusting d-space and offset?to be close to the Kraft values (CalibratedFoils2013.prj).
I also attach a set of foils (Fe, Cu, Au L3, Au L2, Au L1, Pb L3, Pb L2, Pb L1 edges) measured in 2016 (again, using Si(111) on our ID line), also with the same calibration values (FeCu_Au_Pb.prj).? I'm pretty certain these use the same d-spacing as the 2013 Foils to?at least 5 digits.? ?For completeness, all of the raw data files are also under https://github.com/XraySpectroscopy/XASDataLibrary/tree/master/data
In my experience, the Pb L3 edge value has the biggest variation in the literature, with values ranging from 13035 to 13055 eV (possibly a typo somewhere along the line).? Fortunately, the Kraft-based calibration splits the difference and puts the value at 13040 eV.
For W in particular, I will look if I have measured this recently on our ID line.? I can tell you that I use CdWO4 as a phosphor and use that to focus our X-ray beam.? ?I use this trick all the time: any tail from the beam penetrating the phosphor is shortest at the peak of the white-line and for CdWO4 that is always between 10210 and 10215 eV.
I hope that helps.? I am interested in trying to get all these values as accurately as possible, so any comments or suggestions would be most welcome.
--Matt
On Tue, May 5, 2020 at 5:14 PM Bare, Simon R
mailto:srbare@slac.stanford.edu> wrote: All:____
__ __
We are wondering if others agree that the reported values for the W L3 and W L2 edges are *incorrect*. We recently noticed the following:____
__ __
The ?Edge? ? defined by the inflection point of the absorption edge step____
__ __
When using the Ir L_3 edge (11215.0 eV) as a calibration, the W L_3 - and L_2 -edges are *10203.4 eV* and *11542.4 eV*, respectively. ____
__ __
When using the Pt L_3 edge (11564.0 eV) as a calibration, the W L_3 - and L_2 -edges are *10203.3 eV* and *11542.4 eV*, respectively.____
__ __
These observations are thus different than the reported values of *10207.0 eV* and *11544.0 eV* for the L_3 and L_2 edges, respectively.____
__ __
Thanks in advance for the discussion and feedback.____
__ __
__ __
Simon R Bare____
/Distinguished Scientist____/
/SSRL, MS69____/
/SLAC National Accelerator Lab____/
/2575 Sand Hill Road____/
/Menlo Park CA 94025____/
__ __
simon.bare@slac.stanford.edu mailto:simon.bare@slac.stanford.edu____
Ph: 650-926-2629____
__ __
____ __ __
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-- --Matt Newville
http://cars.uchicago.edu> 630-252-0431 _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-- --Matt Newville
http://cars.uchicago.edu> 630-252-0431 _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
------------------------------
Message: 2
Date: Fri, 8 May 2020 10:54:07 +1200
From: Mike Massey
On May 8, 2020, at 10:19 AM, Matthew Marcus
wrote: ?For elements like P and S, people often use the energies of peaks. These are more immune to noise, energy-resolution effects and overabsorption than inflection points are. For instance, on ALS 10.3.2, I used the sulfate peak of gypsum set at 2482.74eV. I forget where I got that number. Going down to soft X-rays, a common convention for the carbon edge is to use a pair of sharp peaks in CO2 gas at 292.74 and 294.96eV. mam
On 5/7/2020 3:09 PM, Mike Massey wrote: Hi Matt, Indeed, in my experience (which is limited to one beamline at one synchrotron facility for P XAS), once it is calibrated, the energy selection tends to be quite stable, so I think you're on-target there. The trouble I still run into, though, is comparability of data between studies. The difficulty is magnified by the fact that people tend to identify certain near-edge features by the energy range at which they occur. I do the same, of course, but I also try to carefully document the material and energy I used to calibrate the monochromator. For the P K-edge, it doesn't really seem like people have settled on a convention for calibrating the monochromator, unlike in the case of iron, for example (where one just uses a foil and sets some feature of that spectrum to their preferred value). If everyone was using the same thing all would be happy, but most people use different materials and different values. So datasets for P at the K-edge really aren't too comparable just yet. Sorry to hijack the conversation, it's just an issue I've been mulling over for a few years. The discussion of energy calibration values made me think of it again. Best, Mike
On May 8, 2020, at 8:51 AM, Matt Newville
wrote: ? Hi Mike,
On Tue, May 5, 2020 at 10:56 PM Mike Massey
mailto:mmassey@gmail.com> wrote: On a tangentially related topic, I find that phosphorus K-edge XAS energy calibration conventions are still in a bit of a "Wild West" state, with a wide variety of materials and values in use for energy calibration. As an extreme example, one or two frequently cited papers in my field from the 2000s don't even report the material or value used for energy calibration, and only show portions of the spectra on an energy axis with values relative to an unknown E0.
I have never measured a P K edge, or indeed any edge lower in energy than the S K edge (ignoring some X-ray raman work). But if one is using a Si(111) double-crystal monochromator where P or S is approximately the low-energy (high-angle) limit, then it really should be that the calibration does not drift much and cannot be too wrong at low energies.
That is, a mono calibration is controlled by a d-spacing and angular offset. Normally (or perhaps, in my experience), "re-calibrating" is done by changing the angular offset, leaving the d-spacing alone. That is, the d-spacing is presumably known, at least to within some thermal drift. If that is the case that the d-spacing really is not changing and what needs to be refined is the angular offset, then setting the offset at relatively high energy edges will be much more sensitive, and changing the angular offset to that a high-energy edge is correct should move lower energy edges by a smaller amount. The corollary is that you have to move the offset a lot to move the P K edge around, and that would have a larger (and ever-increasing) impact on higher energy edges such as Ca, Fe, Cu or Mo.
The counter-argument is also true: d-spacing has a bigger effect on the high-angle / low-energy edges.
So, if you believe the mono d-spacing (or you believe the beamline scientist who believes it ;)) then calibrate at the highest energy you can. The Kraft values don't go very low in energy.
All that said, if using a different mono crystal such as InSb or more exotic crystals, I have no idea how stable those are.
I too have picked my own material and value, and will be the first to acknowledge that I did so out of necessity and ease of comparison to other available data, rather than because I thought it was correct.
The issue of calibration conventions and values definitely seems to be one that merits continued discussion. It has been interesting to watch things evolve over time in the case of iron, for example (it's good to know that 7110.75 is a candidate calibration value...) I appreciate Matt's detailed thoughts, and the data that he's been working with. Thanks Matt!
Cheers,
Mike
On May 6, 2020, at 3:32 PM, Matt Newville
mailto:newville@cars.uchicago.edu> wrote: ? Hi Simon,
This is definitely a timely discussion for me, as I've been spending part of the quartine working on collating data and expanding datasets for an XAFS spectral database. I'm hoping to have something ready for public comment and to start asking for contributions of data in a few weeks, but I'll be happy to have more discussion about that sooner too.
I generally believe that the monochromator I use at GSECARS is both well-calibrated and reasonably accurate. That is, with 2 angular encoders with a resolution of >130,000 lines per degree and an air-bearing, I believe the angular accuracy and repeatability are very good. I believe there are equally good moons in existence. As Matthew Marcus pointed to the Kraft paper (which used an older source but 4-bounce mono to improve resolution), we find that Fe foil is definitely better defined as 7110.75 and Cu foil is between 8980.0 and 8980.5 eV. That is, we've measured multiple foils, found their first derivatives, and refined the d-spacing and angular offset. We do this about once per run, and the offsets tend to be very consistent. For sure, there is some question about whether the Kraft numbers are perfect. For sure, putting Fe foil at 7110.75 +/- 0.25 eV appears to be "most right" to us.
I also believe that we should probably re-measure these metal foils (and other compounds) with a single calibration set for both Si(111) and Si(311). We will probably have time to do that this summer in the time between "beamline staff can get back to the beamline" and "open for outside users".
What I can tell you now is: I have some data on W metal, WO2, and WO3 measured all at the same time on our bending magnet line, with Si(111). An Athena project for this is attached (W.prj). I cannot vouch for the absolute calibration.
I also attach a set of foils (V, Fe, Cu, Mo) measured with the same calibration (and Si(111) on our ID line), after adjusting d-space and offset to be close to the Kraft values (CalibratedFoils2013.prj).
I also attach a set of foils (Fe, Cu, Au L3, Au L2, Au L1, Pb L3, Pb L2, Pb L1 edges) measured in 2016 (again, using Si(111) on our ID line), also with the same calibration values (FeCu_Au_Pb.prj). I'm pretty certain these use the same d-spacing as the 2013 Foils to at least 5 digits. For completeness, all of the raw data files are also under https://github.com/XraySpectroscopy/XASDataLibrary/tree/master/data
In my experience, the Pb L3 edge value has the biggest variation in the literature, with values ranging from 13035 to 13055 eV (possibly a typo somewhere along the line). Fortunately, the Kraft-based calibration splits the difference and puts the value at 13040 eV.
For W in particular, I will look if I have measured this recently on our ID line. I can tell you that I use CdWO4 as a phosphor and use that to focus our X-ray beam. I use this trick all the time: any tail from the beam penetrating the phosphor is shortest at the peak of the white-line and for CdWO4 that is always between 10210 and 10215 eV.
I hope that helps. I am interested in trying to get all these values as accurately as possible, so any comments or suggestions would be most welcome.
--Matt
On Tue, May 5, 2020 at 5:14 PM Bare, Simon R
mailto:srbare@slac.stanford.edu> wrote: All:____
__ __
We are wondering if others agree that the reported values for the W L3 and W L2 edges are *incorrect*. We recently noticed the following:____
__ __
The ?Edge? ? defined by the inflection point of the absorption edge step____
__ __
When using the Ir L_3 edge (11215.0 eV) as a calibration, the W L_3 - and L_2 -edges are *10203.4 eV* and *11542.4 eV*, respectively. ____
__ __
When using the Pt L_3 edge (11564.0 eV) as a calibration, the W L_3 - and L_2 -edges are *10203.3 eV* and *11542.4 eV*, respectively.____
__ __
These observations are thus different than the reported values of *10207.0 eV* and *11544.0 eV* for the L_3 and L_2 edges, respectively.____
__ __
Thanks in advance for the discussion and feedback.____
__ __
__ __
Simon R Bare____
/Distinguished Scientist____/
/SSRL, MS69____/
/SLAC National Accelerator Lab____/
/2575 Sand Hill Road____/
/Menlo Park CA 94025____/
__ __
simon.bare@slac.stanford.edu mailto:simon.bare@slac.stanford.edu____
Ph: 650-926-2629____
__ __
____ __ __
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-- --Matt Newville
http://cars.uchicago.edu> 630-252-0431 _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov mailto:Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
-- --Matt Newville
http://cars.uchicago.edu> 630-252-0431 _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
------------------------------ Subject: Digest Footer _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit ------------------------------ End of Ifeffit Digest, Vol 207, Issue 14 ****************************************
participants (1)
-
Christopher Chantler