Is good EXAFS data possible with 0.01 transmission coefficient
Hi IFEFFIT community, I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my sample is a powder mixture of carbon powder, MnO2 powder, and potassium hydroxide liquid. The transmission coefficient of 6.5keV x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to 0.01. I am told that the ionization chamber detectors can easily detect the x-ray beam after this ~0.01 transmission loss (by increasing the gain on the detector), but I would like to ask the IFEFFIT community if there will be other problems with the EXAFS technique when the transmission coefficient is so low. Does anybody have comments? Much thanks -- if you have any info I greatly appreciate it! Damon Turney City College of New York
Is most of the transmission loss due to the Mn or the matrix (C, KOH, etc)? Is that 0.01 above or below the edge? If most of the absorption is due to the matrix with the Mn providing an edge jump of <1, then I'd say that you have a shot. However, you become very sensitive to harmonics and glitches. If the Mn provides an edge jump of much less than 1, them fluorescence would be the way to go, if possible. With thick samples in transmission, you have an issue with hole effect. For instance, it's possible that your 1% transmission comes entirely from 1% area of pinholes, with the actual material being almost completely opaque. If that happens, you get a reduced edge jump and reduced EXAFS amplitude, with distortions resembling overabsorption. If you're in the thick-matrix scenario, a possible workaround would be to make a sample which contains only matrix and is as nearly as possible exactly as thick as the real sample, and measure its transmission, and subtract. mam On 7/25/2013 1:07 PM, Damon Turney wrote:
Hi IFEFFIT community,
I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my sample is a powder mixture of carbon powder, MnO2 powder, and potassium hydroxide liquid. The transmission coefficient of 6.5keV x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to 0.01. I am told that the ionization chamber detectors can easily detect the x-ray beam after this ~0.01 transmission loss (by increasing the gain on the detector), but I would like to ask the IFEFFIT community if there will be other problems with the EXAFS technique when the transmission coefficient is so low. Does anybody have comments?
Much thanks -- if you have any info I greatly appreciate it! Damon Turney City College of New York _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Oops--Matthew's interpretation is doubtless what you actually meant--i.e., that It < 0.01 I0. Therefore disregard my previous comment!
And I concur with Matthew's comments, now that I understand the circumstances.
--Scott Calvin
Sarah Lawrence College
On Jul 25, 2013, at 4:33 PM, Matthew Marcus
Is most of the transmission loss due to the Mn or the matrix (C, KOH, etc)? Is that 0.01 above or below the edge? If most of the absorption is due to the matrix with the Mn providing an edge jump of <1, then I'd say that you have a shot. However, you become very sensitive to harmonics and glitches. If the Mn provides an edge jump of much less than 1, them fluorescence would be the way to go, if possible. With thick samples in transmission, you have an issue with hole effect. For instance, it's possible that your 1% transmission comes entirely from 1% area of pinholes, with the actual material being almost completely opaque. If that happens, you get a reduced edge jump and reduced EXAFS amplitude, with distortions resembling overabsorption.
If you're in the thick-matrix scenario, a possible workaround would be to make a sample which contains only matrix and is as nearly as possible exactly as thick as the real sample, and measure its transmission, and subtract. mam
On 7/25/2013 1:07 PM, Damon Turney wrote:
Hi IFEFFIT community,
I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my sample is a powder mixture of carbon powder, MnO2 powder, and potassium hydroxide liquid. The transmission coefficient of 6.5keV x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to 0.01. I am told that the ionization chamber detectors can easily detect the x-ray beam after this ~0.01 transmission loss (by increasing the gain on the detector), but I would like to ask the IFEFFIT community if there will be other problems with the EXAFS technique when the transmission coefficient is so low. Does anybody have comments?
Much thanks -- if you have any info I greatly appreciate it! Damon Turney City College of New York _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Hi Damon,
So you're saying you expect the total absorption of the sample to be less than 0.01 absorption lengths? In other words, if the detectors and electronics were identical, you would expect It = I0 exp(-0.01)?
While possible to measure in transmission, such a sample would normally be measured in fluorescence. If done that way, the measurement is routine.
Or am I misunderstanding your description? (I am not entirely sure what you mean by "transmission coefficient.")
--Scott Calvin
Sarah Lawrence College
On Jul 25, 2013, at 4:07 PM, Damon Turney
Hi IFEFFIT community,
I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my sample is a powder mixture of carbon powder, MnO2 powder, and potassium hydroxide liquid. The transmission coefficient of 6.5keV x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to 0.01. I am told that the ionization chamber detectors can easily detect the x-ray beam after this ~0.01 transmission loss (by increasing the gain on the detector), but I would like to ask the IFEFFIT community if there will be other problems with the EXAFS technique when the transmission coefficient is so low. Does anybody have comments?
Much thanks -- if you have any info I greatly appreciate it! Damon Turney City College of New York _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
I'm pretty sure he has the opposite problem - the transmission is <0.01. That's assuming that "transmission coefficient" == "transmission". mam On 7/25/2013 1:33 PM, Scott Calvin wrote:
Hi Damon,
So you're saying you expect the total absorption of the sample to be less than 0.01 absorption lengths? In other words, if the detectors and electronics were identical, you would expect It = I0 exp(-0.01)?
While possible to measure in transmission, such a sample would normally be measured in fluorescence. If done that way, the measurement is routine.
Or am I misunderstanding your description? (I am not entirely sure what you mean by "transmission coefficient.")
--Scott Calvin Sarah Lawrence College
On Jul 25, 2013, at 4:07 PM, Damon Turney
wrote: Hi IFEFFIT community,
I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my sample is a powder mixture of carbon powder, MnO2 powder, and potassium hydroxide liquid. The transmission coefficient of 6.5keV x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to 0.01. I am told that the ionization chamber detectors can easily detect the x-ray beam after this ~0.01 transmission loss (by increasing the gain on the detector), but I would like to ask the IFEFFIT community if there will be other problems with the EXAFS technique when the transmission coefficient is so low. Does anybody have comments?
Much thanks -- if you have any info I greatly appreciate it! Damon Turney City College of New York _______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit