Fellow X-Ray Absorption Enthusiasts, I have recently compiled a model that gives excellent visual fits in R, q, and k space for bond spacing in a BCC structure. This model gives bond spacings that make sense, and are very close to what would be expected from this set. The R factors are very low, and the enot values correspond quite well to the edge. However, our amplitude values are much larger than typically expected. They come in at the range of 1.8 up to 5.0, but only on a few data sets. On all the rest the amplitude values are 0.4 to 1.0. Could this increase in amplitude be attributed to the fact that we ran florescence measurements instead of transmission, and have a weaker signal coming to the detector? What else could be causing this in only one data set? All samples used in this model have the same structure. Thanks in advance to any replies, your help and time is appreciated. Gavin gavin_garside@yahoo.com
Hi Gavin, What are the uncertainties on the high S02 values? Fluorescence is unlikely to be the culprit. While it can affect your ability to normalize properly, you're unlikely to account for a factor of 2 by normalization if the data is relatively decent. And self- absorption tends to suppress S02, not exaggerate it. Why did you switch to fluorescence on just the handful of data sets? That might provide us a clue. --Scott Calvin Sarah Lawrence College On Jul 30, 2010, at 10:47 PM, Gavin Garside wrote:
Fellow X-Ray Absorption Enthusiasts,
I have recently compiled a model that gives excellent visual fits in R, q, and k space for bond spacing in a BCC structure. This model gives bond spacings that make sense, and are very close to what would be expected from this set. The R factors are very low, and the enot values correspond quite well to the edge. However, our amplitude values are much larger than typically expected. They come in at the range of 1.8 up to 5.0, but only on a few data sets. On all the rest the amplitude values are 0.4 to 1.0. Could this increase in amplitude be attributed to the fact that we ran florescence measurements instead of transmission, and have a weaker signal coming to the detector? What else could be causing this in only one data set? All samples used in this model have the same structure. Thanks in advance to any replies, your help and time is appreciated.
Scott,
Thank you for a quick response. The value I am getting for SO2 in the fit most
of interest is 2.95 plus/minus 3.72. So with the error bar I am in range, but I
was just suspicious of it before I make any claims about it. All my experiments
were done in florescence because we have ordered bulk material. By creating a
sample that would work in fluorescence I may have introduced dislocations or
imperfections that would have effected the physical properties of interest in
this sample.
Gavin Garside
University of Utah
________________________________
From: Scott Calvin
I have recently compiled a model that gives excellent visual fits in R, q, and k space for bond spacing in a BCC structure. This model gives bond spacings that make sense, and are very close to what would be expected from this set. The R factors are very low, and the enot values correspond quite well to the edge. However, our amplitude values are much larger than typically expected. They come in at the range of 1.8 up to 5.0, but only on a few data sets. On all the rest the amplitude values are 0.4 to 1.0. Could this increase in amplitude be attributed to the fact that we ran florescence measurements instead of transmission, and have a weaker signal coming to the detector? What else could be causing this in only one data set? All samples used in this model have the same structure. Thanks in advance to any replies, your help and time is appreciated.
Hi Gavin, The problem isn't the value of S02; it's the uncertainty. I've noticed that Ifeffit has a tendency to push up the best fit value of S02 when it's very uncertain, although I'd have to think more about how it determines error bars to confirm that. But it does make sense, in a "Price is Right" kind of way--negative S02's presumably fit horribly, because they turn chi(k) upside down, and so that biases uncertain S02's to move the whole range up. In any case, your focus should be on reducing that uncertainty. 1) One way to do that is to fit with multiple k-weights, assuming you're not doing that already. Check the kw 1, kw 2, and kw 3 boxes all together and run a fit. The reason this works is that the EXFAS equation shows no k-dependence for S02, but a k^2 dependence for sigma^2, which often shows a high correlation with it in fits. Fitting multiple k-weights sometimes helps break that correlation. 2) Along the same lines, if you can squeeze out any additional k-range that may help. 3) If you're fitting coordination numbers, then adding additional scattering shells with some physically defensible scheme for constraining coordination numbers to a small number of parameters can help a lot. 4) Another good technique is to fit multiple samples simultaneously, constraining S02 to be the same for all of them. Or fit the sample and a standard measured in a similar way simultaneously, again constraining S02 to be the same. 5) Along the same lines, you could fit a standard measured in a similar way to determine S02, and then constrain the fit of your sample to take on that value. 4 and 5 are similar, so you may wonder if I have a preference. I'd say that if the samples, beam, detectors, data, and data reduction are all well behaved, then #5 is probably best, and has the benefit of being a technique with a long pedigree. If you're a little suspicious of something in the chain, though (for example, it's difficult to tell if you've been consistent in normalizing your standard and sample, because one has a big white line and the other doesn't), then #4 has the benefit that it distributes the error in the parameters you are fitting between sample and standard. This is good both because your sample has less error than otherwise, and because the values for the standard act as a "canary in a coal mine," warning you by their deviation from known values as to the magnitude of the errors you're looking at. --Scott Calvin Sarah Lawrence College On Jul 31, 2010, at 11:56 AM, Gavin Garside wrote:
Scott,
Thank you for a quick response. The value I am getting for SO2 in the fit most of interest is 2.95 plus/minus 3.72. So with the error bar I am in range, but I was just suspicious of it before I make any claims about it. All my experiments were done in florescence because we have ordered bulk material. By creating a sample that would work in fluorescence I may have introduced dislocations or imperfections that would have effected the physical properties of interest in this sample.
Gavin Garside University of Utah
From: Scott Calvin
To: XAFS Analysis using Ifeffit Sent: Sat, July 31, 2010 4:52:35 AM Subject: Re: [Ifeffit] Large Amplitude Values Hi Gavin,
What are the uncertainties on the high S02 values?
Fluorescence is unlikely to be the culprit. While it can affect your ability to normalize properly, you're unlikely to account for a factor of 2 by normalization if the data is relatively decent. And self-absorption tends to suppress S02, not exaggerate it.
Why did you switch to fluorescence on just the handful of data sets? That might provide us a clue.
--Scott Calvin Sarah Lawrence College
On Jul 30, 2010, at 10:47 PM, Gavin Garside wrote:
Fellow X-Ray Absorption Enthusiasts,
I have recently compiled a model that gives excellent visual fits in R, q, and k space for bond spacing in a BCC structure. This model gives bond spacings that make sense, and are very close to what would be expected from this set. The R factors are very low, and the enot values correspond quite well to the edge. However, our amplitude values are much larger than typically expected. They come in at the range of 1.8 up to 5.0, but only on a few data sets. On all the rest the amplitude values are 0.4 to 1.0. Could this increase in amplitude be attributed to the fact that we ran florescence measurements instead of transmission, and have a weaker signal coming to the detector? What else could be causing this in only one data set? All samples used in this model have the same structure. Thanks in advance to any replies, your help and time is appreciated.
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
participants (2)
-
Gavin Garside
-
Scott Calvin