Hi Bhoopesh,
I agree that high energy sites will be occupied by Cd first. Upon
saturation, the cations will move to low energy sites. That being said, the
Cd loading in Sample 1 was higher than Sample 2. The higher amplitude in
Sample 1 FTmag could be due to presence of more Cd loading.
I have not try Cd-Na presuming low Z would have weak scattering effect. I
should try fitting that.
I will need to go back to Athena to look into the k-range for FT (originally
2:10 Å-1 for both samples). After doing some comparison, S2 k-space data
beyond 9 Å-1 seems to be unreliable to me now.
Anyway, I greatly appreciate your advice. Thank you.
Alan
On 11 January 2011 00:57, Bhoopesh Mishra
Hi Alan, Looking at the real part of FT, I am convinced that O atom will justify for peaks at 1.8 and 2.3 A in sample 1. One of the possibilities for sample 2 to be different is, it has higher ss2 for Cd-O (as can be seen from lower amplitude of first peak). If Cd loading in sample 1 is lower than sample 2, then it makes sense to me that Cd is bound to higher affinity (and less disordered) sites on Sodium Titanate nanotube in the sample1. As the Cd loading increases, Cd starts going to lower affinity albeit more disordered sites, making Cd-O bonding more disordered than sample 1.
Other possibilities include coexistence of O and Na in the first shell of sample 2, which might be interfering destructively and dampening the peak at 2.3 A). Have you tried that scenario? It might not be trivial to distinguish O bonding with high ss2 from O and Na in the first shell. But you can try that out by splitting the first shell at two different distances with smaller ss2 values. It is not obvious to me that your data supports 2nd shell Cd-Ti interaction. In the fit you described, your E_not2 and ss_2 are very high. The amp_2 is highly correlated with ss_2, and amp_2 value is close to its error bar. Put together, these two parameters makes me suspicious of your 2nd shell fit. Your Chi data does not necessarily show clear Cd-Ti interaction either. In my opinion Cd-Ti interaction would result in high amplitude of oscillation in chi (and correspondingly strong peak in FT). However, things can behave differently in case of nanomaterials and you might have some contribution of Ti in your spectra. If this is a significant part of your manuscript, you will have to convince the reviewers.
Good Luck, Bhoopesh
On Sun, Jan 9, 2011 at 11:43 PM, Alan Du
wrote: Hi Bhoopesh and Scott,
I should have given a description of my project. Yes Scott, the work is to investigate the binding mechanisms of aqueous cadmium onto sodium titanate nanotubes. Spectrum of Sample 1 and 2 obtained from merging 9 scans and 4 scans, respectively.
A quick check in Athena and, indeed, the white line of Samples are higher than CdO. I'm not sure the reason behind it though. It is likely that cadmium binds to the surface of substrate rather than inside the bulk. The lack of distinct peaks after 1.8 Å means that there are not many scatters around the absorber?
Bhoopesh, as requested, I have attached the real part of FT ( http://img585.imageshack.us/i/ftreal.jpg/). I haven't got a chance to interpret them.
From preliminary fitting of Sample 1, the major and minor peaks at 1.8 and 2.3 Å could be described by a Cd-O path (CdO). This interests me because Sample 2 does not have a peak at 2.3 Å, meaning there is another single scattering path for Sample 2?.
The peaks at 3 Å were fitted with Cd-Ti path (CdTiO3). No multiple scattering paths used. The best fit goes something like this:
**************************************************************************** Independent points = 13.166992187 Number of variables = 8.000000000 Chi-square = 1534.709946959 Reduced Chi-square = 297.021921317 R-factor = 0.000128095 Measurement uncertainty (k) = 0.000060423 Measurement uncertainty (R) = 0.004455442 Number of data sets = 1.000000000
Guess parameters +/- uncertainties (initial guess): amp = 0.9242390 +/- 0.0509920 (1.0000) enot = 1.3950420 +/- 0.5425380 (0.0000) delr = -0.0872060 +/- 0.0051060 (0.0000) ss = 0.0113250 +/- 0.0008480 (0.0030) amp_2 = 0.2441320 +/- 0.1905250 (1.0000) enot_2 = 22.5261260 +/- 4.5990590 (0.0000) delr_2 = 0.2510860 +/- 0.0719080 (0.0000) ss_2 = 0.0274690 +/- 0.0128040 (0.0030)
Correlations between variables: amp_2 and ss_2 --> 0.9342 enot_2 and delr_2 --> 0.9133 amp and ss --> 0.8865 enot and delr --> 0.8632 amp_2 and delr_2 --> 0.3040 delr_2 and ss_2 --> 0.2888 All other correlations are below 0.25
k-range = 2.000 - 9.000 dk = 1.000 k-window = hanning k-weight = 3 R-range = 1.000 - 4.000 dR = 0.000 R-window = hanning fitting space = R background function = none phase correction = none
R-factor for this data set = 0.00270
***********************************************************************************
The above enot_2 is on the high side. I am not entirely familiar with the parameters yet. Are there other parameters I should worry about?
Cheers,
Alan J. DU Nanyang Technological University, Singapore
-- Alan J. DU Research Student Nanyang Technological University, Singapore School of Civil & Environmental Engineering Blk N1, N1-B4b-04, 50 Nanyang Avenue, Singapore 639798 Mobile: +65 9621 6991 Fax: +65 6861 5254