Dear Mengling,
The basic issues are: will there be enough Tl fluorescence counts in
the detector, and will you be able to see a decent edge. From you
spreadsheet of concentrations, the elements that stand out are: Fe
and to a lesser extent Zn, and Pb.
The Fe and Zn will dominate the fluorescence getting to your detector.
This hurts sensitivity to Tl because the solid-state detectors can
handle a limited flux themselves (typically on the order of 100K Hz).
The high Fe concentration means the Fe Kalpha will dominate the
emission when you shine 12.7keV x-rays on the sample. The
simple-but-effective remedy is to put aluminum foil on the detector,
as aluminum will absorb Fe fluorescence (at 6.4keV) about 4x more than
the Tl fluorescence (at 10.3keV). You might need a few hundred
microns of aluminum foil, and may need to try a few different amounts
to cut down the Fe counts sufficiently.
The other issue is Pb. The Tl L3 edge is at 12.658keV. once you hit
the Pb L3 edge, at 13.1 keV, the Pb Lalpha fluorescence will dominate
the Tl Lalpha fluorescence -- these lines (10.3 and 10.5 keV,
respectively) are too close together to effectively filter
electronically. Even well below the Pb L3 edge, you'll start to see
some Pb Lalpha fluorescence. Since the Pb concentration might be 100x
the Tl concentration, that would be something to look for, but there's
not a lot that can be done about it. Anyway, that really only limits
the EXAFS, not so much the ability to get XANES. The Tl
concentrations are low, but I think that at a good APS BM line like
5-BM, getting Tl L3 XANES is just a matter of time.
Hope that helps,
--Matt
On Thu, Sep 29, 2011 at 9:39 AM, Mengling Yi Stuckman
Dear Ifeffit Community,
I am currently working on project "Evaluating Speciation of Thallium in Coal Fly Ash". The project was accepted by APS and we will conduct our measurement on beamline 5-BM-D.
My fly ash sample has Tl concentration ranging from 0.5ppm to 13ppm. How can I know whether the sample can generate enough photon counts if using fluorescence mode? what would be the appropriate sample preparation procedure I should follow? Is there any interference I should pay attention to?
I understand that I should use Hephaestus, but I don't know how to do that if my sample contains lots of different oxides. I attached the elemental composition in this email. Most metals should be in form of oxides, or sulfates.
Any input would be highly appreciated,
Mengling
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit