[Ifeffit] Reported W L3-edge and L2-edge energy

Matthew Marcus mamarcus at lbl.gov
Thu May 7 17:17:20 CDT 2020


For elements like P and S, people often use the energies of peaks. 
These are more immune to noise, energy-resolution effects and 
overabsorption than inflection points are.  For instance, on ALS 10.3.2, 
I used the sulfate peak of gypsum set at 2482.74eV.  I forget where I 
got that number.  Going down to soft X-rays, a common convention for the 
carbon edge is to use a pair of sharp peaks in CO2 gas at 292.74 and 
294.96eV.
	mam

On 5/7/2020 3:09 PM, Mike Massey wrote:
> Hi Matt,
> 
> 
> Indeed, in my experience (which is limited to one beamline at one 
> synchrotron facility for P XAS), once it is calibrated, the energy 
> selection tends to be quite stable, so I think you're on-target there.
> 
> The trouble I still run into, though, is comparability of data between 
> studies. The difficulty is magnified by the fact that people tend to 
> identify certain near-edge features by the energy range at which they 
> occur. I do the same, of course, but I also try to carefully document 
> the material and energy I used to calibrate the monochromator.
> 
> For the P K-edge, it doesn't really seem like people have settled on a 
> convention for calibrating the monochromator, unlike in the case of 
> iron, for example (where one just uses a foil and sets some feature of 
> that spectrum to their preferred value). If everyone was using the same 
> thing all would be happy, but most people use different materials and 
> different values. So datasets for P at the K-edge really aren't too 
> comparable just yet.
> 
> Sorry to hijack the conversation, it's just an issue I've been mulling 
> over for a few years. The discussion of energy calibration values made 
> me think of it again.
> 
> 
> Best,
> 
> 
> 
> Mike
> 
> 
> 
> 
>> On May 8, 2020, at 8:51 AM, Matt Newville <newville at cars.uchicago.edu> 
>> wrote:
>>
>> 
>> Hi Mike,
>>
>>
>> On Tue, May 5, 2020 at 10:56 PM Mike Massey <mmassey at gmail.com 
>> <mailto:mmassey at gmail.com>> wrote:
>>
>>     On a tangentially related topic, I find that phosphorus K-edge XAS
>>     energy calibration conventions are still in a bit of a "Wild West"
>>     state, with a wide variety of materials and values in use for
>>     energy calibration. As an extreme example, one or two frequently
>>     cited papers in my field from the 2000s don't even report the
>>     material or value used for energy calibration, and only show
>>     portions of the spectra on an energy axis with values relative to
>>     an unknown E0.
>>
>>
>> I have never measured a P K edge, or indeed any edge lower in energy 
>> than the S K edge (ignoring some X-ray raman work).  But if one is 
>> using a Si(111) double-crystal monochromator where P or S is 
>> approximately the low-energy (high-angle) limit, then it really should 
>> be that the calibration does not drift much and cannot be too wrong at 
>> low energies.
>>
>> That is, a mono calibration is controlled by a d-spacing and angular 
>> offset. Normally (or perhaps, in my experience), "re-calibrating" is 
>> done by changing the angular offset, leaving the d-spacing alone.  
>> That is, the d-spacing is presumably known, at least to within some 
>> thermal drift.
>> If that is the case that the d-spacing really is not changing and what 
>> needs to be refined is the angular offset, then setting the offset at 
>> relatively high energy edges will be much more sensitive, and changing 
>> the angular offset to that a high-energy edge is correct should move 
>> lower energy edges by a smaller amount.   The corollary is that you 
>> have to move the offset a lot to move the P  K edge around, and that 
>> would have a larger (and ever-increasing) impact on higher energy 
>> edges such as Ca, Fe, Cu or Mo.
>>
>> The counter-argument is also true:  d-spacing has a bigger effect on 
>> the high-angle / low-energy edges.
>>
>> So, if you believe the mono d-spacing (or you believe the beamline 
>> scientist who believes it ;)) then calibrate at the highest energy you 
>> can.   The Kraft values don't go very low in energy.
>>
>> All that said, if using a different mono crystal such as InSb or more 
>> exotic crystals, I have no idea how stable those are.
>>
>>
>>     I too have picked my own material and value, and will be the first
>>     to acknowledge that I did so out of necessity and ease of
>>     comparison to other available data, rather than because I thought
>>     it was correct.
>>
>>     The issue of calibration conventions and values definitely seems
>>     to be one that merits continued discussion. It has been
>>     interesting to watch things evolve over time in the case of iron,
>>     for example (it's good to know that 7110.75 is a candidate
>>     calibration value...) I appreciate Matt's detailed thoughts, and
>>     the data that he's been working with. Thanks Matt!
>>
>>
>>     Cheers,
>>
>>
>>
>>     Mike
>>
>>
>>
>>
>>
>>>     On May 6, 2020, at 3:32 PM, Matt Newville
>>>     <newville at cars.uchicago.edu <mailto:newville at cars.uchicago.edu>>
>>>     wrote:
>>>
>>>     
>>>     Hi Simon,
>>>
>>>     This is definitely a timely discussion for me, as I've been
>>>     spending part of the quartine working on collating data and
>>>     expanding datasets for an XAFS spectral database.  I'm hoping to
>>>     have something ready for public comment and to start asking for
>>>     contributions of data in a few weeks, but I'll be happy to have
>>>     more discussion about that sooner too.
>>>
>>>     I generally believe that the monochromator I use at GSECARS is
>>>     both well-calibrated and reasonably accurate.  That is, with 2
>>>     angular encoders with a resolution of >130,000 lines per degree
>>>     and an air-bearing, I believe the angular accuracy and
>>>     repeatability are very good.  I believe there are equally good
>>>     moons in existence.   As Matthew Marcus pointed to the Kraft
>>>     paper (which used an older source but 4-bounce mono to improve
>>>     resolution), we find that Fe foil is definitely better defined as
>>>     7110.75 and Cu foil is between 8980.0 and 8980.5 eV.  That is,
>>>     we've measured multiple foils, found their first derivatives, and
>>>     refined the d-spacing and angular offset.  We do this about once
>>>     per run, and the offsets tend to be very consistent.   For sure,
>>>     there is some question about whether the Kraft numbers are
>>>     perfect.   For sure, putting Fe foil at 7110.75 +/- 0.25 eV
>>>     appears to be "most right" to us.
>>>
>>>     I also believe that we should probably re-measure these metal
>>>     foils (and other compounds) with a single calibration set for
>>>     both Si(111) and Si(311).  We will probably have time to do that
>>>     this summer in the time between "beamline staff can get back to
>>>     the beamline" and "open for outside users".
>>>
>>>     What I can tell you now is:  I have some data on W metal, WO2,
>>>     and WO3 measured all at the same time on our bending magnet line,
>>>     with Si(111).  An Athena project for this is attached (W.prj). 
>>>      I cannot vouch for the absolute calibration.
>>>
>>>     I also attach a set of foils (V, Fe, Cu, Mo) measured with the
>>>     same calibration (and Si(111) on our ID line), after adjusting
>>>     d-space and offset to be close to the Kraft values
>>>     (CalibratedFoils2013.prj).
>>>
>>>     I also attach a set of foils (Fe, Cu, Au L3, Au L2, Au L1, Pb L3,
>>>     Pb L2, Pb L1 edges) measured in 2016 (again, using Si(111) on our
>>>     ID line), also with the same calibration values
>>>     (FeCu_Au_Pb.prj).  I'm pretty certain these use the same
>>>     d-spacing as the 2013 Foils to at least 5 digits.   For
>>>     completeness, all of the raw data files are also under
>>>     https://github.com/XraySpectroscopy/XASDataLibrary/tree/master/data
>>>
>>>     In my experience, the Pb L3 edge value has the biggest variation
>>>     in the literature, with values ranging from 13035 to 13055 eV
>>>     (possibly a typo somewhere along the line).  Fortunately, the
>>>     Kraft-based calibration splits the difference and puts the value
>>>     at 13040 eV.
>>>
>>>     For W in particular, I will look if I have measured this recently
>>>     on our ID line.  I can tell you that I use CdWO4 as a phosphor
>>>     and use that to focus our X-ray beam.   I use this trick all the
>>>     time: any tail from the beam penetrating the phosphor is shortest
>>>     at the peak of the white-line and for CdWO4 that is always
>>>     between 10210 and 10215 eV.
>>>
>>>     I hope that helps.  I am interested in trying to get all these
>>>     values as accurately as possible, so any comments or suggestions
>>>     would be most welcome.
>>>
>>>     --Matt
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>     On Tue, May 5, 2020 at 5:14 PM Bare, Simon R
>>>     <srbare at slac.stanford.edu <mailto:srbare at slac.stanford.edu>> wrote:
>>>
>>>         All:____
>>>
>>>         __ __
>>>
>>>         We are wondering if others agree that the reported values for
>>>         the W L3 and W L2 edges are *incorrect*. We recently noticed
>>>         the following:____
>>>
>>>         __ __
>>>
>>>         The “Edge” – defined by the inflection point of the
>>>         absorption edge step____
>>>
>>>         __ __
>>>
>>>         When using the Ir L_3 edge (11215.0 eV) as a calibration, the
>>>         W L_3 - and L_2 -edges are *10203.4 eV* and *11542.4 eV*,
>>>         respectively. ____
>>>
>>>         __ __
>>>
>>>         When using the Pt L_3 edge (11564.0 eV) as a calibration, the
>>>         W L_3 - and L_2 -edges are *10203.3 eV* and *11542.4 eV*,
>>>         respectively.____
>>>
>>>         __ __
>>>
>>>         These observations are thus different than the reported
>>>         values of *10207.0 eV* and *11544.0 eV* for the L_3 and L_2
>>>         edges, respectively.____
>>>
>>>         __ __
>>>
>>>         Thanks in advance for the discussion and feedback.____
>>>
>>>         __ __
>>>
>>>         __ __
>>>
>>>         Simon R Bare____
>>>
>>>         /Distinguished Scientist____/
>>>
>>>         /SSRL, MS69____/
>>>
>>>         /SLAC National Accelerator Lab____/
>>>
>>>         /2575 Sand Hill Road____/
>>>
>>>         /Menlo Park CA 94025____/
>>>
>>>         __ __
>>>
>>>         simon.bare at slac.stanford.edu
>>>         <mailto:simon.bare at slac.stanford.edu>____
>>>
>>>         Ph: 650-926-2629____
>>>
>>>         __ __
>>>
>>>         <image001.png>
>>>         ____
>>>
>>>         __ __
>>>
>>>         _______________________________________________
>>>         Ifeffit mailing list
>>>         Ifeffit at millenia.cars.aps.anl.gov
>>>         <mailto:Ifeffit at millenia.cars.aps.anl.gov>
>>>         http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>>>         Unsubscribe:
>>>         http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>>>
>>>
>>>
>>>     -- 
>>>     --Matt Newville <newville at cars.uchicago.edu
>>>     <http://cars.uchicago.edu>> 630-252-0431
>>>     <W.prj>
>>>     <CalibratedFoils2013.prj>
>>>     <FeCu_AuPb.prj>
>>>     _______________________________________________
>>>     Ifeffit mailing list
>>>     Ifeffit at millenia.cars.aps.anl.gov
>>>     <mailto:Ifeffit at millenia.cars.aps.anl.gov>
>>>     http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>>>     Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>>     _______________________________________________
>>     Ifeffit mailing list
>>     Ifeffit at millenia.cars.aps.anl.gov
>>     <mailto:Ifeffit at millenia.cars.aps.anl.gov>
>>     http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>>     Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>>
>>
>>
>> -- 
>> --Matt Newville <newville at cars.uchicago.edu 
>> <http://cars.uchicago.edu>> 630-252-0431
>> _______________________________________________
>> Ifeffit mailing list
>> Ifeffit at millenia.cars.aps.anl.gov
>> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>> Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
> 
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
> 


More information about the Ifeffit mailing list