> Hi Christian,
>
> On Thu, Feb 6, 2020, 8:40 PM Christian Wittee Lopes <chriswittee at gmail.com>
> wrote:
>
>> Dear all,
>>
>> Recently I was questioned about the EXAFS detection limit when describing
>> different metal species in a bimetallic sample.
>>
>> By checking Pd and Cu K-edges, for example, I found Pd metal
>> nanoparticles and CuO clusters, respectively. But additional techniques
>> tell me I can have copper atoms in intimate contact with the Pd
>> nanoparticles. What would be the minimum amount of these "single atoms"
>> needed to be detected by EXAFS? is there a detection limit or it depends on
>> several parameters?
>>
>
> As Chris Chantler says, there are a lot of things that can influence this,
> so there really isn't one simple answer. Also, as Chris says, advances in
> analytic methods have been (mostly) been improving the situation.
>
> At my beamline, we often get asked questions about detection limits.
> We're typically working in a different context than
> nanoparticles/catalysts, but I think the basic ideas are about the same.
>
> A good starting rule-of-thumb for absolute detection limits is 1 ppm by
> atomic weight. You might be able to do better sometimes, but there are
> situations where XANES at 10 ppm is very hard. For sure, a matrix of
> light elements is much better than a matrix of heavy elements.
>
> For very dilute samples, one will be using fluorescence XAFS measurements
> with a solid-state detector or know very well why you are doing something
> different. These solid-state detectors and electronics are fundamentally
> limited to have energy resolutions of ~120 eV (often 250 eV) and maximum
> total count rates of 5 MHz (often 0.5 MHz). Many beamlines use "a
> handful" (2 to 16) parallel detectors, and some have up to 100 (but often
> with each having a lower individual maximum count rate, and perhaps
> less-than-ideal energy resolution).
> With a count rate of a few MHz total and a sample with 1ppm of "element of
> interest", the elastic and Compton scattering and/or fluorescence from
> other elements will dominate that total count rate and the energy
> resolution will give non-zero background in the fluorescence spectrum.
> That means that even seeing a peak from 1 ppm of an element in an X-ray
> fluorescence spectrum with a solid-state detector is challenging. Not
> impossible, but definitely not routine.
>
> For sure, adding more detectors or counting for a long time can help. But
> those are linear in time and the number of detectors (and no beamline has
> 1000 parallel detectors). Low Z matrices like water, biological material,
> carbon-rich materials are easier. Samples with nearby or overlapping
> fluorescence lines are much harder. That is 10 ppm Zn in water: yes, 1
> ppm Zn in water: maybe, 10 ppm Zn in CaCO3: maybe, 100 ppm Zn in Cu metal:
> no. For sure, XANES at 1ppm is sometimes possible. Getting interpretable
> XAFS would take a lot longer, perhaps days of counting.
>
> Using filters and/or Bent Laue Analyzers in front of a solid-state (or
> integrating) detector can sometimes help to eliminate the unwanted scatter
> signals before they get to the solid-state detector. Using crystal
> analyzers ("wavelength" vs "energy" dispersive fluorescence) can help -
> they have lower backgrounds and are not limited by the total scatter - but
> the solid angle for these tend to be small. Using crystal analyzer arrays
> are probably really needed to get the best detection limits. A few
> beamlines do regularly do HERFD analysis with arrays of crystal analyzers,
> and many of the rest of us are trying to catch up. Still, I believe that
> "1 ppm" is around the state of the art, if not "heroic".
>
> All of that is for the detection limit of an atomic species. If you are
> asking about detecting Cu in/on/with Pd nanoparticles with CuO also
> present, the answer is far worse. Cu XAS measurements will be an average
> of all Cu atoms in the illuminated volume -- you cannot avoid the CuO.
> Seeing that 1% of the Cu atoms are bound to Pd and not to oxygen would be
> very challenging.
>
> Hope that helps,
> --Matt
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
>
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20200209/6dab4f1e/attachment.html>
More information about the Ifeffit
mailing list