[Ifeffit] Fe K edge spectrum of scorodite
Matthew Marcus
mamarcus at lbl.gov
Mon Jul 6 11:47:16 CDT 2015
Here are my versions of processed files, made from what you sent. The .e is the pre-edge removed, post-edge normalized, shifted so that the Fe K-edge is at 7110.75eV.
The .b is k^3*chi(k), with E0 set so that it comes halfway up the edge jump. The .chi is the chi(k), with the abscissa interpolated to a regular grid, the way Artemis likes it.
mam
On 7/6/2015 9:12 AM, Case Van Genuchten wrote:
> Dear Matthew,
>
> Sure, the raw data and/or spectrum can be added to the database. For this sample, the 1st derivative of the Fe foil was 7.112 keV. I can also provide individual scans for the database, if needed. Is there any interest in the As K-edge of this sample, too? I have that one as well.
> Case
>
> On Mon, Jul 6, 2015 at 5:48 PM, Matthew Marcus <mamarcus at lbl.gov <mailto:mamarcus at lbl.gov>> wrote:
>
> Thanks! May I add it to the ALS 10.3.2 reference collection?
>
> What is the standard of energy calibration? Is the Fe metal edge at 7110.75eV, 7112eV or what?
> mam
>
> On 7/6/2015 7:11 AM, Case Van Genuchten wrote:
>
> Dear Shaofeng,
>
> I am attaching the averaged, raw data and Fe K-edge EXAFS spectrum (extracted using SixPack) of scorodite. The mineral was prepared following the procedure outlined in Paktunc et al. (2008). The data were collected at the Stanford Synchrotron Radiation Lightsource in transmission mode with the sample housed in a liquid nitrogen cryostat. You can find additional details on the preparation of the sample and data collection from the publication where the spectrum is reported and modeled (van Genuchten et al. 2014, Geochimica et Cosmochimica Acta).
>
> Please let me know if you have any other questions about data collection.
>
> Best,
>
> Case M. van Genuchten
> Environmental Geochemistry Laboratory
> Institute of Earth Surface Dynamics
> University of Lausanne, Lausanne, CH
>
>
>
> On Mon, Jul 6, 2015 at 4:02 AM, Shaofeng Wang <wangshaofeng at iae.ac.cn <mailto:wangshaofeng at iae.ac.cn> <mailto:wangshaofeng at iae.ac.cn <mailto:wangshaofeng at iae.ac.cn>>> wrote:
>
> Dear all,
> Can anyone have Fe K edge exafs spectrum of scorodite share a copy?
> Thanks in advance.
> Shaofeng
> --------------------------------------
> Shaofeng Wang, Ph.D of Geochemistry
> Environmental Molecular Science Group
> Institute of Applied Ecology, Chinese Academy of Sciences
> Shenyang, 110016, China
> wangshaofeng at iae.ac.cn <mailto:wangshaofeng at iae.ac.cn> <mailto:wangshaofeng at iae.ac.cn <mailto:wangshaofeng at iae.ac.cn>>
> www.iae.cas.cn <http://www.iae.cas.cn> <http://www.iae.cas.cn>
>
>
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov <mailto:Ifeffit at millenia.cars.aps.anl.gov> <mailto:Ifeffit at millenia.cars.aps.anl.gov <mailto:Ifeffit at millenia.cars.aps.anl.gov>>
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>
>
>
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov <mailto:Ifeffit at millenia.cars.aps.anl.gov>
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov <mailto:Ifeffit at millenia.cars.aps.anl.gov>
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>
>
>
>
> _______________________________________________
> Ifeffit mailing list
> Ifeffit at millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
>
-------------- next part --------------
6.8807500E+3 -5.6686879E-4
6.8907510E+3 4.0188098E-4
6.9007500E+3 1.9576439E-4
6.9107500E+3 6.6377126E-4
6.9207510E+3 7.3053432E-4
6.9307520E+3 4.2579199E-4
6.9407500E+3 5.1885990E-4
6.9507480E+3 8.5548572E-4
6.9607510E+3 1.1406525E-3
6.9707500E+3 1.1071908E-3
6.9807480E+3 9.7453779E-4
6.9907510E+3 9.5817003E-4
7.0007490E+3 9.1372709E-4
7.0107520E+3 1.3349763E-3
7.0207490E+3 1.0275797E-3
7.0307490E+3 1.3605076E-3
7.0407490E+3 1.6912761E-3
7.0507480E+3 1.6655628E-3
7.0607500E+3 2.3310543E-3
7.0707500E+3 3.0608769E-3
7.0807490E+3 3.4280701E-3
7.0907510E+3 4.6546218E-3
7.0911001E+3 5.1562926E-3
7.0914482E+3 5.1379598E-3
7.0918018E+3 5.1800186E-3
7.0921509E+3 5.1840505E-3
7.0924990E+3 4.9898361E-3
7.0928481E+3 5.0541580E-3
7.0932021E+3 5.1741299E-3
7.0935508E+3 5.1947828E-3
7.0938999E+3 5.5255872E-3
7.0942490E+3 5.5841443E-3
7.0945981E+3 5.4232348E-3
7.0949521E+3 5.7453403E-3
7.0953008E+3 5.6040450E-3
7.0956499E+3 5.8876155E-3
7.0959990E+3 5.8254543E-3
7.0963481E+3 6.0572238E-3
7.0966982E+3 6.2847678E-3
7.0970518E+3 6.5249893E-3
7.0974009E+3 6.7258098E-3
7.0977500E+3 6.5698550E-3
7.0981001E+3 6.4360630E-3
7.0984492E+3 6.5597185E-3
7.0987988E+3 6.7354134E-3
7.0991479E+3 6.7586658E-3
7.0994980E+3 7.1900967E-3
7.0998521E+3 7.3942537E-3
7.1002021E+3 7.2445427E-3
7.1005518E+3 7.2462349E-3
7.1009009E+3 7.3070318E-3
7.1012510E+3 8.0021702E-3
7.1016011E+3 7.8822904E-3
7.1019512E+3 8.0508466E-3
7.1023008E+3 8.2164365E-3
7.1026509E+3 8.3241699E-3
7.1030010E+3 8.5529276E-3
7.1033511E+3 8.4430149E-3
7.1037012E+3 8.8325896E-3
7.1040508E+3 9.0436364E-3
7.1044009E+3 9.1322768E-3
7.1047510E+3 9.2499360E-3
7.1051011E+3 9.7290400E-3
7.1054512E+3 1.0202789E-2
7.1058018E+3 1.0184573E-2
7.1061519E+3 1.0412129E-2
7.1065020E+3 1.0559413E-2
7.1068481E+3 1.0903092E-2
7.1071982E+3 1.1292792E-2
7.1075488E+3 1.1254787E-2
7.1078989E+3 1.1751585E-2
7.1082500E+3 1.1875168E-2
7.1086001E+3 1.2481090E-2
7.1089512E+3 1.3308561E-2
7.1093018E+3 1.3569955E-2
7.1096519E+3 1.4215299E-2
7.1099980E+3 1.4651619E-2
7.1103491E+3 1.5519120E-2
7.1107002E+3 1.6379659E-2
7.1110508E+3 1.7498053E-2
7.1114019E+3 1.9280512E-2
7.1117480E+3 2.1458479E-2
7.1120991E+3 2.5736959E-2
7.1124502E+3 3.2488927E-2
7.1128008E+3 4.0696480E-2
7.1131519E+3 4.3416918E-2
7.1134980E+3 4.2018272E-2
7.1138491E+3 4.3228885E-2
7.1142002E+3 4.6393724E-2
7.1145518E+3 4.4920125E-2
7.1148979E+3 3.9243296E-2
7.1152490E+3 3.4286267E-2
7.1156011E+3 3.1454499E-2
7.1159521E+3 3.0908519E-2
7.1162988E+3 3.1535559E-2
7.1166499E+3 3.2785327E-2
7.1170020E+3 3.5014785E-2
7.1173481E+3 3.7418321E-2
7.1177002E+3 4.0268283E-2
7.1180518E+3 4.4214300E-2
7.1183979E+3 4.9028349E-2
7.1187500E+3 5.5243193E-2
7.1191021E+3 6.3646648E-2
7.1194492E+3 7.2755320E-2
7.1197998E+3 8.4100138E-2
7.1201519E+3 9.6376169E-2
7.1204990E+3 1.0890239E-1
7.1208511E+3 1.2241849E-1
7.1211982E+3 1.3763460E-1
7.1215498E+3 1.5663065E-1
7.1218979E+3 1.8100515E-1
7.1222500E+3 2.1084773E-1
7.1226021E+3 2.4466526E-1
7.1229492E+3 2.7948362E-1
7.1233008E+3 3.1435361E-1
7.1236489E+3 3.4792411E-1
7.1240010E+3 3.8097774E-1
7.1243491E+3 4.1431893E-1
7.1247012E+3 4.4986482E-1
7.1250479E+3 4.8817377E-1
7.1254009E+3 5.3069557E-1
7.1257480E+3 5.7753898E-1
7.1261011E+3 6.3244947E-1
7.1264492E+3 6.9484979E-1
7.1268008E+3 7.6658149E-1
7.1271489E+3 8.4516995E-1
7.1275020E+3 9.3127267E-1
7.1278501E+3 1.0196375E+0
7.1282021E+3 1.1124751E+0
7.1285498E+3 1.2030849E+0
7.1288979E+3 1.2862571E+0
7.1292510E+3 1.3565443E+0
7.1295991E+3 1.4068305E+0
7.1299521E+3 1.4423019E+0
7.1302998E+3 1.4691341E+0
7.1306479E+3 1.4920315E+0
7.1310010E+3 1.5116633E+0
7.1313501E+3 1.5253123E+0
7.1316982E+3 1.5302798E+0
7.1320508E+3 1.5261334E+0
7.1323989E+3 1.5133530E+0
7.1327480E+3 1.4944884E+0
7.1331011E+3 1.4721429E+0
7.1334492E+3 1.4500458E+0
7.1337979E+3 1.4292176E+0
7.1341509E+3 1.4120328E+0
7.1345000E+3 1.3989839E+0
7.1348491E+3 1.3896494E+0
7.1352021E+3 1.3855464E+0
7.1355508E+3 1.3863191E+0
7.1358999E+3 1.3903745E+0
7.1362480E+3 1.3963094E+0
7.1366021E+3 1.4016349E+0
7.1369512E+3 1.4042144E+0
7.1372998E+3 1.4040196E+0
7.1376489E+3 1.3995903E+0
7.1379980E+3 1.3923826E+0
7.1383511E+3 1.3818878E+0
7.1387002E+3 1.3695264E+0
7.1390498E+3 1.3552926E+0
7.1393989E+3 1.3403201E+0
7.1397480E+3 1.3251940E+0
7.1401021E+3 1.3091658E+0
7.1404512E+3 1.2925683E+0
7.1407510E+3 1.2769457E+0
7.1411250E+3 1.2573849E+0
7.1417598E+3 1.2220012E+0
7.1424189E+3 1.1854193E+0
7.1430942E+3 1.1508599E+0
7.1437881E+3 1.1202539E+0
7.1445020E+3 1.0939228E+0
7.1452358E+3 1.0721081E+0
7.1459902E+3 1.0532905E+0
7.1467632E+3 1.0369454E+0
7.1475518E+3 1.0205347E+0
7.1483608E+3 1.0037559E+0
7.1491890E+3 9.8493877E-1
7.1500381E+3 9.6441562E-1
7.1509058E+3 9.4406507E-1
7.1517900E+3 9.2380230E-1
7.1526929E+3 9.0454361E-1
7.1536221E+3 8.8431867E-1
7.1545601E+3 8.6395840E-1
7.1555239E+3 8.4384721E-1
7.1565078E+3 8.2560447E-1
7.1575059E+3 8.1182733E-1
7.1585249E+3 8.0359363E-1
7.1595640E+3 7.9981878E-1
7.1606191E+3 8.0034802E-1
7.1616982E+3 8.0443086E-1
7.1627920E+3 8.1168026E-1
7.1639072E+3 8.2106291E-1
7.1650420E+3 8.3386074E-1
7.1661919E+3 8.4926217E-1
7.1673628E+3 8.6709007E-1
7.1685542E+3 8.8651611E-1
7.1697651E+3 9.0773016E-1
7.1709912E+3 9.2882447E-1
7.1722432E+3 9.5027891E-1
7.1735049E+3 9.7216458E-1
7.1747930E+3 9.9717112E-1
7.1760962E+3 1.0182272E+0
7.1774238E+3 1.0309371E+0
7.1787632E+3 1.0423678E+0
7.1801279E+3 1.0515047E+0
7.1815078E+3 1.0575832E+0
7.1829082E+3 1.0621943E+0
7.1843291E+3 1.0655866E+0
7.1857661E+3 1.0687883E+0
7.1872231E+3 1.0718619E+0
7.1887012E+3 1.0760759E+0
7.1901948E+3 1.0801393E+0
7.1917090E+3 1.0831919E+0
7.1932441E+3 1.0844331E+0
7.1947949E+3 1.0835607E+0
7.1963662E+3 1.0804234E+0
7.1979590E+3 1.0758645E+0
7.1995669E+3 1.0695226E+0
7.2011958E+3 1.0609348E+0
7.2028462E+3 1.0494385E+0
7.2045122E+3 1.0361004E+0
7.2061982E+3 1.0224238E+0
7.2079009E+3 1.0095351E+0
7.2096250E+3 9.9756046E-1
7.2113701E+3 9.8721122E-1
7.2131309E+3 9.7762953E-1
7.2149121E+3 9.6950896E-1
7.2167109E+3 9.6362357E-1
7.2185298E+3 9.6025634E-1
7.2203701E+3 9.6002187E-1
7.2222271E+3 9.6188017E-1
7.2241050E+3 9.6549093E-1
7.2259990E+3 9.6980267E-1
7.2279141E+3 9.7462655E-1
7.2298462E+3 9.7941541E-1
7.2317998E+3 9.8366984E-1
7.2337739E+3 9.8619160E-1
7.2357598E+3 9.8678106E-1
7.2377729E+3 9.8636600E-1
7.2398018E+3 9.8629506E-1
7.2418481E+3 9.8567290E-1
7.2439150E+3 9.8498094E-1
7.2460039E+3 9.8468710E-1
7.2481089E+3 9.8615613E-1
7.2502310E+3 9.8943351E-1
7.2523750E+3 9.9339933E-1
7.2545361E+3 9.9849046E-1
7.2567178E+3 1.0025760E+0
7.2589180E+3 1.0071901E+0
7.2611392E+3 1.0121528E+0
7.2633770E+3 1.0167137E+0
7.2656318E+3 1.0211457E+0
7.2679092E+3 1.0260867E+0
7.2702031E+3 1.0315643E+0
7.2725190E+3 1.0355871E+0
7.2748521E+3 1.0394187E+0
7.2772080E+3 1.0427554E+0
7.2795752E+3 1.0441525E+0
7.2819702E+3 1.0431969E+0
7.2843770E+3 1.0405549E+0
7.2868062E+3 1.0363564E+0
7.2892520E+3 1.0300584E+0
7.2917212E+3 1.0231791E+0
7.2942080E+3 1.0155625E+0
7.2967119E+3 1.0080689E+0
7.2992378E+3 1.0018587E+0
7.3017822E+3 9.9658966E-1
7.3043428E+3 9.9239872E-1
7.3069219E+3 9.9014761E-1
7.3095239E+3 9.8945488E-1
7.3121431E+3 9.9039297E-1
7.3147798E+3 9.9104450E-1
7.3174399E+3 9.9204880E-1
7.3201182E+3 9.9352598E-1
7.3228130E+3 9.9509788E-1
7.3255259E+3 9.9723879E-1
7.3282578E+3 9.9908948E-1
7.3310122E+3 1.0010026E+0
7.3337852E+3 1.0015001E+0
7.3365752E+3 1.0021348E+0
7.3393828E+3 1.0021389E+0
7.3422148E+3 1.0012142E+0
7.3450601E+3 1.0013425E+0
7.3479282E+3 1.0013175E+0
7.3508149E+3 1.0011318E+0
7.3537202E+3 1.0012767E+0
7.3566431E+3 1.0010335E+0
7.3595850E+3 1.0014912E+0
7.3625498E+3 1.0022563E+0
7.3655278E+3 1.0021988E+0
7.3685308E+3 1.0023260E+0
7.3715518E+3 1.0024668E+0
7.3745908E+3 1.0025762E+0
7.3776440E+3 1.0037302E+0
7.3807212E+3 1.0049905E+0
7.3838169E+3 1.0073916E+0
7.3869312E+3 1.0094226E+0
7.3900649E+3 1.0113368E+0
7.3932168E+3 1.0127903E+0
7.3963940E+3 1.0144847E+0
7.3995840E+3 1.0139297E+0
7.4027939E+3 1.0134147E+0
7.4060220E+3 1.0121613E+0
7.4092700E+3 1.0105493E+0
7.4125371E+3 1.0082544E+0
7.4158232E+3 1.0062145E+0
7.4191289E+3 1.0030253E+0
7.4224541E+3 1.0001608E+0
7.4257930E+3 9.9725463E-1
7.4291572E+3 9.9452219E-1
7.4325410E+3 9.9250378E-1
7.4359380E+3 9.9133111E-1
7.4393608E+3 9.9035985E-1
7.4427979E+3 9.8992570E-1
7.4462549E+3 9.9041867E-1
7.4497319E+3 9.9208854E-1
7.4532280E+3 9.9308620E-1
7.4567388E+3 9.9577226E-1
7.4602759E+3 9.9784425E-1
7.4638271E+3 1.0000934E+0
7.4673979E+3 1.0020299E+0
7.4709888E+3 1.0036406E+0
7.4746001E+3 1.0045451E+0
7.4782319E+3 1.0045775E+0
7.4818789E+3 1.0044622E+0
7.4855459E+3 1.0030843E+0
7.4892329E+3 1.0020236E+0
7.4929351E+3 1.0007815E+0
7.4966582E+3 9.9891894E-1
7.5004019E+3 9.9850248E-1
7.5041660E+3 9.9812755E-1
7.5079448E+3 9.9745795E-1
7.5117510E+3 9.9724127E-1
7.5155669E+3 9.9755208E-1
7.5194092E+3 9.9709197E-1
7.5232661E+3 9.9793103E-1
7.5271392E+3 9.9900346E-1
7.5310381E+3 9.9950006E-1
7.5349531E+3 1.0007013E+0
7.5388838E+3 1.0017345E+0
7.5428359E+3 1.0026317E+0
7.5468091E+3 1.0027007E+0
7.5508032E+3 1.0032072E+0
7.5548081E+3 1.0033975E+0
7.5588398E+3 1.0031116E+0
7.5628882E+3 1.0025495E+0
7.5669580E+3 1.0017148E+0
7.5710430E+3 1.0003333E+0
7.5751450E+3 9.9912267E-1
7.5792690E+3 9.9760526E-1
7.5834141E+3 9.9582283E-1
7.5875762E+3 9.9462697E-1
7.5917598E+3 9.9365941E-1
7.5959600E+3 9.9369498E-1
7.6001768E+3 9.9368168E-1
7.6044150E+3 9.9370382E-1
7.6086768E+3 9.9463892E-1
7.6129492E+3 9.9634338E-1
7.6172490E+3 9.9772075E-1
7.6215601E+3 9.9924018E-1
7.6258940E+3 1.0000055E+0
7.6302500E+3 1.0014107E+0
7.6346182E+3 9.9967303E-1
7.6390142E+3 1.0013649E+0
7.6434209E+3 1.0028680E+0
7.6478511E+3 1.0028737E+0
7.6522988E+3 1.0020018E+0
7.6567632E+3 1.0015365E+0
7.6612510E+3 1.0006471E+0
7.6657559E+3 9.9999106E-1
7.6702842E+3 9.9914271E-1
7.6748242E+3 9.9843129E-1
7.6793872E+3 9.9779200E-1
7.6839678E+3 9.9752337E-1
7.6885718E+3 9.9705725E-1
7.6931880E+3 9.9697024E-1
7.6978281E+3 9.9713000E-1
7.7024858E+3 9.9818392E-1
7.7071611E+3 9.9874514E-1
7.7118608E+3 9.9985166E-1
7.7165718E+3 1.0007250E+0
7.7213081E+3 1.0009611E+0
7.7260610E+3 1.0014117E+0
7.7308330E+3 1.0017669E+0
7.7356230E+3 1.0018652E+0
7.7404321E+3 1.0016020E+0
7.7452651E+3 1.0008880E+0
7.7501099E+3 1.0000356E+0
7.7549800E+3 1.0000037E+0
7.7598628E+3 9.9916070E-1
7.7647700E+3 9.9864591E-1
7.7696958E+3 9.9802107E-1
7.7746401E+3 9.9745190E-1
7.7795981E+3 9.9710041E-1
7.7845801E+3 9.9725659E-1
7.7895820E+3 9.9780275E-1
7.7946021E+3 9.9810993E-1
7.7996421E+3 9.9906319E-1
7.8047012E+3 9.9948455E-1
7.8097729E+3 1.0004267E+0
7.8148701E+3 1.0014844E+0
7.8199868E+3 1.0021962E+0
7.8251172E+3 1.0029380E+0
7.8302729E+3 1.0032835E+0
7.8354419E+3 1.0033661E+0
7.8406372E+3 1.0033166E+0
7.8458462E+3 1.0027527E+0
7.8510742E+3 1.0019864E+0
7.8563218E+3 1.0017713E+0
7.8615908E+3 1.0011680E+0
7.8668789E+3 1.0005971E+0
7.8721812E+3 1.0001833E+0
7.8775088E+3 9.9985657E-1
-------------- next part --------------
0.0000000E+0 0.0000000E+0
5.0000000E-2 0.0000000E+0
1.0000000E-1 0.0000000E+0
1.5000000E-1 0.0000000E+0
2.0000000E-1 0.0000000E+0
2.5000000E-1 0.0000000E+0
3.0000000E-1 0.0000000E+0
3.5000000E-1 0.0000000E+0
4.0000000E-1 0.0000000E+0
4.5000000E-1 0.0000000E+0
5.0000000E-1 0.0000000E+0
5.5000000E-1 0.0000000E+0
6.0000000E-1 0.0000000E+0
6.5000000E-1 0.0000000E+0
7.0000000E-1 0.0000000E+0
7.5000000E-1 0.0000000E+0
8.0000000E-1 0.0000000E+0
8.5000000E-1 0.0000000E+0
9.0000000E-1 0.0000000E+0
9.5000000E-1 0.0000000E+0
1.0000000E+0 0.0000000E+0
1.0500000E+0 0.0000000E+0
1.1000000E+0 0.0000000E+0
1.1500000E+0 0.0000000E+0
1.2000000E+0 0.0000000E+0
1.2500000E+0 0.0000000E+0
1.3000000E+0 0.0000000E+0
1.3500000E+0 0.0000000E+0
1.4000000E+0 0.0000000E+0
1.4500000E+0 0.0000000E+0
1.5000000E+0 0.0000000E+0
1.5500000E+0 0.0000000E+0
1.6000000E+0 0.0000000E+0
1.6500000E+0 0.0000000E+0
1.7000000E+0 0.0000000E+0
1.7500000E+0 0.0000000E+0
1.8000000E+0 0.0000000E+0
1.8500000E+0 0.0000000E+0
1.9000000E+0 0.0000000E+0
1.9500000E+0 0.0000000E+0
2.0000000E+0 0.0000000E+0
2.0500000E+0 0.0000000E+0
2.1000000E+0 0.0000000E+0
2.1500000E+0 2.5069672E-1
2.2000000E+0 2.0779410E-1
2.2500000E+0 1.6834500E-1
2.3000000E+0 1.3647574E-1
2.3500000E+0 1.1237556E-1
2.4000000E+0 9.0352726E-2
2.4500000E+0 6.8237947E-2
2.5000000E+0 4.4565368E-2
2.5500000E+0 1.7522238E-2
2.6000000E+0 -1.0384693E-2
2.6500000E+0 -3.6815998E-2
2.7000000E+0 -6.1081139E-2
2.7500000E+0 -8.6058312E-2
2.8000000E+0 -1.1080324E-1
2.8500000E+0 -1.3399236E-1
2.9000000E+0 -1.5292497E-1
2.9500000E+0 -1.6551279E-1
3.0000000E+0 -1.7228540E-1
3.0500000E+0 -1.7364540E-1
3.1000000E+0 -1.7014174E-1
3.1500000E+0 -1.6306042E-1
3.2000000E+0 -1.5317213E-1
3.2500000E+0 -1.3959761E-1
3.3000000E+0 -1.2311687E-1
3.3500000E+0 -1.0429919E-1
3.4000000E+0 -8.3615344E-2
3.4500000E+0 -6.2051067E-2
3.5000000E+0 -4.0376285E-2
3.5500000E+0 -1.8309759E-2
3.6000000E+0 6.2880464E-3
3.6500000E+0 2.8543842E-2
3.7000000E+0 4.2583213E-2
3.7500000E+0 5.3713622E-2
3.8000000E+0 6.2705139E-2
3.8500000E+0 6.8238449E-2
3.9000000E+0 7.1858360E-2
3.9500000E+0 7.4323467E-2
4.0000000E+0 7.6623013E-2
4.0500000E+0 7.9136644E-2
4.1000000E+0 8.2533170E-2
4.1500000E+0 8.5500546E-2
4.2000000E+0 8.7096129E-2
4.2500000E+0 8.6643209E-2
4.3000000E+0 8.3880485E-2
4.3500000E+0 7.9139823E-2
4.4000000E+0 7.2728541E-2
4.4500000E+0 6.4193348E-2
4.5000000E+0 5.2928906E-2
4.5500000E+0 3.9287986E-2
4.6000000E+0 2.4715883E-2
4.6500000E+0 1.0611448E-2
4.7000000E+0 -2.5219640E-3
4.7500000E+0 -1.4094699E-2
4.8000000E+0 -2.4570619E-2
4.8500000E+0 -3.3650201E-2
4.9000000E+0 -4.0504443E-2
4.9500000E+0 -4.4734758E-2
5.0000000E+0 -4.5744215E-2
5.0500000E+0 -4.4400000E-2
5.1000000E+0 -4.1165276E-2
5.1500000E+0 -3.6995066E-2
5.2000000E+0 -3.2248298E-2
5.2500000E+0 -2.7655850E-2
5.3000000E+0 -2.3723417E-2
5.3500000E+0 -2.1589429E-2
5.4000000E+0 -2.1312801E-2
5.4500000E+0 -2.1838962E-2
5.5000000E+0 -2.2152481E-2
5.5500000E+0 -2.2917641E-2
5.6000000E+0 -2.3629425E-2
5.6500000E+0 -2.3578288E-2
5.7000000E+0 -2.1676407E-2
5.7500000E+0 -1.8213828E-2
5.8000000E+0 -1.3891939E-2
5.8500000E+0 -9.0864700E-3
5.9000000E+0 -4.7752565E-3
5.9500000E+0 2.7324018E-5
6.0000000E+0 4.9209173E-3
6.0500000E+0 9.5288391E-3
6.1000000E+0 1.4280764E-2
6.1500000E+0 1.9581377E-2
6.2000000E+0 2.4551913E-2
6.2500000E+0 2.8632830E-2
6.3000000E+0 3.2384119E-2
6.3500000E+0 3.4926385E-2
6.4000000E+0 3.5278156E-2
6.4500000E+0 3.3577371E-2
6.5000000E+0 3.0220046E-2
6.5500000E+0 2.4967395E-2
6.6000000E+0 1.8441727E-2
6.6500000E+0 1.1235519E-2
6.7000000E+0 3.7903975E-3
6.7500000E+0 -2.8131243E-3
6.8000000E+0 -8.3193828E-3
6.8500000E+0 -1.2780425E-2
6.9000000E+0 -1.5576675E-2
6.9500000E+0 -1.6647718E-2
7.0000000E+0 -1.6079537E-2
7.0500000E+0 -1.5186283E-2
7.1000000E+0 -1.4130744E-2
7.1500000E+0 -1.2630329E-2
7.2000000E+0 -1.0925775E-2
7.2500000E+0 -8.7693334E-3
7.3000000E+0 -6.6848735E-3
7.3500000E+0 -4.6246213E-3
7.4000000E+0 -3.6596730E-3
7.4500000E+0 -2.9021575E-3
7.5000000E+0 -2.6234797E-3
7.5500000E+0 -3.2309885E-3
7.6000000E+0 -3.1435682E-3
7.6500000E+0 -2.9971895E-3
7.7000000E+0 -3.0139932E-3
7.7500000E+0 -2.7726986E-3
7.8000000E+0 -2.8213434E-3
7.8500000E+0 -2.3069846E-3
7.9000000E+0 -1.4311136E-3
7.9500000E+0 -1.2620130E-3
8.0000000E+0 -1.0123717E-3
8.0500000E+0 -7.3398373E-4
8.1000000E+0 -4.8484649E-4
8.1500000E+0 7.4017549E-4
8.2000000E+0 2.1382622E-3
8.2500000E+0 4.6313069E-3
8.3000000E+0 6.8218266E-3
8.3500000E+0 8.8825873E-3
8.4000000E+0 1.0487878E-2
8.4500000E+0 1.2310788E-2
8.5000000E+0 1.1888282E-2
8.5500000E+0 1.1474629E-2
8.6000000E+0 1.0323438E-2
8.6500000E+0 8.7928551E-3
8.7000000E+0 6.6145944E-3
8.7500000E+0 4.6120342E-3
8.8000000E+0 1.5385333E-3
8.8500000E+0 -1.2390167E-3
8.9000000E+0 -4.0396845E-3
8.9500000E+0 -6.6192309E-3
9.0000000E+0 -8.4681131E-3
9.0500000E+0 -9.5260783E-3
9.1000000E+0 -1.0346974E-2
9.1500000E+0 -1.0581138E-2
9.2000000E+0 -9.8482085E-3
9.2500000E+0 -8.1636319E-3
9.3000000E+0 -6.8406266E-3
9.3500000E+0 -4.1409325E-3
9.4000000E+0 -1.9457665E-3
9.4500000E+0 3.5031377E-4
9.5000000E+0 2.3243180E-3
9.5500000E+0 3.9014845E-3
9.6000000E+0 4.7325755E-3
9.6500000E+0 4.8080163E-3
9.7000000E+0 4.5247248E-3
9.7500000E+0 3.2617296E-3
9.8000000E+0 2.2176555E-3
9.8500000E+0 8.9758917E-4
9.9000000E+0 -6.2211501E-4
9.9500000E+0 -9.8289066E-4
1.0000000E+1 -1.3792429E-3
1.0050000E+1 -1.9086467E-3
1.0100000E+1 -1.9669103E-3
1.0150000E+1 -1.8002952E-3
1.0200000E+1 -1.9205676E-3
1.0250000E+1 -9.9111351E-4
1.0300000E+1 -3.2451014E-5
1.0350000E+1 6.6958861E-4
1.0400000E+1 1.8568619E-3
1.0450000E+1 2.8814923E-3
1.0500000E+1 3.5800622E-3
1.0550000E+1 3.7887531E-3
1.0600000E+1 4.2282850E-3
1.0650000E+1 4.3013847E-3
1.0700000E+1 3.9503569E-3
1.0750000E+1 3.3205990E-3
1.0800000E+1 2.3326630E-3
1.0850000E+1 1.0070880E-3
1.0900000E+1 -2.9317409E-4
1.0950000E+1 -1.8905099E-3
1.1000000E+1 -3.4885853E-3
1.1050000E+1 -4.6099469E-3
1.1100000E+1 -5.2522496E-3
1.1150000E+1 -5.2275401E-3
1.1200000E+1 -5.2255776E-3
1.1250000E+1 -4.8952026E-3
1.1300000E+1 -3.6932045E-3
1.1350000E+1 -2.0935859E-3
1.1400000E+1 -6.6194905E-4
1.1450000E+1 5.9690855E-4
1.1500000E+1 1.5871385E-3
1.1550000E+1 1.8664235E-3
1.1600000E+1 1.3530660E-3
1.1650000E+1 2.9740643E-3
1.1700000E+1 3.9203033E-3
1.1750000E+1 3.5859766E-3
1.1800000E+1 2.8478486E-3
1.1850000E+1 2.2028666E-3
1.1900000E+1 1.3807331E-3
1.1950000E+1 6.2797141E-4
1.2000000E+1 -1.9259811E-4
1.2050000E+1 -9.0238430E-4
1.2100000E+1 -1.4229546E-3
1.2150000E+1 -1.7985056E-3
1.2200000E+1 -2.1431819E-3
1.2250000E+1 -2.1614564E-3
1.2300000E+1 -1.6690045E-3
1.2350000E+1 -8.4831206E-4
1.2400000E+1 -9.6239249E-5
1.2450000E+1 8.7848738E-4
1.2500000E+1 1.4466574E-3
1.2550000E+1 1.7313968E-3
1.2600000E+1 2.0982021E-3
1.2650000E+1 2.2977652E-3
1.2700000E+1 2.1916150E-3
1.2750000E+1 1.6813328E-3
1.2800000E+1 8.5169744E-4
1.2850000E+1 3.0173965E-4
1.2900000E+1 -1.4562779E-4
1.2950000E+1 -9.0459174E-4
1.3000000E+1 -1.5311589E-3
1.3050000E+1 -2.1960625E-3
1.3100000E+1 -2.7343104E-3
1.3150000E+1 -2.9239768E-3
1.3200000E+1 -2.6595139E-3
1.3250000E+1 -2.2942391E-3
1.3300000E+1 -1.7633157E-3
1.3350000E+1 -1.1315809E-3
1.3400000E+1 -5.4600282E-4
1.3450000E+1 3.7060907E-4
1.3500000E+1 1.1831368E-3
1.3550000E+1 1.8240318E-3
1.3600000E+1 2.2901451E-3
1.3650000E+1 2.4196880E-3
1.3700000E+1 2.3457258E-3
1.3750000E+1 1.9532791E-3
1.3800000E+1 1.1929978E-3
1.3850000E+1 5.9288860E-4
1.3900000E+1 8.6049582E-5
1.3950000E+1 -6.0651756E-4
1.4000000E+1 -1.2080115E-3
1.4050000E+1 -1.6886822E-3
-------------- next part --------------
2.1238099E+0 2.6283996E+0
2.1652771E+0 2.3801796E+0
2.2070751E+0 2.1392322E+0
2.2492673E+0 1.9144751E+0
2.2918262E+0 1.7149962E+0
2.3347713E+0 1.5274819E+0
2.3779708E+0 1.3489049E+0
2.4212480E+0 1.1449513E+0
2.4648553E+0 9.0939488E-1
2.5087125E+0 6.1182981E-1
2.5528943E+0 2.5109186E-1
2.5972677E+0 -1.4476572E-1
2.6417182E+0 -5.7887724E-1
2.6863498E+0 -1.0346130E+0
2.7315202E+0 -1.5540212E+0
2.7763731E+0 -2.1217464E+0
2.8217172E+0 -2.7336794E+0
2.8672673E+0 -3.3511406E+0
2.9127471E+0 -3.9023521E+0
2.9584580E+0 -4.3461430E+0
3.0043545E+0 -4.6943049E+0
3.0502512E+0 -4.9346895E+0
3.0964882E+0 -5.0760424E+0
3.1426610E+0 -5.1142608E+0
3.1890489E+0 -5.0719163E+0
3.2355695E+0 -4.8941068E+0
3.2820366E+0 -4.5931407E+0
3.3286858E+0 -4.1636383E+0
3.3754901E+0 -3.6246723E+0
3.4224046E+0 -2.9582225E+0
3.4692616E+0 -2.2334858E+0
3.5164642E+0 -1.4281666E+0
3.5634001E+0 -5.3262899E-1
3.6106886E+0 5.8577406E-1
3.6579095E+0 1.5971531E+0
3.7053957E+0 2.2561892E+0
3.7526951E+0 2.8951234E+0
3.8002826E+0 3.4523596E+0
3.8478017E+0 3.8712678E+0
3.8954341E+0 4.2291747E+0
3.9431758E+0 4.5330329E+0
3.9908776E+0 4.8410800E+0
4.0386682E+0 5.1570752E+0
4.0865797E+0 5.5678929E+0
4.1344298E+0 5.9901851E+0
4.1823810E+0 6.3609813E+0
4.2304393E+0 6.6128186E+0
4.2784410E+0 6.7082186E+0
4.3265341E+0 6.6191662E+0
4.3747457E+0 6.3981543E+0
4.4228814E+0 6.0011758E+0
4.4711173E+0 5.3717277E+0
4.5194649E+0 4.4282728E+0
4.5677504E+0 3.2448368E+0
4.6161015E+0 1.9504632E+0
4.6644228E+0 6.5350501E-1
4.7128467E+0 -6.2814425E-1
4.7613589E+0 -1.8118818E+0
4.8098116E+0 -2.9788488E+0
4.8583341E+0 -4.0445414E+0
4.9068490E+0 -4.9047530E+0
4.9554232E+0 -5.5038055E+0
5.0040890E+0 -5.7396871E+0
5.0527213E+0 -5.7117444E+0
5.1014295E+0 -5.4420508E+0
5.1500887E+0 -5.0523868E+0
5.1988270E+0 -4.5605861E+0
5.2475392E+0 -4.0376293E+0
5.2963379E+0 -3.5566806E+0
5.3451962E+0 -3.3044585E+0
5.3939000E+0 -3.3320255E+0
5.4428261E+0 -3.5157853E+0
5.4916952E+0 -3.6463341E+0
5.5405468E+0 -3.8701653E+0
5.5894567E+0 -4.1114230E+0
5.6384563E+0 -4.2880218E+0
5.6874064E+0 -4.1452222E+0
5.7363314E+0 -3.6494581E+0
5.7853412E+0 -2.9891798E+0
5.8343252E+0 -2.0670307E+0
5.8833625E+0 -1.2951112E+0
5.9324051E+0 -3.6977230E-1
5.9815079E+0 6.8029988E-1
6.0305734E+0 1.6989305E+0
6.0796112E+0 2.7418642E+0
6.1287422E+0 3.9541973E+0
6.1778342E+0 5.3520536E+0
6.2270044E+0 6.4696573E+0
6.2761502E+0 7.5931630E+0
6.3253889E+0 8.6448369E+0
6.3744807E+0 9.2446134E+0
6.4237673E+0 9.2493343E+0
6.4729186E+0 8.7932129E+0
6.5221518E+0 7.8814218E+0
6.5713488E+0 6.3302211E+0
6.6206456E+0 4.5358854E+0
6.6699256E+0 2.4387511E+0
6.7191793E+0 2.8135574E-1
6.7685026E+0 -1.5785821E+0
6.8178264E+0 -3.2249122E+0
6.8671066E+0 -4.5957881E+0
6.9163879E+0 -5.3868111E+0
6.9657534E+0 -5.6860033E+0
7.0150945E+0 -5.4352054E+0
7.0644171E+0 -5.2706142E+0
7.1138310E+0 -4.9696439E+0
7.1632386E+0 -4.4800551E+0
7.2126089E+0 -3.9317197E+0
7.2619719E+0 -3.1413335E+0
7.3113438E+0 -2.4289664E+0
7.3607870E+0 -1.6609074E+0
7.4102309E+0 -1.4334187E+0
7.4596472E+0 -1.1393815E+0
7.5090469E+0 -1.0986878E+0
7.5585489E+0 -1.4551974E+0
7.6079590E+0 -1.3640312E+0
7.6574424E+0 -1.3373248E+0
7.7069259E+0 -1.3825174E+0
7.7564095E+0 -1.2755069E+0
7.8058763E+0 -1.3479795E+0
7.8553501E+0 -1.0846283E+0
7.9048958E+0 -6.5965393E-1
7.9543514E+0 -6.3118878E-1
8.0039127E+0 -5.0739096E-1
8.0534633E+0 -3.7216107E-1
8.1030035E+0 -2.4902846E-1
8.1524720E+0 4.4242572E-1
8.2020275E+0 1.2174253E+0
8.2515807E+0 2.6590968E+0
8.3011330E+0 3.9408809E+0
8.3506973E+0 5.1992599E+0
8.4002544E+0 6.2278749E+0
8.4499152E+0 7.4356427E+0
8.4994842E+0 7.3039764E+0
8.5490722E+0 7.1884685E+0
8.5986531E+0 6.5960381E+0
8.6482527E+0 5.7471734E+0
8.6978586E+0 4.4160279E+0
8.7474692E+0 3.2056634E+0
8.7970935E+0 1.1690259E+0
8.8467297E+0 -7.1559157E-1
8.8962917E+0 -2.6957029E+0
8.9459516E+0 -4.6198879E+0
8.9956243E+0 -6.0902426E+0
9.0452163E+0 -6.9843944E+0
9.0949115E+0 -7.7565227E+0
9.1445426E+0 -8.1429779E+0
9.1941908E+0 -7.8210823E+0
9.2438572E+0 -6.5585137E+0
9.2935289E+0 -5.7934935E+0
9.3431436E+0 -3.6267688E+0
9.3928651E+0 -1.8992236E+0
9.4425213E+0 3.1762168E-2
9.4921897E+0 1.7551041E+0
9.5418770E+0 3.2474797E+0
9.5915869E+0 4.1605730E+0
9.6413206E+0 4.3168484E+0
9.6910055E+0 4.3392894E+0
9.7407074E+0 3.1907387E+0
9.7904260E+0 2.3001686E+0
9.8400969E+0 1.2051319E+0
9.8897965E+0 -5.2777235E-1
9.9395206E+0 -8.9617936E-1
9.9892660E+0 -1.2375543E+0
1.0038958E+1 -1.8908150E+0
1.0088763E+1 -2.1014640E+0
1.0138449E+1 -1.7714768E+0
1.0188234E+1 -2.2508472E+0
1.0237964E+1 -1.3525580E+0
1.0287662E+1 -1.7689848E-1
1.0337449E+1 3.9515025E-1
1.0387202E+1 1.7732194E+0
1.0436916E+1 2.9988097E+0
1.0486663E+1 4.1010205E+0
1.0536438E+1 4.2632809E+0
1.0586240E+1 4.9474607E+0
1.0635942E+1 5.2675046E+0
1.0685743E+1 5.0126347E+0
1.0735518E+1 4.4056820E+0
1.0785325E+1 3.4366969E+0
1.0835087E+1 1.7419281E+0
1.0884828E+1 2.1818736E-1
1.0934607E+1 -1.7510081E+0
1.0984414E+1 -4.1276487E+0
1.1034198E+1 -5.7816075E+0
1.1084015E+1 -7.1701442E+0
1.1133805E+1 -7.2106644E+0
1.1183569E+1 -7.3207984E+0
1.1233363E+1 -7.3838430E+0
1.1283213E+1 -6.1360460E+0
1.1332966E+1 -3.7341457E+0
1.1382820E+1 -1.7523967E+0
1.1432586E+1 4.9284006E-1
1.1482398E+1 1.6522022E+0
1.1532247E+1 3.8218174E+0
1.1582020E+1 1.1569683E+0
1.1631895E+1 3.8228281E+0
1.1681678E+1 6.2533807E+0
1.1731513E+1 6.3226955E+0
1.1781333E+1 4.9529901E+0
1.1831129E+1 4.2184236E+0
1.1880975E+1 2.7493712E+0
1.1930803E+1 1.6361443E+0
1.1980680E+1 1.6021317E-1
1.2030478E+1 -1.1173268E+0
1.2080322E+1 -2.3026596E+0
1.2130152E+1 -2.8582370E+0
1.2180032E+1 -3.7876142E+0
1.2229839E+1 -4.0484935E+0
1.2279701E+1 -3.8613172E+0
1.2329548E+1 -1.9950580E+0
1.2379383E+1 -1.0217681E+0
1.2429276E+1 1.0194478E+0
1.2479089E+1 2.6528154E+0
1.2528969E+1 3.0671045E+0
1.2578826E+1 3.9141237E+0
1.2628685E+1 4.5844029E+0
1.2678535E+1 4.7416402E+0
1.2728387E+1 4.1519312E+0
1.2778291E+1 2.6020919E+0
1.2828122E+1 7.1851677E-1
1.2878019E+1 5.3888641E-1
1.2927853E+1 -1.4024492E+0
1.2977743E+1 -2.6760630E+0
1.3027630E+1 -4.2268110E+0
1.3077513E+1 -5.6930518E+0
1.3127344E+1 -6.7042970E+0
1.3177226E+1 -6.5813938E+0
1.3227118E+1 -5.5574306E+0
1.3277004E+1 -5.0735922E+0
1.3326899E+1 -3.0546043E+0
1.3376796E+1 -2.2675437E+0
1.3426631E+1 -2.0774862E-1
1.3476531E+1 2.1747807E+0
1.3526437E+1 3.7489992E+0
1.3576292E+1 5.4271072E+0
1.3626210E+1 6.1333511E+0
1.3676073E+1 6.1766218E+0
1.3726008E+1 5.8722122E+0
1.3775892E+1 4.2111411E+0
1.3825777E+1 1.9726238E+0
1.3875669E+1 1.1494081E+0
1.3925585E+1 -7.4486891E-1
1.3975503E+1 -2.5982937E+0
1.4025376E+1 -4.0646689E+0
1.4075309E+1 -5.3262561E+0
More information about the Ifeffit
mailing list