[Ifeffit] Is good EXAFS data possible with 0.01 transmission coefficient

Scott Calvin scalvin at sarahlawrence.edu
Thu Jul 25 15:33:40 CDT 2013


Hi Damon,

So you're saying you expect the total absorption of the sample to be less than 0.01 absorption lengths? In other words, if the detectors and electronics were identical, you would expect It = I0 exp(-0.01)?

While possible to measure in transmission, such a sample would normally be measured in fluorescence. If done that way, the measurement is routine. 

Or am I misunderstanding your description? (I am not entirely sure what you mean by "transmission coefficient.")

--Scott Calvin
Sarah Lawrence College

On Jul 25, 2013, at 4:07 PM, Damon Turney <dturney at ccny.cuny.edu> wrote:

> Hi IFEFFIT community,
> 
> I will make measurements soon at Brookhaven's NSLS to do EXAFS, and my
> sample is a powder mixture of carbon powder, MnO2 powder, and
> potassium hydroxide liquid.  The transmission coefficient of 6.5keV
> x-rays (the Mn K edge is at ~6.5keV) through the sample is ~0.001 to
> 0.01.  I am told that the ionization chamber detectors can easily
> detect the x-ray beam after this ~0.01 transmission loss (by
> increasing the gain on the detector), but I would like to ask the
> IFEFFIT community if there will be other problems with the EXAFS
> technique when the transmission coefficient is so low.  Does anybody
> have comments?
> 
> Much thanks -- if you have any info I greatly appreciate it!
> Damon Turney
> City College of New York