[Ifeffit] RE: About Cu2+ xafs
Scott Calvin
SCalvin at slc.edu
Sun Feb 13 18:10:48 CST 2005
Hi Igor (and Anatoly),
I could be wrong, but if I recall correctly a self-consistent field
calculation by FEFF8 [i]calculates[/i] what it thinks E0 should be
relative to the derivative peak, and then uses that as the "new"
baseline E0. In other words, a d-E0 found via the SCF FEFF
calculation is relative to a different starting point! You can figure
out what the SCF FEFF calculation assigned as the baseline E0 by
looking at the file xmu.dat and seeing what energy k = 0 corresponds
to.
--Scott Calvin
Sarah Lawrence College
>
>
>The second problem is of philosophical nature and more important
>than the first one because its clarification will provide the answer
>for the first problem and also some peace of mind. I do not
>understand why the E0 corrections (d_E0's) are so different. The
>fact that the correction obtained from SCF calculation is larger
>(and apparently worse) than that obtained from the nSCF calculation
>(absolute values) seems counter intuitive to me. The E0 value I
>picked for background subtraction is 8989 eV, and applying the
>corrections, I find an edge energy of 8990.1 eV and 8981.5 eV
>respectively. This second value falls in the pre-edge region,
>pretty much on the 1s-3d transition bump. Despite the evidences, I
>tend to trust the SCF's d_E0 value the most but would like to
>understand the physics (or lack of) behind the different feff
>calculations!
>
>I apologize if I bored you with unecessary details! Thank you very
>much for your attention and willingness to help! I really appreciate
>it!
>
>Best regards,
>
>Igor
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://millenia.cars.aps.anl.gov/pipermail/ifeffit/attachments/20050213/23b5c5dd/attachment.html>