Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you

Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
--
Carlo Segre, Duchossois Professor of Physics
Illinois Institute of Technology
________________________________
From: Helen Chen via Ifeffit

You can also use a cryostat with only vacuum and no cooling, if desired. Good luck! Mike
On Mar 23, 2025, at 08:16, Carlo Segre via Ifeffit
wrote: This Message Is From an External Sender This message came from outside your organization. Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
-- Carlo Segre, Duchossois Professor of Physics Illinois Institute of Technology From: Helen Chen via Ifeffit
Sent: Tuesday, March 18, 2025 8:05 PM To: ifeffit@millenia.cars.aps.anl.gov Cc: Helen Chen Subject: [Ext][Ifeffit] Question about Air sensitive sample holder This Message Is From an External Sender This message came from outside your organization. Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you
ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov

Hi Helen, As an alternative, for a highly air-sensitive material, you might consider a small can that can be sealed under inert atmosphere. APS Sector 20 (now 25ID) had, and may still have, a small Be dome. You could use a small aluminium can, with windows that are less permeable to air/O2/CO2/H20...e.g. Be sheet, diamond, maybe even graphite. (not under vacuum), maybe even thin aluminium. If you use a slotted holder for the sample, you can use that for alignment if you can't see the sample. For shipping, a mason jar also sealed under inert atmosphere works well. -R. On 2025-03-22 12:16 p.m., Carlo Segre via Ifeffit wrote:
Hi Helen, We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. ZjQcmQRYFpfptBannerEnd Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
-- Carlo Segre, Duchossois Professor of Physics Illinois Institute of Technology ------------------------------------------------------------------------ *From:* Helen Chen via Ifeffit
*Sent:* Tuesday, March 18, 2025 8:05 PM *To:* ifeffit@millenia.cars.aps.anl.gov *Cc:* Helen Chen *Subject:* [Ext][Ifeffit] Question about Air sensitive sample holder Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. ZjQcmQRYFpfptBannerEnd Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you ifeffit mailing list:https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... to unsubscribe, send mail toifeffit-leave@millenia.cars.aps.anl.gov

Hi Helen, +1 on using Aluminized Mylar and Aluminized Kapton (and we have some of each at sector 13 if you want to borrow some). The aluminum layer on Mylar is very thin (you can see visible light through) but it holds helium. In an undulator beam, Mylar might last 10 seconds. The thin aluminum layer helps a lot, but I would not guarantee this by itself for multiple hours. Kapton is leaky but holds up better in an undulator beam (but you can burn a micro-hole in it over several hours with 10^13 photons/sec in a few square microns: loving APS-U!) The Aluminized version is a non-trivial thickness (several microns) of aluminum, so not suitable for tender beams, but should be OK above 5 keV (wait, which "V" in VTC -- Yb K or L3 edge?). I'll say that we have had decent luck (with old APS undulator) with samples in an otherwise regular Ziplock bag with a Kapton film window (held on with carefully cut packing tape) for X-rays in and out and over-pressuring the bag with blow-off N2 (which is free) or helium (which is not). But, oxygen-sensitivity can vary a lot, so I’m not sure there is a single “best” solution. --Matt On Sat, Mar 22, 2025 at 3:06 PM matthew marcus via Ifeffit < ifeffit@millenia.cars.aps.anl.gov> wrote:
Aluminized mylar has a much lower permeability to oxygen than polyethylene and still works with a heat-sealer. I'm assuming you're in the hard- or tender X-ray range. mam On 3/22/2025 12: 16 PM, Carlo Segre via Ifeffit wrote: > Hi Helen, We ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd
Aluminized mylar has a much lower permeability to oxygen than polyethylene and still works with a heat-sealer. I'm assuming you're in the hard- or tender X-ray range. mam
On 3/22/2025 12:16 PM, Carlo Segre via Ifeffit wrote:
Hi Helen, We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. ZjQcmQRYFpfptBannerEnd Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
-- Carlo Segre, Duchossois Professor of Physics Illinois Institute of Technology ------------------------------------------------------------------------ *From:* Helen Chen via Ifeffit
*Sent:* Tuesday, March 18, 2025 8:05 PM *To:* ifeffit@millenia.cars.aps.anl.gov *Cc:* Helen Chen *Subject:* [Ext][Ifeffit] Question about Air sensitive sample holder Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization. ZjQcmQRYFpfptBannerEnd Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov
ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov
-- --Matt Newville <newville at cars.uchicago.edu> 630-327-7411

Hi,
We often use transmission gas cells with continuous gas flow for measurements in the soft X-ray range, and we are wondering if these cells are also a good option for measuring air/moisture-sensitive samples in the hard X-ray range.
Two ~150 nm SiN windows isolate the sample and the gas from the environment, and we have never had problems with X-ray-induced window damage or leaks. Even with continuous helium flow, the pressure in the chamber was <1E-6 mbar.
A single window is sufficient for emission experiments and the design can be adapted to allow for large take-off angles.
Does somebody already have experience using cells with SiN/SiC membranes in the hard X-ray range and know what the challenges or limitations are?
Best
Richard
On 24/03/2025 20:35, Matt Newville via Ifeffit wrote:
Hi Helen,
+1 on using Aluminized Mylar and Aluminized Kapton (and we have some of
each at sector 13 if you want to borrow some).
The aluminum layer on Mylar is very thin (you can see visible light
through) but it holds helium. In an undulator beam, Mylar might last 10
seconds. The thin aluminum layer helps a lot, but I would not guarantee
this by itself for multiple hours.
Kapton is leaky but holds up better in an undulator beam (but you can burn
a micro-hole in it over several hours with 10^13 photons/sec in a few
square microns: loving APS-U!) The Aluminized version is a non-trivial
thickness (several microns) of aluminum, so not suitable for tender beams,
but should be OK above 5 keV (wait, which "V" in VTC -- Yb K or L3 edge?).
I'll say that we have had decent luck (with old APS undulator) with samples
in an otherwise regular Ziplock bag with a Kapton film window (held on with
carefully cut packing tape) for X-rays in and out and over-pressuring the
bag with blow-off N2 (which is free) or helium (which is not). But,
oxygen-sensitivity can vary a lot, so I’m not sure there is a single “best”
solution.
--Matt
On Sat, Mar 22, 2025 at 3:06 PM matthew marcus via Ifeffit <
ifeffit@millenia.cars.aps.anl.govmailto:ifeffit@millenia.cars.aps.anl.gov> wrote:
Aluminized mylar has a much lower permeability to oxygen than polyethylene
and still works with a heat-sealer. I'm assuming you're in the hard- or
tender X-ray range. mam On 3/22/2025 12: 16 PM, Carlo Segre via Ifeffit
wrote: > Hi Helen, We
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd
Aluminized mylar has a much lower permeability to oxygen than
polyethylene and still works with a heat-sealer. I'm assuming you're in
the hard- or tender X-ray range.
mam
On 3/22/2025 12:16 PM, Carlo Segre via Ifeffit wrote:
Hi Helen, We have had luck using a vacuum sealing system such as used
for food. However the Polyethylene is subject to degradation under the
beam so that wrapping with Kapton is helpful as well as double bagging
it. Finally, if you keep it under
ZjQcmQRYFpfptBannerStart
This Message Is From an External Sender
This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd
Hi Helen,
We have had luck using a vacuum sealing system such as used for food.
However the Polyethylene is subject to degradation under the beam so
that wrapping with Kapton is helpful as well as double bagging it.
Finally, if you keep it under Argon when transporting it to the beamline
and then in a helium filled chamber, you should avoid the oxidation.
Carlo
--
Carlo Segre, Duchossois Professor of Physics
Illinois Institute of Technology
------------------------------------------------------------------------
*From:* Helen Chen via Ifeffit

Hello again,
I have a follow up question about handling air sensitive sample. Some of
you mentioned vacuum sealing like which used for food, this reminds me a
aluminized mylar bags. Does anyone know a good source to get thine
aluminized mylar bags. It seems that the standard thickness is about 0.1mm,
just wondering if there is a well known source who makes bags thinner than
that.
Thank you!
Best,
Helen
Mike Massey
You can also use a cryostat with only vacuum and no cooling, if desired. Good luck! Mike On Mar 23, 2025, at 08: 16, Carlo Segre via Ifeffit
wrote: Hi Helen, We have had luck using a vacuum sealing ZjQcmQRYFpfptBannerStart This Message Is From an Untrusted Sender You have not previously corresponded with this sender. See https://urldefense.us/v3/__https://itconnect.uw.edu/email-tags__;!!G_uCfscf7... for additional information. Please contact the UW-IT Service Center, help@uw.edu 206.221.5000, for assistance. ZjQcmQRYFpfptBannerEnd You can also use a cryostat with only vacuum and no cooling, if desired.
Good luck!
Mike
On Mar 23, 2025, at 08:16, Carlo Segre via Ifeffit < ifeffit@millenia.cars.aps.anl.gov> wrote:
Hi Helen, We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
-- Carlo Segre, Duchossois Professor of Physics Illinois Institute of Technology ------------------------------ *From:* Helen Chen via Ifeffit
*Sent:* Tuesday, March 18, 2025 8:05 PM *To:* ifeffit@millenia.cars.aps.anl.gov *Cc:* Helen Chen
*Subject:* [Ext][Ifeffit] Question about Air sensitive sample holder Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you
ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... <https://urldefense.com/v3/__https://millenia.cars.aps.anl.gov/mailman3/lists... > to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov

Uline claims to sell 2.5mil aluminized mylar pouches https://urldefense.us/v3/__https://www.uline.com/BL_5552/Metalized-Food-Bags... I've previously ordered foil pouches from Impak https://urldefense.us/v3/__https://www.impakcorporation.com/flexible_packagi... but they seem to mostly carry 4.0mil with the Al foil. There is just metallized mylar and kapton foil available thinner but you would need to make your own pouches or tape these as a window over a cutout on a thicker Al foil/mylar pouch. -Lyle On Wed, Jun 4, 2025 at 3:14 PM Helen Chen via Ifeffit < ifeffit@millenia.cars.aps.anl.gov> wrote:
Hello again, I have a follow up question about handling air sensitive sample. Some of you mentioned vacuum sealing like which used for food, this reminds me a aluminized mylar bags. Does anyone know a good source to get thine aluminized mylar ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd Hello again, I have a follow up question about handling air sensitive sample. Some of you mentioned vacuum sealing like which used for food, this reminds me a aluminized mylar bags. Does anyone know a good source to get thine aluminized mylar bags. It seems that the standard thickness is about 0.1mm, just wondering if there is a well known source who makes bags thinner than that. Thank you!
Best, Helen
Mike Massey
於 2025年3月22日 週六 下午12:42寫道: You can also use a cryostat with only vacuum and no cooling, if desired. Good luck! Mike On Mar 23, 2025, at 08: 16, Carlo Segre via Ifeffit
wrote: Hi Helen, We have had luck using a vacuum sealing ZjQcmQRYFpfptBannerStart This Message Is From an Untrusted Sender You have not previously corresponded with this sender. See https://urldefense.us/v3/__https://itconnect.uw.edu/email-tags__;!!G_uCfscf7... https://urldefense.us/v3/__https://itconnect.uw.edu/email-tags__;!!G_uCfscf7... for additional information. Please contact the UW-IT Service Center, help@uw.edu 206.221.5000, for assistance. ZjQcmQRYFpfptBannerEnd You can also use a cryostat with only vacuum and no cooling, if desired.
Good luck!
Mike
On Mar 23, 2025, at 08:16, Carlo Segre via Ifeffit < ifeffit@millenia.cars.aps.anl.gov> wrote:
Hi Helen, We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd Hi Helen,
We have had luck using a vacuum sealing system such as used for food. However the Polyethylene is subject to degradation under the beam so that wrapping with Kapton is helpful as well as double bagging it. Finally, if you keep it under Argon when transporting it to the beamline and then in a helium filled chamber, you should avoid the oxidation.
Carlo
-- Carlo Segre, Duchossois Professor of Physics Illinois Institute of Technology ------------------------------ *From:* Helen Chen via Ifeffit
*Sent:* Tuesday, March 18, 2025 8:05 PM *To:* ifeffit@millenia.cars.aps.anl.gov < ifeffit@millenia.cars.aps.anl.gov> *Cc:* Helen Chen *Subject:* [Ext][Ifeffit] Question about Air sensitive sample holder Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and ZjQcmQRYFpfptBannerStart This Message Is From an External Sender This message came from outside your organization.
ZjQcmQRYFpfptBannerEnd Hello, Our group (Jerry Seidler) just started a project involving doing VTC and HERFD on Yb doped CrX3 material. They are thin small crystal sheets and are very air sensitive. We have tried encapsulated them between a layer of laptop tape and film, but after 12hr we see degradation due to air leaks. I am looking for some advice on air sensitive sample holder. What are some typical methods you use to deal with air sensitivity, especially for samples in thin sheets geometry? Thank you
ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... <https://urldefense.com/v3/__https://millenia.cars.aps.anl.gov/mailman3/lists... > to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov
ifeffit mailing list: https://millenia.cars.aps.anl.gov/mailman3/lists/ifeffit.millenia.cars.aps.a... to unsubscribe, send mail to ifeffit-leave@millenia.cars.aps.anl.gov
participants (8)
-
Carlo Segre
-
Helen Chen
-
Lyle Gordon
-
Matt Newville
-
matthew marcus
-
Mike Massey
-
Richard Gnewkow
-
Robert Gordon