Dear Lei,
> Q1. In two beam lines, for exactly the same MoS2 crystal, we got difference in XANES region. Spectrum from DESY has a weak pre-edge peak resolved in XANES while ERSF doesn't. And the derivative shape of the two spectra for MoS2 collected from two synchrotrons are not well overlapped. Is it acceptable?
If you measured the Mo K-edge at 20000 eV at DESY you were most probably using the 311 crystals (at least that is what I or one of my colleagues here would have suggested) As Matthew already mentioned the differences in the pre-edge region might be due to differences in the energy resolution (for instance if you used 111 crystals at ESRF).
> Q2. I tried to overlay the first derivative of the two spectra of MoS2. There is a energy shift at 2.5 eV from the different synchrotrons if we take the one from DESY as standard. Can I apply the energy shift at 2.5 eV for MoS2 from ERSF and then overlay the first derivative of the other NiMo samples with MoS2 to get the energy calibration? Or should I apply 2.5 eV for all the NiMo sulfided samples?
I am not sure if I understand exactly what you want to do. Anyway, if you find a difference of 2.5 eV between the edge position measured on the same sample at different beamlines it makes sense to define the spectra from one beamline to be "correct" and shift all spectra from the other beamline by the respective energy shift (here 2.5 eV). BUT (!) you must be sure that the shift isĀ reproducible... If you measured reference spectra you can easily check if this is true, you should not observe any shift between the energy positions of you metal foil spectra. If you have no reference spectra you could trust encoder values (if you have them) or it becomes a religious thing and you just have to believe (and maybe convince others to believe too...;-)
So you checked the reproducibility of your energy axis and you can be more or less sure that the energy axis did not shift during your measurements. Fine, so there is no need to re-calibrate your one ESRF spectrum with respect to another ESRF spectrum, but if you want to compare them with the DESY spectra you still have to shift either all the DESY spectra or all the ESRF spectra by 2.5 eV. And as Matthew already mentioned you should document what you did so that someone else can trace your energy positions back to a common standard (in your case the metal foil and the usually used k-edge position of 20000.0 eV from the orange book.
> Q3. Since we lost the signal for the foil reference in ERSF, we tried to put the foil in the sample position. The first derivatives of the foil spetra from time to time overlap quite well. Should we set the energy calibration at 0 according to the stable foil signal?
Best regards
Edmund
_______________________________________________ Ifeffit mailing list Ifeffit@millenia.cars.aps.anl.gov http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
_______________________________________________
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit