Mingliang,
This includes a link to a discussion on the mail list about self-absorption and how to understand it.
-Richard
2010/1/29 Welter, Edmund
<edmund.welter@desy.de>
Dear Mingliang,
Thanks for your explain. But I am still puzzled. Following your instruction, I performed *Self-Absorption *for the samples, choosing *XANES(Fluo) *from *algorithm*, but I did not know the real meaning of *Formula*. Does it mean the species formula which I wish to detect? Or the whole composition of the sample?
The composition of the whole sample. You do not need to know the species, you "only" need to know thew elemental composition of your sample. A hairy task if you don't digest your sample completely and put it into an ICP-MS or anything like that. I don't know which precision is necessary to yield usable results. Maybe someone else tested or calculated it.
In principle the matrix produces just a smooth background absorption and knowing the composition it is easy to calculate the amount of this absorption using tabulated scattering factors. But it should also be possible to determine this background experimentally, for instance extrapolating the background before the edge. But for this you would need a blank spectrum to "calibrate" the mued value.
If you follow the link to the "document section" in Athena a very instructive webpage will open and among other things you will find citations of the original papers underlying the four different algorithms used for the self absorption correction.
I used natural mineral to interact with selenite, then I wish to check the speciation of the selenium reaction product. For my case, should I fill the blank with Se or FeSe or both of them (because currently I do not know the exactly containing speciation)?
As said before, the "chemical Formula" of the entire sample. "Chemical formula" here does not mean you have any kind of chemical bond, it just means molar ratio.
For the angles, normally are they 45? Becuase I measured the samples in fluorescence mode, and didnot measure the angles between sample, incoming beam and detector. I tried to input *Se *in the *Formula* entry, and select 45 for the angles, then I got a corrected spectra which was higher than the original one.
45 degrees is the usually used geometry. However, the program asks for the geometry which you used not for the usually used geometry! If it was 45 degrees than its fine. Obviously the algorithm does what it is supposed to do, it amplifies the oscillations. But this does not mean that the result is correct! It will only be "correct" if the input parameters were correct.
But when I using this spectra for LCF analysis, I still could not got overlapping line for the fitted spectra with the sample spectra...
Using pure Se in the self absorption correction is definitely wrong, it overestimates the effect dramatically. Actually the effect more or less disappears for diluted samples. But even if the spectra were properly corrected I can not see any reason why the used set of reference spectra MUST fit the spectrum of your sample. There are many possible reasons for this, the most likely is that you are missing one or more important reference spectrum (Or you are not using enough references/free parameters in your fit...;-)
Best regards,
Edmund Welter
_______________________________________________